Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 2:54 pm
by Paper_Nanny
StillAnotherGuess wrote:
Paper_Nanny wrote:Settings: BPM=0, Pres Max=12, EPAP Max= 12, EPAP Min=8, PS Max= 4, PS Min=0.
Typical ASV settings
Paper_Nanny wrote:Or is it still doing something ASVish because I wasn't able to set PS min= 4?
No. Currently PS Min=0 is sitting on your EPAP value. Think CPAP.
When I had it set with PS min and max=4, that was BiPAP? Why did the old machine let me set it that way, but the new one won't?
StillAnotherGuess wrote:I'm not your enemy.
Did you recently arrive at that conclusion or have you known that all along?

Deborah

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 3:01 pm
by Paper_Nanny
StillAnotherGuess wrote:I also enjoy your humor and your wit.
Hey, now!! Don't be trying to flatter me!! I am not a Southern girl!!

In fact, when Husband and I decided to move from the Midwest, we considered only destinations which were North of the Mason-Dixon line. After all the trouble caused by the flattery I got in FL, I'm scared to go into the South again.


Deborah

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 4:17 pm
by StillAnotherGuess
Paper_Nanny wrote: When I had it set with PS min and max=4, that was BiPAP?
Correct. That was BiPAP therapy (PS Min > 0) and ASV therapy (PS Max > 0)

Simple to remember for your machine:
CPAP therapy: PS Min = 0
Bilevel therapy: PS Min > 0
ASV therapy: PS Max > 0
Paper_Nanny wrote:Why did the old machine let me set it that way, but the new one won't?
I do not know. You need to have someone tell you how to access the Clinicians Menu on your new machine. I have never seen a Respironics BiPAP AutoSV Advanced, neither the old style nor the new PR-1 series.

Did you get the Sleep lab or doctors script for your new machine? How was it (supposed) to be set up?

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 5:16 pm
by Paper_Nanny
StillAnotherGuess wrote:
Paper_Nanny wrote:Why did the old machine let me set it that way, but the new one won't?
I do not know. You need to have someone tell you how to access the Clinicians Menu on your new machine.
The DME provider showed me how to access that menu and gave me the provider manual. Otherwise I wouldn't have been able to set anything on it!
StillAnotherGuess wrote:Did you get the Sleep lab or doctors script for your new machine? How was it (supposed) to be set up?
My original script from Mister Sleep Doctor was for an ASV. After my last appointment with MSD, an order was sent to the DME for a BiPAP set at 8/12. The RT at the DME saw no reason to give me a different type of machine when the one I had could be set to work in BiPAP mode. So, that was how I set it up, with some help from BrianinTN.

I have the new PR System One BiPAP ASV because I wanted a lighter weight machine with multiple humidity settings. So, there was a reason for that machine swap, but still no reason to swap the ASV for a BiPAP.

I just changed the settings on my machine so Min and Max PS are zero. Min PS= 0 and Max PS= 0. Zero is equal to both the Min PS and the Max PS. Zero Makes Me Hungry. Gotta go get a snack now.

Deborah

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 5:37 pm
by StillAnotherGuess
Paper_Nanny wrote: After my last appointment with MSD, an order was sent to the DME for a BiPAP set at 8/12.
I imagine the RTs screwed up when they set-up your new machine. They probably wanted to set PS MIN = 4 and PS Max = 0, but they got it transposed.
Paper_Nanny wrote:I just changed the settings on my machine so Min and Max PS are zero. Min PS= 0 and Max PS= 0. Zero is equal to both the Min PS and the Max PS. Zero Makes Me Hungry. Gotta go get a snack now.
You ARE now following NMs request.
I believe he wanted you to trial your new machine in CPAP mode only.
You have now put your machine in CPAP mode only (PS Min and PS Max = 0).
I assume you are using EPAP Min = 8 and EPAP Max = 12.

I'm sure NM and the board would be interested in seeing your CPAP (only) results.
NM and others may help you with recommendations from that CPAP baseline.

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 6:10 pm
by Dojers
Paper_Nanny wrote: Oh, Silly Guess, that by no means puts me in a unique position. It puts me in exactly the same position as all human beings. We are all solutions seeking a problem. We all have our own special blends of talents and gifts. Part of the human condition is a search to find ways to use those for the benefit of others, while at the same time, finding those whose talents and gifts will be of benefit to us.

It is with that in mind that I ponder your most recent posts to this thread. I am trying to understand why you feel drawn to me and my situation. Why do you feel the need to say these things to me? Are you here to give help me or are you looking for me to give something to you? Or looking for something from someone else here? I ponder, but come to no conclusions. Ponder and ponder... Ponder and wander, meander and contemplate, knit and purl, pearl and pear as I go from here to there.

As for being smarter than a fifth grader, the way I look at it is I don't have to be smarter because I'm bigger. And I am allowed to cuss and they're not.

Deborah
You just literally made me snort ice tea out of my nose - thanks a lot!!

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 6:15 pm
by Dojers
Ok, I am new to actually having a machine that I can read all this crazy info off of. How long does it take a semi-literate blonde to start to understand waveforms and what all those crazy acronyms mean?? I mean, I grok AHI and the OAs and CAs but don't know what's caused when by who did what where and how because of that doing what and so on (if any of that made sense - welcome to my world of insanity - thought I was the only one here! )

Just wondering.....oh, yeah, and how did/do you go about figuring out this stuff? Just keep reading the threads and trying to make sense of it all?

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 7:24 pm
by Paper_Nanny
StillAnotherGuess wrote:
Paper_Nanny wrote: After my last appointment with MSD, an order was sent to the DME for a BiPAP set at 8/12.
I imagine the RTs screwed up when they set-up your new machine. They probably wanted to set PS MIN = 4 and PS Max = 0, but they got it transposed.
How could min be higher than max? As for setting it up wrong, the RT didn't do that. The RT didn't set up my machine. I did that. Me. ME ME ME!!!
StillAnotherGuess wrote:
ME ME ME wrote:I just changed the settings on my machine so Min and Max PS are zero.
You ARE now following NMs request.
I believe he wanted you to trial your new machine in CPAP mode only.
I don't believe that.
Paper_Nanny wrote:
NotMuffy wrote:...a trial of low-level CPAP would have been, as is now, a very viable option.
To set my machine to function like a CPAP, I would set Max Pressure= EPAP Max= EPAP Min with PS= 0? If not, how do I do it? What pressure setting do you think I should go with? I would try it for a few nights or so, to see what happens.
And what came next?? The chicken or the egg? The herring or the dill??
StillAnotherGuess wrote:You have now put your machine in CPAP mode only (PS Min and PS Max = 0). I assume you are using EPAP Min = 8 and EPAP Max = 12.
I believe the appropriate response to that is: Un!#%ingbelievable. A tragic error has occurred. I keep hearing voices, but I cannot tell if they are in my head or not? Why amd I being locked out of the PS min???

Argh!!!

edit to delete two stray lines left from the quote of the original. they didn't belong in this post. they were just straggling here. no change to meaning.

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 7:31 pm
by BrianinTN
:: chews on some popcorn while watching the carnage ::

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 7:32 pm
by Paper_Nanny
Dojers wrote:How long does it take a semi-literate blonde to start to understand waveforms and what all those crazy acronyms mean??
About as long as it would take if the semi-literate blonde died her hair and became a semi-literate brunette. Or, even better, dye it so you become a readhead. That will take care of both the literacy and the blonde problems!!
Dojers wrote:Just wondering.....oh, yeah, and how did/do you go about figuring out this stuff? Just keep reading the threads and trying to make sense of it all?
Yeah, read a lot. Ask a lot of questions. Read more. And not just read on forums like this one, but also searching for stuff on other sites. But be careful when you read. I don't think anything has warnings about content possibly causing you to snort ice tea out of your nose.

Deborah

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 8:01 pm
by Mr Bill
Paper_Nanny wrote:
Mr Bill wrote:In that case is this evidence that the ASV is doing something beneficial right there that simple CPAP or BIPAP would not accomplish?
Is my machine doing things like an ASV? I thought with the current settings, it was acting strictly as a BiPAP would. Settings: BPM=0, Pres Max=12, EPAP Max= 12, EPAP Min=8, PS Max= 4, PS Min=0.

Or is it still doing something ASVish because I wasn't able to set PS min= 4?

Deborah
Well it seems to me you gave it 4 cm/H20 to play with between 8 and 12 so your normal report should look something like your night from 7/11 to 7/12. Plus we can see there are pressure pulses happening. So, yeah, its doing the ASV thing within those two pressures.

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 8:44 pm
by Paper_Nanny
Mr Bill wrote:Well it seems to me you gave it 4 cm/H20 to play with between 8 and 12 so your normal report should look something like your night from 7/11 to 7/12. Plus we can see there are pressure pulses happening. So, yeah, its doing the ASV thing within those two pressures.
I don't know why I didn't know that. Wishful thinking? I wanted it to be in BiPAP, despite that 4 cm H2O difference so I pretended like it didn't matter?

From the BiPAP autoSV Advanced System One Provider Guide (IMPORTANT! Remove this guide before giving the device to the patient. Only medical professionals should adjust pressure settings.):

PS Min- This setting is the minimum difference that is permitted between IPAP and EPAP. You may adjust the setting from 0 cm H2O to the difference between the maxiumum pressure setting minus the maximum EPAP setting.

So I can't run this machine like it was a BiPAP??

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 9:04 pm
by StillAnotherGuess
Paper_Nanny wrote: How could min be higher than max?
PS Min and PS Max are two different animals..
Keep repeating the mantra until it sinks in:
CPAP therapy: PS Min = 0
Bilevel therapy: PS Min > 0
ASV therapy: PS Max > 0
Paper_Nanny wrote:As for setting it up wrong, the RT didn't do that. The RT didn't set up my machine. I did that. Me. ME ME ME!!!
Amateurs, Who needs 'em?
Paper_Nanny wrote:Why amd I being locked out of the PS min???
Amateurs, but that would piss me off, too!

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 9:38 pm
by BrianinTN
StillAnotherGuess wrote: PS Min and PS Max are two different animals..
Keep repeating the mantra until it sinks in:
CPAP therapy: PS Min = 0
Bilevel therapy: PS Min > 0
ASV therapy: PS Max > 0
Wrong, wrong, wrong and also, wrong. (Yes, that's 4 wrongs -- one for the initially silly statement, and three for the next three necessary but not by-themselves-sufficient conditions.)

First off, let's clarify a definition, shall we? In Respironics language, PS (pressure support) is defined as the difference between IPAP and EPAP. If you set Min PS > 0, you are enforcing the constraint that some bilevel support will be employed -- i.e., IPAP will always be larger than EPAP. Max PS specifies the maximum bilevel support the ASV will attempt. I have seen the term "pressure support" used in various literature in different ways, but with respect to a Respironics device, this is what it means.

This is the second time in this thread that you have insisted (incorrectly) that "PS Max > 0" is "ASV therapy." Putting aside the obvious issue that nearly all ASV users employ a backup rate, the condition as specified is still not sufficient. Let's start be creating a random variable x, and specifying that PS Min = PS Max = x, where x > 0. The implications of that, ignoring ramping, are as follows:
  • The user's ASV will run in the equivalent of automatic bilevel mode
  • The difference between IPAP and EPAP will always be static (equal to x), and EPAP and IPAP will increase or decrease concomitantly and by the same absolute values
  • The maximum possible IPAP is constrained by the lesser of either Max EPAP + x or Max Pressure
  • The maximum possible EPAP is constrained by the lesser of Max Pressure - x or Max EPAP
  • The minimum possible IPAP would be min EPAP + x
To put these equations into an example, let's set x to 4 and specify the remaining variables as follows, preserving that condition: Min EPAP = 8, Min PS = 4, Max PS = 4, Max EPAP=15, and Max Pressure = 30. Under these parameters, possible pressures (when not ramping) would be 12/8, 13/9, 14/10,..., 19/15. IPAP will always equal EPAP + 4. Note how constraining Max PS to 4 means that there would be no functional difference at all between setting Max Pressure = 19 and setting Max Pressure = 30.

Similarly, PS Min > 0 is not a sufficient condition for straight traditional bilevel therapy either. If PS Min = x > 0 and PS Max > PS Min, IPAP will again be allowed to vary, but it will always exceed EPAP by an amount greater than or equal to x. Let's pretend this time that Min EPAP = 8, Min PS = 4, Max PS = 15, and Max Pressure = 30. (Note the only difference from my preceding example is that I have specified Max PS = 15). Some possible pressures (again not while ramping) would include 12/8, 13/8, 14/8, 13/9, 14/9, 14/10, 15/10, etc. The maximum possible pressure combination would be 15/30, and the minimum IPAP given that maximum EPAP would be 19/15. Bilevel always? Yes. Static difference between IPAP and EPAP, as most BiPAP and Auto-BiPAP users see? Nope.

And while I'm at it, PS Min = 0 is also not a sufficient condition for CPAP mode. A great many of us run our ASVs with PS Min = 0, including myself when using a FFM. All that does is permit the ASV to try IPAP = EPAP. It does not prohibit IPAP from exceeding EPAP. IPAP and for that matter EPAP can both bounce around, constrained of course by Max Pressure, Max PS, and Max EPAP. Setting PS Min to 0 by itself prohibits neither bilevel nor ASV therapy.
StillAnotherGuess wrote: I imagine the RTs screwed up when they set-up your new machine. They probably wanted to set PS MIN = 4 and PS Max = 0, but they got it transposed.
WTF? I'm not sure where you got the idea that Max PS can be less than Min PS, but once again, wrong. In addition to failing a basic common sense check, I tried to see if my ASV would let me do that. It doesn't. Score a victory for the little man and common sense.

Re: No Answers Yet For Problems With ASV

Posted: Fri Jul 15, 2011 11:00 pm
by dsm
BrianinTN wrote:
StillAnotherGuess wrote: PS Min and PS Max are two different animals..
Keep repeating the mantra until it sinks in:
CPAP therapy: PS Min = 0
Bilevel therapy: PS Min > 0
ASV therapy: PS Max > 0
Wrong, wrong, wrong and also, wrong. (Yes, that's 4 wrongs -- one for the initially silly statement, and three for the next three necessary but not by-themselves-sufficient conditions.)

First off, let's clarify a definition, shall we? In Respironics language, PS (pressure support) is defined as the difference between IPAP and EPAP. If you set Min PS > 0, you are enforcing the constraint that some bilevel support will be employed -- i.e., IPAP will always be larger than EPAP. Max PS specifies the maximum bilevel support the ASV will attempt. I have seen the term "pressure support" used in various literature in different ways, but with respect to a Respironics device, this is what it means.

This is the second time in this thread that you have insisted (incorrectly) that "PS Max > 0" is "ASV therapy." Putting aside the obvious issue that nearly all ASV users employ a backup rate, the condition as specified is still not sufficient. Let's start be creating a random variable x, and specifying that PS Min = PS Max = x, where x > 0. The implications of that, ignoring ramping, are as follows:
  • The user's ASV will run in the equivalent of automatic bilevel mode
  • The difference between IPAP and EPAP will always be static (equal to x), and EPAP and IPAP will increase or decrease concomitantly and by the same absolute values
  • The maximum possible IPAP is constrained by the lesser of either Max EPAP + x or Max Pressure
  • The maximum possible EPAP is constrained by the lesser of Max Pressure - x or Max EPAP
  • The minimum possible IPAP would be min EPAP + x
To put these equations into an example, let's set x to 4 and specify the remaining variables as follows, preserving that condition: Min EPAP = 8, Min PS = 4, Max PS = 4, Max EPAP=15, and Max Pressure = 30. Under these parameters, possible pressures (when not ramping) would be 12/8, 13/9, 14/10,..., 19/15. IPAP will always equal EPAP + 4. Note how constraining Max PS to 4 means that there would be no functional difference at all between setting Max Pressure = 19 and setting Max Pressure = 30.

Similarly, PS Min > 0 is not a sufficient condition for straight traditional bilevel therapy either. If PS Min = x > 0 and PS Max > PS Min, IPAP will again be allowed to vary, but it will always exceed EPAP by an amount greater than or equal to x. Let's pretend this time that Min EPAP = 8, Min PS = 4, Max PS = 15, and Max Pressure = 30. (Note the only difference from my preceding example is that I have specified Max PS = 15). Some possible pressures (again not while ramping) would include 12/8, 13/8, 14/8, 13/9, 14/9, 14/10, 15/10, etc. The maximum possible pressure combination would be 15/30, and the minimum IPAP given that maximum EPAP would be 19/15. Bilevel always? Yes. Static difference between IPAP and EPAP, as most BiPAP and Auto-BiPAP users see? Nope.

And while I'm at it, PS Min = 0 is also not a sufficient condition for CPAP mode. A great many of us run our ASVs with PS Min = 0, including myself when using a FFM. All that does is permit the ASV to try IPAP = EPAP. It does not prohibit IPAP from exceeding EPAP. IPAP and for that matter EPAP can both bounce around, constrained of course by Max Pressure, Max PS, and Max EPAP. Setting PS Min to 0 by itself prohibits neither bilevel nor ASV therapy.
StillAnotherGuess wrote: I imagine the RTs screwed up when they set-up your new machine. They probably wanted to set PS MIN = 4 and PS Max = 0, but they got it transposed.
WTF? I'm not sure where you got the idea that Max PS can be less than Min PS, but once again, wrong. In addition to failing a basic common sense check, I tried to see if my ASV would let me do that. It doesn't. Score a victory for the little man and common sense.

Hi BrianinTN

I realise the issue of settings on the Resp ASVs can get very confusing - just wanted to add some further interpretation ...

SAGuess posted
1. CPAP therapy: PS Min = 0
2. Bilevel therapy: PS Min > 0
3. ASV therapy: PS Max > 0

1 is true
2 is true
3 is true (with qualifications)

On an Respironics ASV (lets ignore epap min & max as they merely add to the confusion)

To set the machine to cpap mode:> you set a 0 PS min (so by default there is no epap-ipap gap ) which is cpap

To set the machine in cpap mode with ASV support:> (yes ! this is a valid setting ) you set PS min = 0 & PS max greater than zero (recomm 8-10). Thus whilst the machine runs as a cpap (epap=ipap), if it detects ASV type events the machine will use the PS max to raise pressure using ASV rules.

To set the machine to Bilevel mode :> you set PS min > 0 (creates an epap and an ipap) and you set PS max = PS min (removes ASV rules ). Also, backup rate works in this setting.

Whilst StillAnotherGuess doesn't always know his manners he does know his machine settings

DSM