Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Thu Sep 30, 2010 9:06 pm
by jdm2857
Unfortunately, the S8 ResLink SpO2 socket is mounted directly to the circuit board. That makes swapping it out much more difficult. I'm not good enough with a soldering iron to risk cooking components on the board.

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Fri Oct 01, 2010 7:33 pm
by DreamDiver
casadelane wrote:I got it to work. ...]
Fantastic! I'll have to wait until I'm the lottery winner before I can do this, though.
It's great that you got it to work.
In our previous attempts at looking at this, it had always been about working around the existing pin-out.
Changing to a standard pin-out architecture makes total sense.
Thinking inside and outside the box.
Excellent.

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Sat Jan 15, 2011 10:33 am
by ze_
casadelane:

Nice job - My 10-yr old Autoset T broke last week and I was about to embark in a similar journey to get O2 data embedded in the S9 stream. Knocked at Nonin door already, had the development kit ready to be ordered.
I was going to try to identify the signals in the edge connector in the back of the S9, but since you got an adapter and had it open - can you help me with a shortcut? Is that expensive adapter simply a 3-wire job or does it have any components in it? Resmed might have gone the inkjet cartridge route - eeprom to ID the module, or simply replaced a DB-9 with an edge connector.
Could you post a picture of the pcb, and perhaps the connectivity, from the xpod lemo to the edge fingers?
I would buy a dev. kit, make a pcb with the lemo and some keying to allow it to go only one way into the S9. Would give away the extra boards (minimum lot for PCB).

Next step would be to find out from Nonin which cheaper solution would be there in the long run - perhaps some other product from them has the same data format inside that can be tapped into.
Another path would be to write a little app to convert logged night data from other oximeters into Resmed format and add to the SD card - problems are to accurately sync them and figure out Resmed's CRC polynomial.

Perhaps Resmed, the FDA, DMEs or "evil forces" will turn around and allow patients to have access to their own data. Imagine if diabetics would have to go to a doctor or "sugar specialist" to get their sugar readings...

Thanks

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Sat Jan 15, 2011 11:18 am
by avi123
dsm wrote:In particular ask if the S9 accepts a Nonin 3012 or Nonin 3212 - these are stated as working with the ResLink as per the latest ResLink manual.

(I'll be asking them too but last time I had no luck despite having aqquired some 3012 & 3212 units - I wrongly assumed the 3011 & 3211 were the correct SpO2 units & tried to adpat one but it just didn't work !!! - Now I know I was WRONG !. Ahh well we live & learn but I do have Nonin 3012 & 3212 units With luck these will work with the S8 Reslink AND the S9 SpO2 adpater.

Many thanks

DSM
I don't understand the issues in this thread. Can you explain?

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Sat Jan 15, 2011 12:01 pm
by HoseCrusher
The problem involves integrating pulse oximeter data into the flow data from the machine. Some models have an accessory SpO2 monitor, but it is very expensive. I think the idea is to use a less expensive finger probe and just wire it up to the machine.

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Fri Jan 21, 2011 1:49 pm
by StevenXXXX
Does anyone have the current prices (or a link to those prices) of the ResMed S9 Oximeter Adapter and the Nonin Xpod Oximeter, etc. IF they were all bought through a ResMed DME supplier?

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Fri Jan 21, 2011 6:28 pm
by casadelane
The main problem is the the ResMed S9 Oximeter Adapter has a connector that is unlike anyone has ever seen before and Nonin does not sell their Xpod Oximeter with a compatible interface to the Resmed Adapter. You have to take apart the ResMed S9 Oximeter Adapter and replace its connector to one that will work with Nonin's Xpod. The Nonin Xpod can be purchased directly from Nonin as an OEM kit for $300 (see my post before) but the Resmed S9 Oximeter is VERY hard to find and purchase. You have to check with your DME supplier to see if they can special order one for you.

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Sat Jan 22, 2011 10:09 am
by StevenXXXX
ResMed's Confidential Home Care Provider Product List (effective January 2011) shows on page 30 that a Complete Oximeter Kit can be purchased which includes the Item # 36940 S9 Oximeter Adapter and the Item # 36942 S9 Oximetry Accessory Kit (shown on page 31) which includes the reusable soft sensor, 3M cable, sensor clip, & Xpod).

So, it looks like everything could be acquired from ResMed through a reseller.

That ResMed Confidential Home Care Provider Product List (which does not have prices) can be found at
http://www.resmed.com/us/documents/product-list-usa.pdf

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Sat Jan 22, 2011 2:38 pm
by Slinky
Oh, yeah, you CAN buy the kit - IF you have loads of money to just blow. The S8 ResLink oximeter kit ran about $800 if I remember correctly.

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Mon Mar 12, 2012 11:28 pm
by -tim
I now have a working "kit" for my S9.

It consists of a "36940": S9 Oximetry Adapter which is the expensive resmed made thing that goes on the back of the S9.
22304: ApneaLink "Accessorises for Oximetry" which consists of a Nonin adapater and one of their sensors which gets taped over a finger.

My fathers DME was having a "get rid of this junk to the first person who asked for it sale". I also ended up with assumed to be worthless USB adapter and few hundred tape things and two Nonin 7000A adapters.

I expect that a full working system could be made up using the 36940 and a magic connector and standard Nonin parts.
Does anyone have a list price on the 36940?

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Tue Mar 20, 2012 12:25 am
by shb
The S9 adapter (ie Resmed's 36940) can be bought (by anyone in public) and a delivered from this site for $279

http://www.cpapusa.com/search/ResMed-S9 ... T2501.html

But it does seem very difficult to get it all working...

DSM... you still got those 10 (qty) nonin 3012 units ?

I've been using an S8 since 2005 - and am upgrading to an S9. I was thinking about shelling out for the Oximetry module to get that data across into my resScan graphs - but - still not quite sure what I need to buy.. I liked DSM's thought back a few months to buy any generic cheap oximetry monitor -BUT THEN WRITE SOME SHAREWARE - to format the data onto the SD card in the same format - then resScan will just process it as if it were produced by the "$1400 (approx)" Resmed/Nonin proprietary system.

Who ever has got it working - can you share the data file - I'd be happy to help reverse engineer and write the app...

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Tue Mar 20, 2012 12:40 am
by shb
Hopefully some of the brains trust that worked this issue through over 18 months are still watching this thread ????????

With an Resmed S9 auto - I'm thinking I only need to get the following two parts:

http://www.nonin.com/documents/OEM%20De ... mation.pdf ($300)
http://www.cpapusa.com/search/ResMed-S9 ... T2501.html ($279)

And then the issue is making a plug that will connect it. Someone please correct me if I left out something...

The other option that DSM suggested early in the thread - was to write bit of shareware - that takes the output from any consumer oximeter (i think they are often priced around $100 total) and drops it onto the SD card so that ResScan will pick it up as if it had come from the full blown Resmed system.

To do that - I think we really just need a sample file -- what is on the SD card --- for those who got it working ??????? Can someone please post an example ?

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Wed Apr 04, 2012 8:17 pm
by rxs0
How to Modify the ResMed S9 Oximeter Adapter with the Nonin 3012 Xpod (format 2) Oximeter:

Given the cheapest price I could find on the ResMed S9 Oximetry Complete Kit was about $1000, with internet advertised prices at $1325 (http://www.cpapusa.com/cpap-sleep-apnea ... T2822.html), I decided to attempt the modifications that casadelane originally posted in this thread:
(viewtopic.php?f=1&t=50805&start=45)

I modified the ResMed Oximetry Adapter to work with a standard PS/2 mini-DIN 6 pin cable. I again used many of the ideas that casadelane described in his great post. Unfortunately, I am unable to access his attached images. I wrote these instructions so that no pictures are required. I can post pictures to this thread at a later time.

The PS/2 cable is exactly the same cable used for non-USB keyboards and mice. I decided to use this cable given its availability at most computer stores and the option for an extra extension cable once you have completed the project for additional length. You can however use whatever connectors you desire such as the Lemo connectors. Cut the 6 foot PS/2 extension cable in half. One half will have the female connection which I used to attach to the ResMed Oximeter Adapter. The other piece will have a male connection which I used to attach the Nonin 3012 Oximeter device.

Parts List:
- ResMed Oximeter Adapter purchased from http://www.cpapusa.com for $225 (free shipping). I got a better deal than the advertised price of $279. Ask Blake for their best price.
(http://www.cpapusa.com/cpap-sleep-apnea ... T2501.html)
The ResMed manual can be downloaded at:
(http://www.resmed.com/int/products/s9-a ... c=patients)

- Nonin OEM 3012 Xpod kit, 8000AA sensor (Adult Articulated Finger Clip Sensor) with data format 2 for data logging purchased from Nonin. Call 800-356-8874 and ask for Myra in customer service. Be sure you ask for format 2 which allows data logging (transmits 5 bytes, 75 times per second). Here is the order form. Select Xpod, 8000AA, and Data Format 2 on the form. Fax or email the form back to Myra, attention OEM: Fax 763-577-5521 or email oem@nonin.com. She will ask you if it is for commercial or research purposes. I said it was for "research".
(http://www.nonin.com/documents/OEM%20De ... mation.pdf) for a total of $327 ($300 for the Xpod and sensor, $21 sales tax, and $6 ground shipping).
Here are the specifications for the Nonin Oximeter in PDF format.
(http://www.nonin.com/documents/Xpod%20S ... ations.pdf).

- Standard 6 foot PS/2 extension cable - $6-10 from Best Buy. The cable has one male and one female connector.

- Other items used included thin solder, soldering iron 35watt, 3-pin computer fan power Y-splitter cable (which I bought at Fry's electronics for use of the Red, Black, and Yellow wires), various heat shrink tubing sizes (3/64", 1/16", 3/32", 3/8"), inexpensive heat gun from Harbor Freight, hemostats, needle nose pliers, metal cookie sheet pan for soldering tray, lighter to burn off plastic insulation, other miscellaneous items.


1: Remove the cover off the ResMed Oximeter Adapter by gently pulling the housing apart. Be sure to take note of how the unit is assembled. I took multiple photographs with my iPhone. Avoid touching the internal chips and circuitry. Gently pull off the tiny plug with the three wires (Red, Black, Green) from the circuit board. There is a small nut the holds the ResMed 3-pin black connector to the plastic housing. ResMed refers to this 3-pin proprietary panel connector as "Adapter Connector". The nut can be removed with a pair of needle nose pliers or a pair of hemostats. As mentioned above, I removed the wire plug from the circuit board and removed the ResMed 3-pin housing connector. I then gently removed the donut from the wire. Do not force the plug through the donut. Take your time. By now, you have a separate 3 wire cable (Red, Black, Yellow wires) with the tiny circuit board plug at one end and the larger 3-pin black ResMed connector at the other end.

2. You can either replace the proprietary 3-pin black ResMed connector with the free end of the PS/2 female portion of the cable (First Method) or keep the ResMed connector intact and solder new contacts along the wire length as described below (Second Method). It is probably easier to perform the First Method. You can also attach a 3 or 4 pin Lemo connector to the 3-wire ResMed cable as casadelane did.

2a. First Method:
Remove the 3-pin ResMed proprietary connector from the small 3 wire cable by melting the solder welds. Use wire crimps or use solder to attach the new female PS/2 cable to the now free end(s) of the 3-wire ResMed Oximetry Adapter cable. I used the PS/2 red, white, and bare wires. I covered the bare wire with black heat shrink tubing for insulation. You can use whatever wires you wish from the PS/2 cable. You just need to be consistent with the other male PS/2 cable. It might be better to use the red, green, and black or bare cables since this will match the relevant cable colors of the Nonin 3012 Oximeter. Cover the solder welds with appropriately sized heat shrink tubing. You can use a lighter or an inexpensive heat gun from Harbor Freight. This method is the easier way to go. You will just need to find a rubber grommet to fit the remaining space between the housing hole and PS/2 cable. You might be able to find the rubber grommet at Lowes hardware section. Alternatively, you can just substitute the ResMed connector with a small 3 or 4-pin connector such as the Lemo connector similar to what casadelane explained in his post. You will likely have to enlarge the ResMed Housing "hole" to accommodate the new connector.

2b. Second Method:
Using cut wire from a 3-pin computer fan power Y-splitter cable and appropriately sized heat shrink tubing, I soldered 3 like colored wires (Red, Black, Yellow wires) approximately 2-3 cm from the end of the ResMed black 3-pin plug. This will allow you to still be able to wrap the ResMed 3-wire cable around the donut on the other end of the cable when you reassemble the unit. I did not cut the original cable wire. I simply burned off a small amount of plastic insulation to access a small amount of bare wire underneath. I then placed a small piece of heat shrink tubing over the new jumper wire and "pulled in" a small amount of the original wire into the shrink tubing for insulation. This takes on a Y shapes configuration. I then drilled a small hole on the left upper side of the ResMed connector (that is if you are facing the front/3-pin connector) to allow access for the female end of the PS/2 extension cable. I drilled the hole just a bit smaller than the cable for a snug fit by hand only. I drilled at this spot taking into account available space that the wires will take up when the unit is reassembled. Insert about a half inch of PS/2 cable into the ResMed Housing. Don't push too much cable in or you won't have the space to reassemble the unit. I then soldered the newly placed "jumper" wires to the ends of the PS/2 wires, which were covered with heat shrink tubing. Be sure to put the heat shrink tubing on before soldering. Also be careful not to prematurely heat the shrink tubing before positioning the tubing over the new solder weld. I made this mistake several times. Also, you can use a short segment of 3/8" heat shrink tubing over the PS/2 cable inside the Oximeter Adapter housing to prevent the cable from being pulled out.

2c. Third Method:
The last method would be to try to recreate a proprietary male connector to attach to the ResMed 3-pin female Black "Adapter Connector". I have not tried this yet. This would involve buying connector pins compatible with the Connector, a plug to hold the pins in place, and a way of securing the male connector to the female ResMed connector. There is a small circumferential groove around the back waist of the ResMed plug in addition to a small bar in the lower 6 o'clock position of the ResMed connector that locks the connectors into place. So far, no one has been able to find a compatible male connector to the proprietary female ResMed adapter connector.

3. Reassemble the ResMed Oximeter Adapter. Be sure to take note of the orientation of the ResMed port connector which plugs into the ResMed S9 machine when you dissemble and reassemble the unit. ResMed refers to this port as the "Module/Adapter port". It has a small notch at one end of the connector to allow it to plug into the ResMed S9 unit in one direction only. I took various photographs with my iPhone to remember how the adapter is assembled. You now have the female end of the PS/2 cable sticking out of the ResMed adapter housing. If you used the second method above, the original ResMed 3-pin black plug in remains intact.

4. Solder the male end of the PS/2 extension cable to the Nonin 3012 cable. Be sure to order this cable without the connector with the wires at the end exposed to make it easier to gain access to the wires. It is very difficult to remove the plastic insulation from the Nonin cable. Be sure not to damage the wires underneath. As mentioned earlier, the wires are documented correctly in the Nonin PDF specification above. I will discuss this below. When you order the Oximeter from Nonin, you have multiple options including a mini-12 connector, USB connector, Lemo 4 pin connector, or without a connector. It would probably be easier to order it without a connector and the wires exposed and ready to solder. Take your time. Be sure to plan placement of you heat shrink tubing over the wires and cables prior to soldering. I placed a long 3/8" heat shrink tubing over the newly joined cables (male PS/2 and Nonin 3012) at the soldering site for wire protection and a professional appearance. This also helps prevent the cables from being pulled apart in case someone trips over the cable.

The correct wire-to-wire connections are as follows:

(-): ResMed Red wire connects to the Nonin Bare/Cable Shield wire.

(+): ResMed Yellow wire connects to the Nonin Red wire.

Serial Output: ResMed Black connects to the Nonin Green wire.

Make sure you keep the PS/2 wires consistent and maintain the connections above.

The Nonin Inputs/Outputs are correct as documented in the PDF file above. The Nonin Red wire is the Power Input, Nonin Cable Shield is Ground/Negative, and the Nonin Green wire is the Serial Output wire. I also confirmed this with a multimeter.

ResMed however uses an unconventional wiring scheme in the Oximeter Adapter. The ResMed Power Input wire is the Yellow wire; Ground/Negative is the Red Wire; and the Serial Output is the Black wire. This is weird, but is works. Just be sure to keep your connections correct. I again confirmed this with a multimeter.

All three wires are needed for operation. All three wires have to be connected for the sensor light to illuminate and the ResMed S9 machine to recognized the Oximeter.

Be sure to put your sensor clip on your finger before starting your therapy or the machine will report the following error, "Check oxi sensor or press dial to accept no oximetry data. This is mentioned in the troubleshooting section of the ResMed Oximeter Adapter manual. You will be able to see the finger sensor red light illuminated once connected to the unit. It will remain illuminated even if the unit is powered off but remains connected to the wall outlet. Also, you will see the live SaO2 and Pulse data in the S9 display while in therapy mode. You can also review overnight O2 saturation readings with the SD card and ResScan software 3.16.

I hope this helps.

Thanks again to casadelane for the original plans. Let me know if you need any photos.

Take care,
Rich

Re: S9 SpO2 module - can anyone ask their rep what one ?

Posted: Thu Apr 05, 2012 5:29 am
by -tim
I'm trying to find the connector. It appears to me a micro- or nano- EXF connector that someone tried to make. Note that it is not a "mini-exf" which is about 20% too large. The outer diamater of the socket is .02 inches or 5.04 mm which seems to indicate that its a clone of an US connector. There are no markings at all on it which means its made by someone hiding from patents. One end seems to be factory terminated with a small rectangular plug so someone in China is willing to send me a box of these things for a few hundred dollars. When I say factory made, this isn't done at the same place that made the S9 in Australia. Its not done the same ways. I'm also looking for other options.

There are 3 parts to the S9 system. The "Resmed Oximeter Adapter", the Model 3012 XPOD adapter and then an ?7000? or ?8000? sensors.
The Resmed Oximeter Adapter" seems to be an isolation power supply and might be rs-232 level shifter or even an rs-232 to 485 type converter. It may even do protocol conversion. It seems to contain an isolated 3.3 or 5.0 v power supply and may have optical isolators. That sort of stuff is typical for medical gear just in case the patient gets a climate line hose cut and O2 probe cut and have bits of each conducting across the heart. I expect a compatible device could be made but the latches would be a pain as would a case. It might be best to pay the $2xx to resmed for this.

The XPOD is a small computer that reads the sensor and sends RS-232 like data down the line. One end is a standard DB9 and the other end can be ordered with out a connector. Resmed seem to attach some cheap connector they found on the end and repackage the things in the bags that Nonin sent them the XPOD in. There are other options for this but I haven't exploored it yet. The word in the medical halls is that Nonin is expensive.

The sensor is two LEDs, one red and one IR and sensors. Pulse rate and O2 levels are determined by looking at the changes in each with each beat of the heart. There are a few types of these from a clamp on to a reusable tape on to a one time use tape on sensor.

Once I recover from spending way too much on the kit I mentioned above, I've told dsm that I would buy some of his stuff and see if I can't find a way to make a cheaper kit.

I'm currently looking at building an extension adaptor that would allow one to examine or tap into the signals between the S9 and the resmed adapters. If they work out, I may end up with 5 or 10 of them incase anyone has a use for one. I'm guessing they will be about $10 or free except to addresses in Bella Vista in which case they are a few grand dearer due to "market conditions".

I expect that a device could be made to feed in two values into the S9's data stream. Maybe something from Xen, maybe a temp probe, maybe a phase of moon detector, I don't know but someone might find something useful. I also wonder what happens if we feed in 3 or 4 values but that is for later.