Re: ResMed S9 AutoSet v DeVilbiss IntelliPAP AutoAdjust
Posted: Sat Oct 30, 2010 1:26 pm
As indicated earlier, I am continuing with the collection of data from the ResMed S9 AutoSet and the DeVilbiss IntelliPAP AutoAdjust - now combining that with collection of ODI (Oxygen DeSaturation Index) from the SPO 7500.
I changed the settings on both APAP machines to the following:
Automatic Mode (from fixed CPAP)
Lower Limit 6.0 cmH2O
Upper Limit 10.0 cmH2O
Ramp for 10 minutes starting at 4.0 cmH2O
Relief (EPR/SmartFlex) ON
Results as follows:
Oct 28/29 (S9) - AHI 0.4, ODI 10.0, 95th Percentile 8.7cmH2O
Oct 29/30 (IP) - AHI 1.6, ODI 15.0, 95th Percentile 6.0cmH2O
Here are the graphs:
Oct 28/29 (S9):

And for Oct 29/30 (IP):


One of the things that is becoming clearer to me is that the DeVilbiss algorithm intentionally seeks the lowest possible pressure, and its increases and decreases in pressure are gradual so as to not cause an arousal from the pressure changes. I am growing to appreciate those aspects of the DeVilbiss algorithm very much.
I am also finding more 'meaning' in the AHI reported by DeVilbiss. Though the DeVilbiss AHI recorded 1.6 last night *seems* to indicate more events than the S9 reported AHI from the night earlier, it was nearly the lowest AHI ever recorded by the IntelliPAP - and for a change, I do *feel* my sleep was improved.
Offered FWIW
I changed the settings on both APAP machines to the following:
Automatic Mode (from fixed CPAP)
Lower Limit 6.0 cmH2O
Upper Limit 10.0 cmH2O
Ramp for 10 minutes starting at 4.0 cmH2O
Relief (EPR/SmartFlex) ON
Results as follows:
Oct 28/29 (S9) - AHI 0.4, ODI 10.0, 95th Percentile 8.7cmH2O
Oct 29/30 (IP) - AHI 1.6, ODI 15.0, 95th Percentile 6.0cmH2O
Here are the graphs:
Oct 28/29 (S9):

And for Oct 29/30 (IP):


One of the things that is becoming clearer to me is that the DeVilbiss algorithm intentionally seeks the lowest possible pressure, and its increases and decreases in pressure are gradual so as to not cause an arousal from the pressure changes. I am growing to appreciate those aspects of the DeVilbiss algorithm very much.
I am also finding more 'meaning' in the AHI reported by DeVilbiss. Though the DeVilbiss AHI recorded 1.6 last night *seems* to indicate more events than the S9 reported AHI from the night earlier, it was nearly the lowest AHI ever recorded by the IntelliPAP - and for a change, I do *feel* my sleep was improved.
Offered FWIW