idamtnboy wrote:Welllllll, if you wouldn't buy from Harbor Freight!!
I learned something a long time ago from other people's bad experience. Nothing from Harbor Freight stays powered on when unattended in my house, so there are no battery charger circuits from Harbor Freight anywhere near me.
idamtnboy wrote:Come on man, get real. How likely is the chance of AC current jumping through the circuitry of a charger to one of the output leads? And if it does happen you've got the circuitry in the converter to stop the flow. It'll probably burn up the interior of the converter. And then you've got all the circuitry of the PAP machine, which would burn up stopping a 120v current. And of course this all presupposes a return path to ground through all of the paraphernalia above, a very unlikely event. Then, if after getting through all that there is somehow 120v imposed on the hose there still almost certainly will be no path to ground. Bed clothes are insulators, and unless you have your hand or feet touching some sort of metal that connects to ground you won't conduct any electricity at all. And whatever kind of floor the bed is on, the floor most likely won't provide a good enough ground to cause a shock. And even if you do conduct 120v it will only be for a very brief moment because all the circuit components that are being breached will burn up, creating an open path. A human can tolerate a 120v shock for a few seconds. I know. Twice as a teenager I was part of a very clear 120v path to ground, once from one finger to my bare feet on dirt, and another from left hand to my right hand, right across my heart.
You suffer from some very common misconceptions about electrical safety. Lots of people have died from those misconceptions, even from non-medical equipment. I'll try to address a few of them.
Lots of people have died in the from being electrocuted by medical devices. I remember back in the 1970's (?) when there was a big push to improve medical device safety and much more stringent electrical safety requirements were instituted. Quite a few electrical devices come in special "hospital" or medical grade versions. Quite a few devices are labeled as not for medical use.
Shorting the 120VAC to the low voltage secondary is not at all unlikely. There's usually a transformer somewhere in the device with 120VAC wound on a coil, a plastic insulator over the inner coil, and the low voltage "safe" side of the power supply wound over the plastic insulator. If the insulation between the 120VAC and the "safe" side melts or gets damaged, the 120 VAC can be connected to the low voltage side.
Reputable manufacturers go to quite a bit of effort to be sure this is unlikely to happen. Unfortunately there's a lot of cheap junk from China (or the USA) being sold these days that's not really up to electrical safety standards.
Tear apart one of the little wall wart DC plugs that everything seems to use these days. Look at where the AC and DC voltages go. Or look at the DC supply in a device that plugs directly into the wall. Imagine what happens if it gets hot and things start to melt. Then be afraid.
Even if the 120VAC gets through the transformer to the electrically isolated low voltage side of the electrical device, it will not necessarily burn anything out. Since it's electrically isolated, the "ground" side of the DC power supply can be at 120VAC and no extra current will flow because there's no path to ground. The device will run just fine until something provides a path to ground. Some medical devices are extra dangerous because your body is likely to be the path to ground.
You're assuming that if 120VAC gets "loose," it will "burn up" something and there will no longer be any current flowing. Even if the 120VAC damages the electrical circuitry in your CPAP, it won't necessarily vaporize it causing an open circuit. "Burned up" non-functional circuitry is often still electrically conductive.
A current that will easily kill a human being may not damage wiring and circuit boards at all. If you become the path to ground for 120 VAC, currents of around .1 amp can easily kill you. Many CPAP machines are fused at around 7 amps, so the lethal current that's killing you won't necessarily burn anything up. It's possible all the current is flowing through the "ground" side of the DC power system and doesn't pass through any electronic devices at all.
Fuses mostly protect from fire, not from electrocution. If you open up most electrical devices and touch something that's got 120VAC on it, it's very easy to get a lethal level of current and not affect the electrical device at all.
As far as humans tolerating 120VAC shocks, that's highly variable. If your skin is dry, you don't have a good ground, etc., you may touch 120VAC and not have any harm. With a good contact, you can get a very quick severe injury. I've already explained why you can't count of the device to rapidly vaporize, act as a fuse and shut off the current.
You are right that bedclothes, beds, etc. are electrical insulators. However, there are quite a few cases where you might get grounded. For instance, turning on or touching the lamp on your headboard, grabbing the telephone, touching another grounded device on the bedside table, standing barefoot by the bed masking up, etc. If you have a 120 VAC "leak" in your equipment, it may well continue to operate for years until that one day you happen to be the path to ground.
Yes, you'll probably be safe. Is "probably" good enough for you or your loved ones?