Length of Time of Obstructive Apena

Post by fdw » Mon Dec 07, 2015 9:08 am

Lately I've been have periods of time in which my Obstructive Apena have lasted nearly a full minute.......Is a pressure increase necessary?

Image
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
LSAT
Posts: 13352
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Length of Time of Obstructive Apena

Post by LSAT » Mon Dec 07, 2015 9:23 am

I think I would switch over from fixed to auto with a pressure range of 12-20 to see if the pressure you are at is sufficient.

WindCpap
Posts: 394
Joined: Fri Oct 09, 2015 6:57 pm

Re: Length of Time of Obstructive Apena

Post by WindCpap » Mon Dec 07, 2015 9:53 am

Given your overall AHI is very low, there is probably a marked difference between laying on your side vs laying on your back, or REM vs NREM sleep. Putting your CPAP in auto mode would probably clear these up. If you are comfortable at 14.4, keep it as the low pressure.

_________________
MachineMask

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Length of Time of Obstructive Apena

Post by fdw » Mon Dec 07, 2015 12:58 pm

LSAT wrote:I think I would switch over from fixed to auto with a pressure range of 12-20 to see if the pressure you are at is sufficient.
I would need to know what parameters I'm looking for to find out what is sufficient.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
jnk...
Posts: 2988
Joined: Fri Sep 19, 2014 12:36 pm
Location: New York State

Re: Length of Time of Obstructive Apena

Post by jnk... » Mon Dec 07, 2015 1:14 pm

If it were me, I might bump pressure up a few tenths as a personal experiment, but I would do so based only on the fact that my airway was closing at all--not on how long it was staying closed. Sometimes those long ones have a slight central component--it is just that the airway closes during them so that it is reported as a pure obstructive event.

Alternatively, I might try, just as a test, less EPR (for example, 2, instead of 3) in order to see if that helped my average AHI over time. But I would file it all under minor tweaking and wouldn't worry much about it other than seeing if it helped my average AHI over the coming weeks. And even that wouldn't matter to me unless it helped how I felt in the mornings. Generally, an obstructive now and then won't destroy sleep. And for some, keeping pressure so high that nothing obstructive ever happens is enough to hurt sleep more than it helps it.

But hey, that's just me. And my opinions are, at times, nonstandard. Especially for this board.

Not a pro--only a guy on the Internet.
-Jeff (AS10/P30i)

Accounts to put on the foe list: Me. I often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13352
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Length of Time of Obstructive Apena

Post by LSAT » Mon Dec 07, 2015 2:02 pm

fdw wrote:
LSAT wrote:I think I would switch over from fixed to auto with a pressure range of 12-20 to see if the pressure you are at is sufficient.
I would need to know what parameters I'm looking for to find out what is sufficient.
I believe that the machine would have gone higher if it was not set on fixed pressure. By switching to auto and setting a max pressure at 20 after a few days you could see how often (if at all) the unit increased the pressure above 14.4. Then you could increase tour fixed pressure or stay on auto.
Sleepyhead graphs will show you this and help determine your "parameters".

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Length of Time of Obstructive Apena

Post by fdw » Mon Dec 07, 2015 2:35 pm

LSAT wrote:
fdw wrote:
LSAT wrote:I think I would switch over from fixed to auto with a pressure range of 12-20 to see if the pressure you are at is sufficient.
I would need to know what parameters I'm looking for to find out what is sufficient.
I believe that the machine would have gone higher if it was not set on fixed pressure. By switching to auto and setting a max pressure at 20 after a few days you could see how often (if at all) the unit increased the pressure above 14.4. Then you could increase tour fixed pressure or stay on auto.
Sleepyhead graphs will show you this and help determine your "parameters".
I really like the idea of auto CPAP all the time especially during periods when I really don't need a higher pressure to keep my airway open. Just not sure I can stay on auto pap all the time.

Anyway you suggested a pressure setting of 12.0 minimum to 20 maximum which sounds exceptiable, just don't want to get blown away and get woke up...

So, if it turns out after a few days in auto pap that my maximum pressure is say 16.8, should that be where I set the CPAP when I go back to fixed?????
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Too tall
Posts: 297
Joined: Sat Dec 14, 2013 10:58 pm

Re: Length of Time of Obstructive Apena

Post by Too tall » Mon Dec 07, 2015 3:23 pm

I had this exact thing thing. Went to the sleep doctor over it and saw the NP. She couldn't explain it. Maybe rem? No answer. One thing I found is that I was taping my mouth shut with 3M durapore when this happened. With a chin strap, I didn't see these little outburst of OSA? I use a nasal pillow. Mine were going for 30 minutes sometimes.
System One RemStar Pro with C-Flex+ (460P)

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Length of Time of Obstructive Apena

Post by Wulfman... » Mon Dec 07, 2015 3:36 pm

You need to keep in mind that without preceding events (flow limitations and/or snores) an APAP will NOT try to interrupt/stop an apnea event in progress. You have very low Flow Limitation and Snore indexes. Consequently, I doubt that using a range would have the desired result you seek.
I'd suggest trying a pressure increase or figuring out if you might have been on your back during the time of that cluster.
And, from looking at that report, I'd speculate that the cluster was "positional".


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
jnk...
Posts: 2988
Joined: Fri Sep 19, 2014 12:36 pm
Location: New York State

Re: Length of Time of Obstructive Apena

Post by jnk... » Mon Dec 07, 2015 3:46 pm

Although, it may be that the flattening of the flow curve might have been enough to cause an algorithm response, in the case of the grab supplied.
-Jeff (AS10/P30i)

Accounts to put on the foe list: Me. I often post misleading, timewasting stuff.

User avatar
ChicagoGranny
Posts: 15360
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Length of Time of Obstructive Apena

Post by ChicagoGranny » Mon Dec 07, 2015 4:00 pm

fdw wrote:So, if it turns out after a few days in auto pap that my maximum pressure is say 16.8, should that be where I set the CPAP when I go back to fixed?????
No.
fdw wrote:I really like the idea of auto CPAP all the time especially during periods when I really don't need a higher pressure to keep my airway open.
Right. That is why autopaps are favored by so many of us.
fdw wrote:Just not sure I can stay on auto pap all the time.
Why? Many of us have used it for years.
fdw wrote:Anyway you suggested a pressure setting of 12.0 minimum to 20 maximum which sounds exceptiable, just don't want to get blown away and get woke up...
If you don't need higher pressure, the machine will not raise the pressure.

You are worried about getting awoken by higher pressure? Waking up is better than continuing to sleep and suffering long apneas. But, higher pressure is not necessarily going to wake you up. My pressure is all over the place, and I rarely awaken.
LSAT: I think I would switch over from fixed to auto with a pressure range of 12-20 to see if the pressure you are at is sufficient.
Then you see how you feel the next morning and make a judgment of how you slept. Follow this up with a review of Sleepyhead details for the night to look at events, pressure, pressure changes and mask leak. You need to see this before deciding what to do on the second night. It's a type of multi-night titration.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Length of Time of Obstructive Apena

Post by fdw » Mon Dec 07, 2015 5:19 pm

ChicagoGranny wrote:
fdw wrote:So, if it turns out after a few days in auto pap that my maximum pressure is say 16.8, should that be where I set the CPAP when I go back to fixed?????
No.
fdw wrote:I really like the idea of auto CPAP all the time especially during periods when I really don't need a higher pressure to keep my airway open.
Right. That is why autopaps are favored by so many of us.
fdw wrote:Just not sure I can stay on auto pap all the time.
Why? Many of us have used it for years.
fdw wrote:Anyway you suggested a pressure setting of 12.0 minimum to 20 maximum which sounds exceptiable, just don't want to get blown away and get woke up...
If you don't need higher pressure, the machine will not raise the pressure.

The WHY is because I have a pacemaker/defibulator and once upon a time diagnosed with CHF. I read that people diagnosed with CHF should NOT use APAP only CPAP.

I would like to use APAP

You are worried about getting awoken by higher pressure? Waking up is better than continuing to sleep and suffering long apneas. But, higher pressure is not necessarily going to wake you up. My pressure is all over the place, and I rarely awaken.
LSAT: I think I would switch over from fixed to auto with a pressure range of 12-20 to see if the pressure you are at is sufficient.
Then you see how you feel the next morning and make a judgment of how you slept. Follow this up with a review of Sleepyhead details for the night to look at events, pressure, pressure changes and mask leak. You need to see this before deciding what to do on the second night. It's a type of multi-night titration.

The WHY is because I have a pacemaker/defibulator and once upon a time diagnosed with CHF. I read that people diagnosed with CHF should NOT use APAP only CPAP.

I would like to use APAP......By the way I'm also taking 5mg of generic Ambien
Last edited by fdw on Mon Dec 07, 2015 5:32 pm, edited 1 time in total.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
ChicagoGranny
Posts: 15360
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Length of Time of Obstructive Apena

Post by ChicagoGranny » Mon Dec 07, 2015 5:30 pm

fdw wrote:The WHY is because I have a pacemaker/defibulator and once upon a time diagnosed with CHF. I read that people diagnosed with CHF should NOT use APAP only CPAP.
I am glad I asked.

If I had those problems, I would be working with a sleep doctor who corresponds with my heart doctor. I would not be getting advice from an internet forum on machine settings.

In a case like yours, forums can offer good help for things like controlling leak, comfort issues and understanding your Sleepyhead reports. I would not trust a forum for things that might affect the heart.

Good luck.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
fdw
Posts: 260
Joined: Wed Aug 27, 2014 7:08 am
Location: D/FW TEXAS

Re: Length of Time of Obstructive Apena

Post by fdw » Mon Dec 07, 2015 5:34 pm

ChicagoGranny wrote:
fdw wrote:The WHY is because I have a pacemaker/defibulator and once upon a time diagnosed with CHF. I read that people diagnosed with CHF should NOT use APAP only CPAP.
I am glad I asked.

If I had those problems, I would be working with a sleep doctor who corresponds with my heart doctor. I would not be getting advice from an internet forum on machine settings.

In a case like yours, forums can offer good help for things like controlling leak, comfort issues and understanding your Sleepyhead reports. I would not trust a forum for things that might affect the heart.

Good luck.

Actually I've been on my own since my insurance does not cover sleep related issues.
ResMed AirSense S10 Autoset (Ins. Pur. 3/16/2016)
AutoCPAP 10-15 EPR 1.0 (Latest Sleep Study 3-7-16)
Started CPAP March 1995 with a pressure of 11.0

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: Length of Time of Obstructive Apena

Post by Jay Aitchsee » Mon Dec 07, 2015 6:41 pm

FDW, can you supply a reference for this statement?
fdw wrote:The WHY is because I have a pacemaker/defibulator and once upon a time diagnosed with CHF. I read that people diagnosed with CHF should NOT use APAP only CPAP.
I am familiar with the warning regarding ASV and CHF but not with APAP and CHF. In fact ResMed says,
The study and this warning do not apply to ResMed CPAP or APAP devices.
referencing their warning regarding ASV and CHF. http://www.resmed.com/us/dam/documents/ ... AQs_v1.pdf