Bwolf

Complex Apnea

Post by Bwolf » Thu Sep 01, 2011 7:47 am

I had a follow-up sleep study done while using a CPAP and they found out that my Apnea was "complex" and the tech said I would need a different kind of machine....I explained that I have used both a regular CPAP, and an Auto-CPAP.....other than a BI-PAP what other types of machines are there? The tech said its not very common and would be more expensive but wanted the doctor to explain it all to me.....does anyone have any ideas? Thanks, Baron

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Complex Apnea

Post by JDS74 » Thu Sep 01, 2011 8:00 am

Bwolf wrote:I had a follow-up sleep study done while using a CPAP and they found out that my Apnea was "complex" and the tech said I would need a different kind of machine....I explained that I have used both a regular CPAP, and an Auto-CPAP.....other than a BI-PAP what other types of machines are there? The tech said its not very common and would be more expensive but wanted the doctor to explain it all to me.....does anyone have any ideas? Thanks, Baron
Baron:
The machine type being referred to is an Auto Servo Ventilator machine (ASV). Both ResMed and Respironics make them and there may be others.

The machine basically operates as an auto adjusting BiPap machine until you have a detected central apnea. Then it kicks in as a ventilator type machine and forces a breath. It has another mode in which it monitors your breathing rate and kicks in if your rate falls below the minimum the machine is set for.

If you would click on the User Control Panel and select your equipment and then select "TEXT" for the display type, it would be helpful for others wanting to make helpful comments. At the same time, it would also be helpful if you would open your original post and edit to add the comment "ASV" in the subject line. That way others with more knowledge than I on ASV type therapy could chime in.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Pugsy
Posts: 65062
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Complex Apnea

Post by Pugsy » Thu Sep 01, 2011 8:05 am


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Bwolf

Re: Complex Apnea

Post by Bwolf » Thu Sep 01, 2011 8:09 am

Thanks to both of you....I'll make the changes to my post as you suggest.

Bwolf
Posts: 3
Joined: Thu Sep 01, 2011 8:11 am

Re: Complex Apnea ASV

Post by Bwolf » Thu Sep 01, 2011 8:13 am

Bwolf wrote:I had a follow-up sleep study done while using a CPAP and they found out that my Apnea was "complex" and the tech said I would need a different kind of machine....I explained that I have used both a regular CPAP, and an Auto-CPAP.....other than a BI-PAP what other types of machines are there? The tech said its not very common and would be more expensive but wanted the doctor to explain it all to me.....does anyone have any ideas? Thanks, Baron

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Complex Apnea

Post by JohnBFisher » Thu Sep 01, 2011 12:22 pm

JDS74 wrote:... That way others with more knowledge than I on ASV type therapy could chime in. ...
You did a good job with the explanation.

Bwolf, we can add a lot more information as you go along with it. And it will take some advice on how to live with it. If people think the respiratory therapists are clueless about living with xPAP, they have *no* clue on living with an ASV machine. What do I mean by that? Well, when you use a CPAP or BiLevel machine, the pressure is fairly steady and leaks in the mask are nothing - in comparison to an ASV machine. You see, when the ASV machine kicks in as a ventilator, it ramps the pressure up toward 25 to 30cm H2O (depends on your specific prescription). But at that level there is not a mask in the world that stays on without leaking. Getting a mask to work at those pressures is "interesting". But by using mask liners and other tricks, you'll learn how to avoid those problems.

And the sleep tech was right. The machines cost a LOT. Hang on to your hat (be seated) when you hear the price. But I would pay almost ANYTHING for that good night of sleep. And my ASV machine is *WONDERFUL*. I sleep like a baby with it. I don't / can not / sleep without it.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
Paper_Nanny
Posts: 430
Joined: Fri Apr 15, 2011 3:52 pm
Location: Southern Oregon

Re: Complex Apnea

Post by Paper_Nanny » Thu Sep 01, 2011 12:35 pm

JohnBFisher wrote:And the sleep tech was right. The machines cost a LOT.
My insurance company didn't have a problem with covering my ASV, so although the cost is a lot more, for me, that was a moot point.

It does take a bit to get used to the way the ASV works and this is a great place for getting ideas and suggestions on how to make the therapy work for you. For almost everyone, it really is a matter of being committed to your therapy and committed to finding ways to deal with any problems that come up.

That said, you could be one of the lucky ones who doesn't have much trouble adjusting to the ASV. I was in that category. I had a few minor problems getting used to it, but those were easily remedied. It has turned out that the ASV may not be the best machine for me to be using, but that is a separate issue.

Read a lot and ask questions if you have any problems. And best wishes!

Deborah

_________________
Mask: FitLife Total Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: DreamSTation DSX500
PR System One BiPAP Auto SV Advanced

Bwolf
Posts: 3
Joined: Thu Sep 01, 2011 8:11 am

Re: Complex Apnea

Post by Bwolf » Thu Sep 22, 2011 8:53 am

Thank you to everyone for your replies....the insurance is currently doing a pre-authorization for the machine...so we will wait and see.

I am concerned however as to why I have central apena...my doctor said that centrol apena is usually only found in people who either: abuse drugs, are extremely overweight, have other complicated medical conditions like HIV, etc.....I don't use drugs, am not overweight, and regularly test negative for other serious medical issues.....am I just an odd case?

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Complex Apnea

Post by JohnBFisher » Thu Sep 22, 2011 11:21 am

Bwolf wrote:Thank you to everyone for your replies....the insurance is currently doing a pre-authorization for the machine...so we will wait and see.

I am concerned however as to why I have central apena...my doctor said that centrol apena is usually only found in people who either: abuse drugs, are extremely overweight, have other complicated medical conditions like HIV, etc.....I don't use drugs, am not overweight, and regularly test negative for other serious medical issues.....am I just an odd case?
Bwolf, just remember that in any graduating class of doctors that one half of them do not finish in the top half! And I fear this is a perfect example. Complex Sleep Apnea Syndrome (CompSAS), is a well known and well documented situation, where the respiratory systems fails to trigger as it should with increased pressure. This is not something new. It is not something unique. And it MUST be addressed as you would address obstructive apneas. Failure to do so *IS* very detrimental to your health. This is not a "MIGHT BE" situation. This is a "DEFINITE" situation.

If, during the sleep study, you exhibited signs of complex sleep apnea, then your doctor should work to resolve the issue. If he fails to do so, go find another doctor who will do so.

Why? Well, I had an idiot doctor, who told me that I could not have central sleep apnea, because it is so rare. (Not that I did not have central apneas. I did during ALL of my sleep studies. I've had increasing problems during the wake/sleep transition - reaching a point where my BiPAP would shut off because I stopped breathing for more than a minute at a time!!) He was just convinced that if it was rare, then I could not have it. Idiot!

I should have paid a LOT more attention to my own symptoms. Since I did not, I developed uncontrolled high blood pressure (even with medication). I also struggled with HORRIBLE headaches. I used Ibuprofen to tackle the headaches. And the high blood pressure and Ibuprofen combined to permanently damage my kidneys. Gee, thanks Doc!

Just because they have a "Doctor" as part of their title does NOT mean that they are competent! If you are having problems with sleep - PAY ATTENTION !!! If the doctor will not, find another who will.

http://www.resmed.com/us/clinicians/abo ... apnea.html

Hope that helps.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Complex Apnea

Post by archangle » Thu Sep 22, 2011 12:03 pm

JohnBFisher wrote:my BiPAP would shut off because I stopped breathing for more than a minute at a time!!
It shuts off if you stop breathing for a minute?

Geez, what horrible device was that? I've heard of shutoff due to big leaks, but to shut off due to a long apnea!!!

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Complex Apnea

Post by archangle » Thu Sep 22, 2011 12:21 pm

Be sure you have a competent doctor who really looks at your sleep study PLUS waveforms before prescribing an ASV machine. Some people panic when they hear "central apnea."

My CPAP machine will occasionally show "central apneas," but when I go look at the waveform, the length of the event is short, and I'm actually still breathing somewhat during the apnea. A dumb doctor might see "x" central apneas per hour and panic.

I'm also suspicious about whether short central apneas mean anything. If your blood O2 level is high enough, maybe your breathing reflex may simply shut down for a short time until your O2/CO2 levels reach a level for you to breathe again.

I do realize that you you can have truly dangerous central apneas. Just be sure to really look at the results before you let someone inflict you with an ASV machine.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
JohnBFisher
Posts: 3821
Joined: Wed Oct 14, 2009 6:33 am

Re: Complex Apnea

Post by JohnBFisher » Thu Sep 22, 2011 12:44 pm

archangle wrote:
JohnBFisher wrote:my BiPAP would shut off because I stopped breathing for more than a minute at a time!!
It shuts off if you stop breathing for a minute?

Geez, what horrible device was that? I've heard of shutoff due to big leaks, but to shut off due to a long apnea!!!
This was a BiPAP only unit (at the time) and I had the "Auto Off" feature enabled. The assumption the engineers probably made was that if the user was not connected (no change to inspiration pressure), then the user must have taken the mask off. Thus, after one minute it would turn itself off. Now, I don't know about you, but even if a central apnea is normal during the wake/sleep transition, then one that lasts more than one minute during that transition is NOT normal. Though I have a scattering of central apneas throughout the night, the wake/sleep transition is the most difficult for me.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Additional Comments: User of xPAP therapy for over 20 yrs. Resmed & Respironics ASV units with EEP=9cm-14cm H2O; PSmin=4cm H2O; PSmax=15cm H2O; Max=25cm H2O
"I get up. I walk. I fall down. Meanwhile, I keep dancing” from Rabbi Hillel
"I wish to paint in such a manner as if I were photographing dreams." from Zdzisław Beksiński