AVAPS-AE on MY Trilogy - Ongoing

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Madalot
Posts: 4285
Joined: Sat Jan 30, 2010 7:47 am

Re: Trilogy AVAPS-AE - Ongoing UPDATE

Post by Madalot » Thu Oct 03, 2013 5:16 am

jamiswolf wrote:Hi Maddie,
Going back two posts to your question regarding EPAP pressures. I'm a little rusty thinking about all this stuff. You were surprised that your machine was only showing 10cm, but your study recommended 12cm. Whats your IPAP running...just to give me more info.

Next post you commented that you sleep better with avaps, yet you tend to feel more well rested on the ae setting. How are you determining that you sleep better with plain avaps? By machine data?
My new doctor didn't like the unevenness of the 02 line on my pulse ox report and wanted my IPAP increased from 15 to 18. I was able to do that slowly over the course of about two weeks. My IPAP is now 18-23. My EPAP is at 9 (regular AVAPS) which is where it has been for a while.

My comment about sleeping better on regular AVAPS is based on my perceptions. I don't feel I wake up as often and when I do wake, the machine doesn't seem part of the equation. On AE, the FOTs were bothersome, at times, and when awakened, they became part of the equation.

I was hoping that AE would show that it raised my EPAP significantly when sleeping supine, but that wasn't the case. The average EPAP was typically 9.5, which is barely higher than my set EPAP. The last sleep study was pretty clear that an EPAP of 12 was needed to sleep supine. But then again, they weren't using a Trilogy, but the other equipment (can't remember it's name off hand) that can be set to both AVAPS & ASV modes. I suppose there could be a slight discrepancy in how the equipment functions.

Unless there are comments to the contrary, I think my smartest move is to wait until my Trilogy returns then try AE again on that machine. I'll go back through this thread and figure out the starting point. One change at a time.

Until then, regular AVAPS and snoozing in the afternoons. It's still better than before when I slept for HOURS every afternoon. Little cat naps, in my recliner, aren't the end of the world.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
Madalot
Posts: 4285
Joined: Sat Jan 30, 2010 7:47 am

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by Madalot » Thu Oct 24, 2013 9:50 am

I didn't want people to think I'd dropped the ball on this, so here's the latest.

I got my Trilogy back. I went through the settings & options to see what's new, different, etc. I didn't see much that was obviously changed. A new option under a couple of the modes - MPV - which a little research stands for "Mouth Piece Ventilation" which is something I don't need at this stage. Beyond that, I don't *see* much change from the loaner I had.

I plan to give AVAPS-AE another try on THIS machine. My doctor has okay'd my redoing the test, just to see *IF* the updated Trilogy does better or behaves differently than the loaner machine.

If the results are about the same, we feel that the best option would be to increase my set EPAP to 10 on regular AVAPS. I currently have it set to 9, but the AE testing was showing an average EPAP around 9.5. If I can get myself to tolerate EPAP of 10, I *might* be on target for better settings, better sleep and better stamina the next day.

My doctor wrote an official prescription to increase my EPAP to 10 (we need to keep the official records accurate).

I'll start the AVAPS-AE testing again soon and see where it goes. I got my flu shot last week and had flu-like symptoms for several days afterwards. Now, I seem to have pulled/aggravated muscles in my upper back on my left side, causing horrible spasms and pain in my back and chest. I have been MISERABLE and not up to doing anything other than just trying to find a comfortable position to sleep.

Once this has passed and feel more like myself, I'll be trying AVAPS-AE again.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

giblets

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by giblets » Fri Apr 17, 2015 8:48 am

Not sure if the user is still about, but read this OLD thread. The oscilations are indeed expected (almost always 10 in a row). Its the way the trilogy detects if there is an obstruction or limitation or not. It always tests to see if it can improve things by increasing the EPAP, test again, and if no improvement then drop the EPAP, test again, then drop the epap, test again, if it's worse...increase the EPAP...test again... every minute or so.

User avatar
Madalot
Posts: 4285
Joined: Sat Jan 30, 2010 7:47 am

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by Madalot » Fri Apr 17, 2015 9:43 am

giblets wrote:Not sure if the user is still about, but read this OLD thread. The oscilations are indeed expected (almost always 10 in a row). Its the way the trilogy detects if there is an obstruction or limitation or not. It always tests to see if it can improve things by increasing the EPAP, test again, and if no improvement then drop the EPAP, test again, then drop the epap, test again, if it's worse...increase the EPAP...test again... every minute or so.
Oh yeah, I'm still here. I gave up on AVAPS-AE a while ago, using straight AVAPS with the EPAP at a set 10. If pain forces me on my back again, I may try AVAPS-AE again, but for now, I'm okay with the way it is. I wake up every 1.5 - 2 hours, probably because of pain, but can usually move around and go back to sleep.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by raisedfist » Fri Jun 24, 2016 7:12 am

Hi Madalot,

Are you still around? Hope you are doing well.

I am having problems with AVAPS-AE; I am still having oxygen desaturations - presumably during REM sleep. I keep waking up once the EPAP ramps up - I can barely breathe when it's at 9 or 10. I also wonder about the auto-breath rate. In watching the screen it does seem to select a nice, long inspiration time so maybe it's helpful but I'm not sure.

I am wondering - should I raise the PS minimum and shorten the EPAP range? All my sleep studies show I suffer from hypoventilation and thus all recorded sleep events are obstructive hypopneas - no discrete apneas.

In straight bi-level, you are supposed to raise the IPAP for hypopneas and EPAP for apneas. I used IPAP 20/EPAP 4 before switching over to AVAPS with the Trilogy. With AVAPS, I assume I am having desaturations because the IPAP is low and the EPAP has to kick in extra hard in order to prevent/fix the hypopneas. I am meeting my target tidal volume each night.

My settings are:

AVAPS-AE
Min PS: 6
Max PS: 16
EPAP: 4-14
Breath Rate: Auto


I am going to try:

Min PS: 12
Max PS: 20
EPAP: 4-10
Breath Rate: 12
Inspiration time: 2.0 seconds

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Madalot
Posts: 4285
Joined: Sat Jan 30, 2010 7:47 am

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by Madalot » Sun Jun 26, 2016 5:12 am

raisedfist wrote:Hi Madalot,

Are you still around? Hope you are doing well.

I am having problems with AVAPS-AE; I am still having oxygen desaturations - presumably during REM sleep. I keep waking up once the EPAP ramps up - I can barely breathe when it's at 9 or 10. I also wonder about the auto-breath rate. In watching the screen it does seem to select a nice, long inspiration time so maybe it's helpful but I'm not sure.

I am wondering - should I raise the PS minimum and shorten the EPAP range? All my sleep studies show I suffer from hypoventilation and thus all recorded sleep events are obstructive hypopneas - no discrete apneas.

In straight bi-level, you are supposed to raise the IPAP for hypopneas and EPAP for apneas. I used IPAP 20/EPAP 4 before switching over to AVAPS with the Trilogy. With AVAPS, I assume I am having desaturations because the IPAP is low and the EPAP has to kick in extra hard in order to prevent/fix the hypopneas. I am meeting my target tidal volume each night.

My settings are:

AVAPS-AE
Min PS: 6
Max PS: 16
EPAP: 4-14
Breath Rate: Auto


I am going to try:

Min PS: 12
Max PS: 20
EPAP: 4-10
Breath Rate: 12
Inspiration time: 2.0 seconds
Not around much, but saw this and got your PM. Sent you a PM. I stopped AVAPS-AE a long time ago because of the FOT's. Early in my treatment when I was having desaturations, my doctor raised my IPAP. That's about all I can tell you.

_________________
Mask: FlexiFit HC431 Full Face CPAP Mask with Headgear
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Trilogy 100. S/T AVAPS, IPAP 18-23, EPAP 10, BPM 7

roman.kostjuk
Posts: 2
Joined: Wed Mar 14, 2012 4:31 pm

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by roman.kostjuk » Thu Jun 27, 2019 4:34 pm

Hi everyone. I see that's old topic. But maybe someone can help me.
I'm highly in need to reed Logs on SD Card from Triilogy machine with BrowseUUTLogs app.
Please help

Nikilet2
Posts: 3
Joined: Tue Jun 04, 2019 8:38 pm

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by Nikilet2 » Fri May 20, 2022 5:18 pm

I want to post re my Trilogy 100. I have the option to POST REPLY but can't find an option to post. Can someone help?

User avatar
Pugsy
Posts: 64024
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by Pugsy » Fri May 20, 2022 6:22 pm

Nikilet2 wrote:
Fri May 20, 2022 5:18 pm
I want to post re my Trilogy 100. I have the option to POST REPLY but can't find an option to post. Can someone help?
newtopic.PNG

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
chunkyfrog
Posts: 34394
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by chunkyfrog » Fri May 20, 2022 6:34 pm

Pugsy, has anyone heard from Madalot lately?
I think of her often.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AVAPS-AE on MY Trilogy - Ongoing

Post by palerider » Fri May 20, 2022 8:48 pm

Pugsy wrote:
Fri May 20, 2022 6:22 pm
Nikilet2 wrote:
Fri May 20, 2022 5:18 pm
I want to post re my Trilogy 100. I have the option to POST REPLY but can't find an option to post. Can someone help?
newtopic.PNG
Probably should have added some arrows, just in case they still can't find it:
2022-05-20_21-47.png

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.