Not getting enough air
Not getting enough air
New CPAP, REMSTAR Plus C-Flex set at 9cm. I feel like I'm not getting enough air, just like being congested and trying to breath through your nose. No problem falling asleep once I relax, but I can do that anytime, anywhere. Can I increase the pressure myself?
I was diagnosed about 8 yrs ago, used the machine for a few months. I stopped using the machine because of discomfort with headgear and I didn't feel any more rested. It wasn't worth the hassle. I began using it again about 14 months ago at my husband request. Recently had new sleep study done and Dr prescribed 6cm(thought I would die). I had to check with my original DME to get the pressure prescribed the first time(9cm.) as my Dr didn't have any record of my sleep study (curious). DME reset pressure at 9cm. I did notice a difference this time. Still don't feel more rested but no more headaches upon waking. Using Swift FX Bella nasal pillows, LOVE it!!
Thanks for any help.
I was diagnosed about 8 yrs ago, used the machine for a few months. I stopped using the machine because of discomfort with headgear and I didn't feel any more rested. It wasn't worth the hassle. I began using it again about 14 months ago at my husband request. Recently had new sleep study done and Dr prescribed 6cm(thought I would die). I had to check with my original DME to get the pressure prescribed the first time(9cm.) as my Dr didn't have any record of my sleep study (curious). DME reset pressure at 9cm. I did notice a difference this time. Still don't feel more rested but no more headaches upon waking. Using Swift FX Bella nasal pillows, LOVE it!!
Thanks for any help.
Re: Not getting enough air
This machine?
https://www.cpap.com/cpap-machine/respi ... cflex.html
PR System Plus CPAP... no data available...just wanted to let you know. That SD card gathers hours of use only. You have zero way to evaluate the data to see if therapy is optimal which might explain the not so feeling great. No way to know if leaks are a problem. No way to know if the 9 cm pressure is optimal and perhaps you just need more time or maybe simply more hours of sleep. Sort of flying blind without data.
Regarding not enough air...are you using ramp feature starting out low? If so try not using ramp.
If you aren't using ramp...what CFlex setting are you using if you are using it. Might try changing the CFlex setting and see if that help with how the breathing feels.
Pressure of 9 cm should be more than adequate to prevent any stifling sensations from not enough pressure. Now a setting of 4 or 5 or 6 cm..yeah can be a bit stifling for some people especially someone who was used to a higher pressure from previous use.
https://www.cpap.com/cpap-machine/respi ... cflex.html
PR System Plus CPAP... no data available...just wanted to let you know. That SD card gathers hours of use only. You have zero way to evaluate the data to see if therapy is optimal which might explain the not so feeling great. No way to know if leaks are a problem. No way to know if the 9 cm pressure is optimal and perhaps you just need more time or maybe simply more hours of sleep. Sort of flying blind without data.
Regarding not enough air...are you using ramp feature starting out low? If so try not using ramp.
If you aren't using ramp...what CFlex setting are you using if you are using it. Might try changing the CFlex setting and see if that help with how the breathing feels.
Pressure of 9 cm should be more than adequate to prevent any stifling sensations from not enough pressure. Now a setting of 4 or 5 or 6 cm..yeah can be a bit stifling for some people especially someone who was used to a higher pressure from previous use.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- chunkyfrog
- Posts: 34397
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Not getting enough air
Cpap without data is like treating diabetes without pricking your finger.
If it works at all, it's just a coincidence.
At 8 years, your insurance (if you have any) should cover a new machine.
This time, please read the info about full data machines, maybe even automatic machines, before getting anything.
The right equipment makes all the difference.
If it works at all, it's just a coincidence.
At 8 years, your insurance (if you have any) should cover a new machine.
This time, please read the info about full data machines, maybe even automatic machines, before getting anything.
The right equipment makes all the difference.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Not getting enough air
Link for Jan's blog about full data machines and "bricks".
http://maskarrayed.wordpress.com/
I couldn't tell for sure which machine OP was referring to..."new" was mentioned so I assumed the PR S1.
Even if it is the older Remstar Plus...still no data.
http://maskarrayed.wordpress.com/
I couldn't tell for sure which machine OP was referring to..."new" was mentioned so I assumed the PR S1.
Even if it is the older Remstar Plus...still no data.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Not getting enough air
i wake up in the middle of the night feeling that im not getting enough air, i shut of unit for 20-30 min then i can use it. Unit is a Resmed Airsense 10
Re: Not getting enough air
This thread is over 10 years old.
Please start a new topic/thread of your own to get more specific ideas about your problem.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Not getting enough air
Beware that my blog hasn't been updated and the info on machines is outdated. If you are getting a new machine, we are now recommending that the ONLY models you should accept are a ResMed 10 or 11 Autoset or Autoset for Her. Those are basic machines that have maximum flexibility for basic treatment and are fully data capable so your doctor and we can help you. Do not accept anything else. While they are auto-adjusting machines, they can be set with a fixed pressure if that works best for you.Pugsy wrote: ↑Tue Mar 27, 2012 11:31 amLink for Jan's blog about full data machines and "bricks".
http://maskarrayed.wordpress.com/
I couldn't tell for sure which machine OP was referring to..."new" was mentioned so I assumed the PR S1.
Even if it is the older Remstar Plus...still no data.
The machine you have is what we refer to as a brick. It only works at a fixed pressure which may or may not be best for you, and there's NO useful data to help adjust and optimize your therapy.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Not getting enough air
Check dates of posts folks....I made that post back in 2012.
Zombie thread alert.
Janknitz....you replied to old poster ...not the new account that today resurrected this old post and who I told to start their own thread.
Zombie thread alert.
Janknitz....you replied to old poster ...not the new account that today resurrected this old post and who I told to start their own thread.
Janknitz wrote: ↑Mon Jun 05, 2023 1:37 pmBeware that my blog hasn't been updated and the info on machines is outdated. If you are getting a new machine, we are now recommending that the ONLY models you should accept are a ResMed 10 or 11 Autoset or Autoset for Her. Those are basic machines that have maximum flexibility for basic treatment and are fully data capable so your doctor and we can help you. Do not accept anything else. While they are auto-adjusting machines, they can be set with a fixed pressure if that works best for you.Pugsy wrote: ↑Tue Mar 27, 2012 11:31 amLink for Jan's blog about full data machines and "bricks".
http://maskarrayed.wordpress.com/
I couldn't tell for sure which machine OP was referring to..."new" was mentioned so I assumed the PR S1.
Even if it is the older Remstar Plus...still no data.
The machine you have is what we refer to as a brick. It only works at a fixed pressure which may or may not be best for you, and there's NO useful data to help adjust and optimize your therapy.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Not getting enough air
Yes, I saw that, but wanted to put that out there because any newbies looking for my blog might not get that the info is outdated.Janknitz....you replied to old poster ...not the new account that today resurrected this old post and who I told to start their own thread.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm