Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 9:07 am
by Madalot
Desktop computer hard drive is obviously NOT fixed as DirectView won't work again. Back to the netbook.
Waveforms from last night, sample with pulsations and sample without. Only change to settings was Auto Trak rather than Flow Trigger. It's obvious which set is which:
Daily Detail looked fine to me. Nothing out of whack.
Oh -- Pulmonette asked me to check the Error Log to see if it would give us any hints. Cannot find it. Help on that please?
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 10:19 am
by Pulmonette
1. Insert an SD Card into the SD Card slot on the Trilogy Ventilator.
2. Press down the Audio Pause and Down Button at the same time.
3. Scroll down in the menu until you come to ‘Write Event Log to SD Card’ Select OK.
4. Remove the SD Card from the Trilogy Ventilator.
5. Connect an SD Card Reader to your PC.
6. Place the SD Card into the SD Card reader.
7. Double click on the Trilogy Tool Box Icon.
8. Click on the BrowseUUTLogs Menu.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 11:05 am
by Madalot
Pulmonette wrote:1. Insert an SD Card into the SD Card slot on the Trilogy Ventilator.
2. Press down the Audio Pause and Down Button at the same time.
3. Scroll down in the menu until you come to ‘Write Event Log to SD Card’ Select OK.
4. Remove the SD Card from the Trilogy Ventilator.
5. Connect an SD Card Reader to your PC.
6. Place the SD Card into the SD Card reader.
7. Double click on the Trilogy Tool Box Icon.
8. Click on the BrowseUUTLogs Menu.
Okay -- all good until #7. All I get is a Trilogy FOLDER (both on XP & Windows 7). I don't see a Tool Box icon or UUTLOGs anywhere. Just a lot of files.
There are BIN, EDF, & Microsoft Excel files in the folder.
2C_TV111050272_17_1373996834.bin (example of one bin file)
AD_201309_000.edf (example of one edf file)
EL_TV111050272_20130716 (example of one excel file)
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 11:41 am
by Pulmonette
Oh, right.
"Don't ask me where I got it."
Got a DropBox?
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 11:51 am
by Madalot
Pulmonette wrote:Oh, right.
"Don't ask me where I got it."
Got a DropBox?
Say, What? You lost me.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:01 pm
by Madalot
Is there a separate "Error" Log on this puppy? The "Event" Log is onscreen, plus that's what is in the Excel files on the SD card (I looked). Are you talking about something different? Because I'll be damned if I can see or find it.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:12 pm
by robysue
Madalot wrote:Pulmonette wrote:Oh, right.
"Don't ask me where I got it."
Got a DropBox?
Say, What? You lost me.
DropBox.com provides space for users to share folders and files with other users. A small (approximately 2 Gig) DropBox account is free. It also allows you to sync folders and files across multiple computers extremely easily.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:16 pm
by Madalot
robysue wrote:Madalot wrote:Pulmonette wrote:Oh, right.
"Don't ask me where I got it."
Got a DropBox?
Say, What? You lost me.
DropBox.com provides space for users to share folders and files with other users. A small (approximately 2 Gig) DropBox account is free. It also allows you to sync folders and files across multiple computers extremely easily.
Thanks, robysue!! I do NOT, but I'm game if you want, Pulmonette. I'll set up an account and see how it works. Give me a few minutes...
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:27 pm
by Madalot
Pulmonette wrote:Oh, right.
"Don't ask me where I got it."
Got a DropBox?
Thanks to robysue, yes I do. And the Trilogy folder is in it. How do YOU access it?
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:31 pm
by Pulmonette
Post the link.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:34 pm
by Madalot
https://www.dropbox.com/sh/pnm43shsca1ggy8/Rxj81YMaji
Okay -- VERY new to this. Try this. Please let me know if it works.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:39 pm
by Pulmonette
Fine.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 12:42 pm
by Madalot
Pulmonette wrote:Fine.
Cool. I'm going to go dry my hair, then sit down for a while. If I get sleepy enough, I will mask up for a while and see if those pesky pulses occur while sleeping in a recliner...
Got to remember to put the card back in the Trilogy...still in the card reader at the moment.
Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 2:25 pm
by Madalot
Got in my recliner and masked up. Here's the waveforms from this. As soon as I started to doze off, I felt the pulses. I never really fell asleep. Close, but not completely.

Re: Trilogy AVAPS-AE - Ongoing
Posted: Mon Sep 02, 2013 2:36 pm
by -SWS
I'm definitely favoring the faulty-valve hypothesis. Madalot, when do you get your Trilogy back?