Re: ASV: AHI still high, suggestions?
Posted: Thu May 26, 2011 8:39 pm
That's like asking why not trust the ASV's automatic algorithm. One reason is that it didn't seem to work for you. The other is that the machines do not learn from mistakes - they will go down to the minimum again and again, until jogged out of it - and when the minimum (in this case minimum PS leading to minimum IPAP) is too low, you'll be having the events that happen at the pressure the machine reverted to.BrianinTN wrote:Mr Bill wrote: So, Looking at your recent 90% EPAP at 7 and average PS of 3 results. I wonder if you would get similar results by setting Min EPAP at 7 and min PS at 3 or EPAP min at 6 and min PS at 4. In other words set you min EPAP low enough that your unit is occasionally bumping the EPAP up and set you min PS to 3 or 4 to give you a comfortable bilevel range. I'm thinking your min EPAP should be low enough that you should see your machine bumping it just a bit throughout the night. If you have no bumps in the Min EPAP, then it may be too high.I have a question for the two of you or JBF. I understand why min EPAP matters; the titration protocols call for addressing obstructive events by raising EPAP. From what I've read in your other posts, ozij, the ASV only prevents those events -- it does not "cure" them by opening the pathways, and because it takes time for the ASV to make adjustments, those are all reasons you want to set a proper floor for EPAP. (Let me know if I have misunderstood anything.)ozij wrote: I would consider a week or 2 at EPAP min =7 PS=3, I agree with Mr. Bills analysis. I would definitely use the way I felt as a guide to anything.
It's the PS part that has me a little confused. Let me quote -SWS from another thread (viewtopic.php?f=1&t=60687&st=0&sk=t&sd= ... 05#p574405) where he talks about experimenting with settings to improve AHI:Here's my question for you: since raising min PS is practically speaking equivalent to raising min IPAP, why not let the ASV figure out the appropriate level of PS?-SWS wrote: Here are some of the dynamics you might encounter in an exploratory grid like that:
- as PS Min decreases: a) iatrogenic central events might decrease, but b) primary central events might increase, and/or c) obstructive events might increase
- as PS Min increases: a) iatrogenic central events might increase, but b) primary central events might decrease, and/or c) obstructive events might increase
It keeps away the obstructive events that appear when the IPAP is lower.At EPAP=7 and PS=3, you're basically enforcing a floor of IPAP=10. What does a higher IPAP floor accomplish?
That depends on what you need. Some people feel choked if the IPAP is too low.from Mr. Bill, you mentioned a "comfortable bilevel range" -- but isn't a higher baseline inhaling pressure going to be inherently less comfortable?
It won't necessarily stay there, and and average does not tell youi how much time you are at each pressure.And if the higher IPAP is what's required to eliminate the hypopneas, isn't the ASV going to "get there" (and stay there) anyway? Most nights on the ASV, I've seen an average PS around 3, even when min PS is set to 0.
The ASV in "out of the box Auto mode" did not work for you at home.In a nutshell, I thought this area was exactly the area where an ASV is supposed to shine compared to conventional treatment modalities like an Auto BiPAP -- so I'm trying to better understand the theoretical background for why applying this particular constraint to the ASV should improve its performance.