ResMed Pressure Adjustment

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Don Up North

ResMed Pressure Adjustment

Post by Don Up North » Sat Oct 24, 2015 2:41 pm

I would like to know how to adjust the pressures on my ResMed BiPap (VPAP). My respiratory therapist will not tell me how.

DWaldman
Posts: 41
Joined: Sat Oct 24, 2015 2:35 pm

Re: ResMed Pressure Adjustment

Post by DWaldman » Sat Oct 24, 2015 3:01 pm

Do a search for the clinician's manual for the machine you have, example, (resmed s9 clinician's manual) and you will find it. Google is your friend.

User avatar
LSAT
Posts: 13364
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: ResMed Pressure Adjustment

Post by LSAT » Sat Oct 24, 2015 3:02 pm

What Resmed Bipap do you have...S8...S9...Airsense 10...??????????

big_dave
Posts: 202
Joined: Thu May 17, 2012 11:17 am
Location: Eugene, OR

Re: ResMed Pressure Adjustment

Post by big_dave » Sun Oct 25, 2015 12:17 pm

Don Up North wrote:I would like to know how to adjust the pressures on my ResMed BiPap (VPAP). My respiratory therapist will not tell me how.
Depending on which machine model you have, one of these should get you into the clinical menu:

* Hold down the knob and the home button for 3 seconds. (AirSense 10 series)
* Hold down the knob and the button with 2 check marks for 3 seconds. (S9 series)
* Hold down the down arrow and right arrow buttons for 3 seconds. (S8 or S7 series)

For more detailed instructions, you can go to:
http://www.apneaboard.com/adjust-cpap-p ... tup-manual

_________________
Machine: PR System One REMStar 60 Series Auto CPAP Machine
Mask: Pilairo Q Nasal Pillow CPAP Mask with Headgear
Additional Comments: PR DS-560 System One; former S8 Autoset II user; Sleepyhead (Mac); I sometimes use the Mirage Activa