justbreathe wrote:... I am wondering what all the fuss is about special titration and all. ...
Well, done properly they try to find the lowest EPAP setting that keeps the airway open. 10cm H2O seems to be where most of them start, but there is no real reason to have it that high. I assume they started you lower and worked up to 10 where the obstructive apneas were cleared. At least that is how it is supposed to go. Here is the protocol for titrating the older ASV unit from Resmed. I suspect it remained mostly the same for the S9 version of their ASV unit:
http://www.resmed.com/us/documents/1011 ... otocol.pdf
One reason they might have set the EPAP to 10cm H2O is to have the ability to push the maximum pressure upto 25cm H2O. The Max PS goes from 9 to 16cm H2O on the older unit. Thus, 10cm H2O (for EPAP) and 15cm H2O (for MaxPS) is 25cm H2O.
justbreathe wrote:... The EPAP is the low setting 10 inches H20 for me. It keeps the airway open and should work just like the cpap. ...
Almost. It can act as a CPAP device. In that case the PS will be 0 (zero) cm H2O. There will be no difference between IPAP and EPAP.
justbreathe wrote:... The PS, Pressure support for me is 3 inches H20. This works like a Bipap and raises to a minimum of 13 IPAP. ...
Exactly. By the way, I find it a little easier if the PS is 4 and not 3. It just separates the IPAP pressure enough to make it easier to breathe.
justbreathe wrote:... The Max PS is 15. This allows the ASV to adjust to a maximum IPAP of 25 inches of H20 when needed to clear an Apnia event. Does not matter if it is Obstructive, Hypopnia or Central in nature. It should get them all. ...
By the way, it is centimeters H2O and not inches H2O. One inch = approximately 2.54 centimeters. So, your 25" would be about 63.5cm. Yikes!
It can go UPTO (but may not) 25cm H2O. The actual pressure depends on the amount of air the unit thinks you should be breathing. If it sees that you will fall short, then it ramps up the pressure to increase that amount. It will ramp upto the 25cm H2O if needed. If it does not need that much, the amount it delivers is much lower.
justbreathe wrote:... I don't see why the EPAP would be 10 on the VPAP ADAPT when was doing fine at 6 inches H20 on straight CPAP except for the centrals. ...
As I noted, I suspect it was to give you the top end of the pressure.
justbreathe wrote:... I am not adjusting well to the higher pressure. ...
You might want to try setting the EPAP value down to 7 or 8 and see how it does. If it suffice and clears most of your apnea events, then you are probably good to go. Do keep a careful track on the AHI values. If those spike up by dropping the EPAP, then the higher value may in fact be needed.
justbreathe wrote:... Am I missing something and not understanding these settings correctly? ...
Nope. I think you have it.
Hope this helps.