Posted: Sat Feb 25, 2006 11:45 pm
by rested gal
Jerry69 wrote:Did you mean 0.4l/s? That would be the leak rate at 11 cm, from interpolation of the chart.
Dang, I never get those pesky decimal points right! I get days of the week wrong too!

Yeah, what you said is what I meant, Jer. Thanks for the catch!! Bless your heart!

< scribbling note to self.... ".4 .4 .4 .4" >

Posted: Sun Feb 26, 2006 4:19 pm
by Jakester
Ok, so here's some leak data on the Breeze venting from EncorePro for this week. If I am reading this right the 90% leak is a little high and the avg leak is a little low.

My AHI was 2.1

Tube is 16" long

90% leak-26.2
Avg leak-17.5

90% Press-10.3
Avg Press-9.5


Leak

Posted: Sun Feb 26, 2006 5:34 pm
by Jerry69
Jakester wrote:Ok, so here's some leak data on the Breeze venting from EncorePro for this week. If I am reading this right the 90% leak is a little high and the avg leak is a little low.

My AHI was 2.1

Tube is 16" long

90% leak-26.2
Avg leak-17.5

90% Press-10.3
Avg Press-9.5
As you say, 17.5 may be slightly low. Mine is usually about 20 l/m at 9 cm pressure, but that includes slight mask leakage at the pillows.

Seems close enough to me. So, you are using 1/4" O.D. thin wall plastic tubing and forcing it into the vent hole.

I'll head down to ACE Hardware tomorrow and get a piece of tubing and give it a try tomorrow night.

Thanks,

Jerry