Re: Cognitive Fog
Posted: Fri Sep 09, 2011 10:16 pm
Your equipment choices from the drop down menu.
Machine... M Series Pro with C-Flex CPAP machine. Respironics sticks all their machines with their Remstar thing highly confusing.
Humidifier.. M Series heated humidifier.
Respironics mask doesn't tell us a lot. There are a lot of them..Does it cover nose only? Cover mouth and nose? or Nasal pillows (only nostrils covered).
Your leak line is fine. Not bad enough to impact therapy or event reporting. The good report is quite good except for the snores.
Don't know what to think about all those snores.
If they are real snores they could be a problem. If it is hose rubbing against something then not real. Unfortunately we can't tell. Do you use any hose management system that keeps the hose up and off the bed covers? Or does it just flop around in bed with you?
If those are real snores that could mean quite a bit of flow limitations and perhaps needs just a bit more pressure to eliminate the bulk of those snores? Do you have your sleep study results? Could these snores be worse because you are on your back? Do you have a partner who can comment on hearing the snores or not?
If it were me I would increase the pressure 0.5 to see if it impacted those snores.
Machine... M Series Pro with C-Flex CPAP machine. Respironics sticks all their machines with their Remstar thing highly confusing.
Humidifier.. M Series heated humidifier.
Respironics mask doesn't tell us a lot. There are a lot of them..Does it cover nose only? Cover mouth and nose? or Nasal pillows (only nostrils covered).
Your leak line is fine. Not bad enough to impact therapy or event reporting. The good report is quite good except for the snores.
Don't know what to think about all those snores.
If they are real snores they could be a problem. If it is hose rubbing against something then not real. Unfortunately we can't tell. Do you use any hose management system that keeps the hose up and off the bed covers? Or does it just flop around in bed with you?
If those are real snores that could mean quite a bit of flow limitations and perhaps needs just a bit more pressure to eliminate the bulk of those snores? Do you have your sleep study results? Could these snores be worse because you are on your back? Do you have a partner who can comment on hearing the snores or not?
If it were me I would increase the pressure 0.5 to see if it impacted those snores.