Re: POLL: Typical AHI for ResMed Users
Posted: Fri Sep 04, 2009 1:14 pm
Velbor wrote:No failure of logic. Just a sense of incompleteness. And defensiveness on my part, with overreaction to a non-existent "accusation" that I might be "unhygenic". Since my own data was being used as an example for discussion, I would much prefer that my own "excessive wakefulness" not be attributed to any presumed violation of the standard "sleep hygiene" rules (which, again being defensive, I would insist that I follow rigorously - being a rigorous - read obsessive - sort of person by nature).-SWS wrote: So where exactly is the above logic failing?
By the way, I regularly cycle through a careful regimen of Ambien or Klonopin, not to get to sleep, but with the goal of lowering my threshold for waking through the night. Codeine works best, but it's not something a "hygenic" person would want to take regularly!
Sorry, didn't realize that this was a puzzlement. The source of information for my now infamous chart, the source for all of the Respironics and ResMed definitions (the PB data was contributed by ozij) is direct quotation from manufacturer clinical manuals. Here is an image of the Respironics defintion page from the M-Series Auto manual:-SWS wrote: I'm still hoping to hear more about those Respironics summary definitions from my query a mere twelve posts up. And thank you for that, BTW! For instance, I'm wondering if the Respironics hypopnea definition(s) were compiled from various patent descriptions, clinician set-up manuals, training or marketing literature, etc. There seems to be a certain amount of discrepancy among the hypopnea definitions contained in the various Respironics text media.
Thanks!! Velbor
Thanks, Velbor! So those two recovery breaths are clearly a hard hypopnea-scoring requirement for the M-Series auto. The 50% amplitude reduction is clearly at odds with the clinician's training simulation claiming 40% amplitude reduction instead.
Also a comment that manufacturers probably don't release every little implementation detail about their algorithms. So while Respironics clearly requires those two recovery breaths for hyopnea scoring, I don't think we can comfortably assume that Resmed doesn't also bother to look---in one way or another--for that same post-hypopneic survival reflex (the large-volume recovery breaths).
A part of that Respironics decision-weighting criteria includes spotting FL in the hypopnea's valley. That would probably imply an obstructive component is primarily associated with that hypopnea's etiology, versus central hypoventilation for instance.
I'm still thinking that some of what Resmed scores as hypopneas may be considered hypoventilation events by Respironics instead.