coreyg wrote:Hi Mars,
In Patient Mode:
Max Latency Time: 0
Min Pressure: 12 (Last night I increased this from 10)
Max Pressure: 16
In Clinician Mode:
Mode: AUTO
Min Pressure: 12
Max Pressure: 16
Comfort Pressure 12 (This automatically increased when I changed min Pressure)
Max Latency Time 0
Latency Time 0
When I get up at night, I take the mask off and I do not turn the sandman off or touch any of the buttons. Within about 10 to 15 seconds, the pressure drops a lot all of a sudden. It's still running but the pressure is very low. When I download the data to the computer and look at the graphs to see what the pressure is set to, it reads 0.7. When I'm about to put the mask back on, I press the on button and the pressure goes back up (I assume to my min pressure) and then I put the mask on. I'm a bit worried that being half asleep, I may forget to press the on button and put the mask back on at the extremely low pressure. So far, I don't think I have but I was wondering what it would do. Will it recognize that I have the mask back on and increase the pressure to where it's supposed to?
Thanks for your help. I think all of the comments about Comfort Pressure where leading in the wrong direction. It doesn't really matter but I'm still not sure what they mean by "no respiratory cycle". I keep thinking that means, if I die (ie stop breathing) with the sandman on, it will drop to the comfort pressure. That really makes me feel better
And thanks for the links on the Command on IFL. I haven't read it yet, but I will.
For last night I got:
OBS AI: 1.2
CNT AI: 0
OBS HI: 6.8
CNT HI: 2.9
Snorings (SNG): 81/hour
Runs: 25
The snorings seem to be what's always increasing my pressure. The CNT HI is higher than normal last night, but that was just one night, I'll see what it does over the next several nights.
-Corey
Hi Corey
Lets see if I can respond to the above one by one -
When I get up at night, I take the mask off and I do not turn the sandman off or touch any of the buttons. Within about 10 to 15 seconds, the pressure drops a lot all of a sudden. It's still running but the pressure is very low.
I would think that doing the above would skew your results, and I cannot think of any advantage it would have. No doubt there are different sequences that suit different people, but correct data is always a must.
What I do is first of all put my mask on, then get a comfortable fitting, then turn the machine on, then get comfortable again, then lie down. For getting up I turn the machine off, then take the mask off. This will allow accurate feedback.
When I download the data to the computer and look at the graphs to see what the pressure is set to, it reads 0.7.
Are you saying you download to your computer in the middle of the night, whilst the machine is still running?
When I'm about to put the mask back on, I press the on button and the pressure goes back up (I assume to my min pressure) and then I put the mask on.
So the machine is still on, you have a pressure of 0.7, you press a button and the pressure goes back up - if you press the bottom right button at this point the machine will turn off.
The pressure would only go back up if you have the mask on, it would not go back up to your minimum pressure if you have the mask off. So the "and
then I put the mask on" does not seem likely.
I think something is missing from this sequence - could you please check it and advise.
I'm a bit worried that being half asleep, I may forget to press the on button and put the mask back on at the extremely low pressure.
If you put the mask on when the machine is on, it will automatically go to your minimum pressure. You do not have to press a button for this to happen. But as that method will give you inaccurate results, I suggest you give my suggested sequence a try.
So far, I don't think I have but I was wondering what it would do. Will it recognize that I have the mask back on and increase the pressure to where it's supposed to?
As stated above, if the machine is on, and the mask is on, it will deliver the pressure.
Should you have the mask on, and the machine off, the most likely thing is that you will sleep without therapy. Or you may not sleep, and realise the machine is off.
I think all of the comments about Comfort Pressure where leading in the wrong direction.
Comfort pressure can be set, but is only activated when both latency times are input. Your latency times are zero, therefore you have never experienced the comfort setting with that zero condition.
Latency times in Auto are similar to ramp times in CPAP.
It doesn't really matter but I'm still not sure what they mean by "no respiratory cycle". I keep thinking that means, if I die (ie stop breathing) with the sandman on, it will drop to the comfort pressure. That really makes me feel better
I think we need to get sorted out on all this other stuff before you worry about "respiratory cycle". It doesn't mean what you think it means, and if you do die with the Sandman on, it will not bring you back to life.
Anyhow, I think the definition is that a respiratory cycle is one breath. For more information on all that stuff (and a lot of other esoteric stuff) you could go to -
viewtopic.php?f=1&t=35298&st=0&sk=t&sd= ... le+sandman
if you have a few hours to spare.
A simpler explanation is at
http://www.answers.com/topic/what-is-th ... tory-cycle
Your total Obs AHi is 8.0, and that needs to be brought down to 5.0 or under, as well as the high number of snoring events.
Lets start sorting out all the other stuff, and go from there.
When you describe a sequence of events, please be very careful to be accurate.
Don't get too hung up over one night, it is trends that count.
cheers