Re: Respironics / EncoreViewer Data Error ?
Posted: Thu May 28, 2009 3:44 pm
ozij wrote:You have to find out where Respironics starts its day and where it ends it, and how it handles sessions.
I have come to learn that suggestions by Ozij should never be taken lightly. She has a gift for considering “outside the box” possibilities. Her suggestion that an explanation of apparent “apneas” of less than 10 seconds duration might be explained by “day” or “session” considerations led me on the following path.6PtStar wrote:I did not look at the 9 second days but all the 6 second days (April 5,6,8 &12 and May16). The 6 second apnea was the only one. All had a OA of 0.1. Respirionics starts it's day at 12 noon and runs until the next 12 noon unless there is a segment of more than one hour.
Jerry has indicated that the Respironics “day” runs from noon to noon. This is the same protocol utilized by ResMed, which is well documented. I cannot recall having seen as unambiguous a statement from Respironics, but I have no reason to doubt it: the Help file for Encore Viewer, “Patterns of Use” indicates, “Across the bottom of the graph is the time of day starting with 12:00 noon.” And “The usage time is counted over a 24-hour time period ….”.
Here is the schedule of my therapy utilization on the night in question (the evening of 25 May 2009). Entries for “Clock Actual Time” are recorded to the nearest quarter-hour):

Thus it is seen that I was out of bed 5 times during this night; one of my arisings was prolonged and atypically involved the hungries and getting a snack. My experience in analyzing Respironics data is that a break of 1 hour 15 minutes DOES cause the machine to begin counting a new “session”, but a break of 45 minutes, as on this night, does NOT result in a new session. The critical cut-off time is presumably one hour.
What is important to note in this context is that all times were within the same “day” from noon to noon. BUT, that is true only for my local time.
We have on a number of occasions been advised on this Forum that the Respironics internal clock (which, unlike ResMed models, cannot be reset by the user) is set to Coordinated Universal Time (UTC) / Greenwich Mean Time. I think that my own data collection can corroborate this. Extracts from the EncoreViewer .xml file for this date show the following:

The .xml file from which this was extracted is time-stamped as having been created on 26 May 2009 at 9:18am, when I ran EncoreViewer. This is precisely the information contained in line 2 of the .xml file, which presumably obtained the information from my computer’s clock.
My initiation of therapy for the night of the evening of 25 May 2009 was at about 12:15am on 26 May, which would be about 4:15am UTC. According to line 47 of the .xml file, the “Therapy Session Date” entry bears almost exactly this time-stamp, presumably generated by the REMstar’s internal clock. Note that line 46 shows that there was only ONE “Therapy Session”, and that line 47 shows the duration of this session without reference to any of my “blower-off” arisings (equal to the final blower-off relative time). [Remember that my calculation of “average duration of apnea” – and presumably James Skinner’s - is not based on either session time nor blower time.]
With this evidence that the machine’s internal clock is based on UTC, is it possible that the Respironics data management system is constrained to cumulating “Average Time in Apnea Per Day” in terms of a fixed 24 hours based on noon-to-noon UTC? My utilization ranges from 04:15 through 12:30 UTC. Of my eight (or perhaps seven) apneas, one occurred at 8:12:04 local, or 12:12:04 UTC. Might the duration of this apnea not have been counted for 5/25/09, but “held over” for 5/26, even while the presence of the apnea itself was counted during 5/25?
This explanation seems possible, though not at all comforting.
The EncoreViewer, presented with data from the evening of 25 May, but which in fact is generated entirely on 26 May, nonetheless “correctly” presents it as data of 25 May, (since 25 May started at noon and does not end until noon - presumably UTC noon, since that is all the Respironics internal clock knows - of 26 May)

The EncoreViewer is able, once the smartcard is downloaded, to make the “UTC to local time” conversion:

It presumably can get the 4 hour time-difference information (normally 5 hours, now adjusted for DST) from my computer:

To review: EncoreViewer presents “days” and “times” correctly for local time (assuming the REMstar internal clock is reasonably accurate). Detail data (e.g. occurrence of an apnea, even if it occurred during the “next” day, after noon UTC) is presented properly, presumably because the Respironics blower firmware is “smart” enough to deal correctly with a single “session”, as Jerry's comment suggested.
But we do not know how the Respironics firmware is cumulating the “Time in Apnea Per Day” data. This is where Ozij’s hypothesis is most intriguing. It does seem possible. If the apnea durations are being cumulated noon-to-noon UTC, and the number of apneas are being counted (per the “Daily Events Per Hour” chart) by session, one could imagine the following scenario:
Three apneas overnight in one session, durations 10, 11, 10 seconds each, but the last one is after noon UTC. So we see 3 apneas, 21 (not 31: the last one is for “tomorrow”) seconds total duration, average duration 7 seconds. The next night there are three apneas, durations 11, 10, 10 seconds each, but the last TWO are after noon UTC. So we see 3 apneas, 21 seconds total duration (10 from night one, and just the first apnea from night 2), for an average duration 7 seconds (with 20 seconds being carried over to night 3). Possible. But sooner or later, this “carrying” will catch up with us.
If this hypothesis is correct, it could presumably become MORE frequent as one’s location is further WEST in the US or Canada, thereby creating a greater differential between local time and UTC. Thus, while noon UTC is 8am on the US East Coast, it would be 5am on the US West Coast, thus allowing more time during which apneas are counted in one “day” but apnea duration cumulated for the next day.
Unfortunately, I don’t believe that I can confirm or rebut Ozij’s hypothesis with my own data. And with Jerry’s data, including consecutive days of single apneas, each supposedly of under 10 seconds duration, it’s hard to imagine conditions under which this explanation would work. To get under 10 seconds average duration, you need to have the machine cumulate LESS apnea time while showing MORE apnea events for the night.
Still I would ask, Jerry, would you be willing to look into the UTC times for your “six second” apneas? This would need to include the data (approximate UTC times and numbers of apneas, and daily “Time in Apnea Per Day”) including the days before and after.
My thanks to you both!! It would be astonishing if this explanation were correct; unfortunately, it would make the "Time in Apnea Per Day" information virtually useless. Other explanations or comments remain very much welcome.
Velbor