Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Thu May 21, 2009 1:14 pm
by SleeplessInNyack
That's good info, SWS; Although the audio sounds like a 33RPM played back on an auto 16-33RPM player!
They should either redo the audio or transcribe the webinars to text.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Fri May 22, 2009 4:21 pm
by Kitmastin
We also offer free CEU credits for clinicians who participate in the courses on
http://www.devilbissclinicaleducation.com site.
Thank you for the feedback on the audio recordings. We are looking at alternatives to record the audio for our webinars. In addition, we will be adding additional educational courses within the next few months.
As, always - we are listening. Feel free to offer up any comments or suggestions.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Fri May 22, 2009 7:27 pm
by SleeplessInNyack
Kitmastin wrote:We also offer free CEU credits for clinicians who participate in the courses on
http://www.devilbissclinicaleducation.com site.
Thank you for the feedback on the audio recordings. We are looking at alternatives to record the audio for our webinars. In addition, we will be adding additional educational courses within the next few months.
As, always - we are listening. Feel free to offer up any comments or suggestions.
Actually, the audio on the second Webinar was not bad, just the first one is a bit funky.
I just want to say that I think it's terrific that your making these courses available.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Fri May 22, 2009 9:36 pm
by Kitmastin
We're glad that you find the webinars useful. We are working on a fix for the audio issues and should have them repaired in the next month. If you have any other suggestions - courses you would like to see presented, etc...please let me know. You can always email me at
Kristin.Mastin@DeVilbisshc.com.
Thanks again for the feedback.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Fri May 22, 2009 9:45 pm
by -SWS
Nice!
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Fri May 22, 2009 10:49 pm
by ozij
Thank you, Kit,
On my desktop,
- the audio quality on the second was reasonable, but the sound was very weak.
Please note that there is a major time lag between the audio and the appearance of case study charts it the second webinar. That makes the last part of the presentation almost impossible to follow - a pity because it is interesting to hear the analysis.
The scoring for the second quiz is totally messed up - (check the first question and you will see that immediately).
Any chance of making those webinars dowloadable Powerpoint shows, in addition to video?
O.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Sat May 23, 2009 11:56 am
by SleeplessInNyack
Kitmastin wrote:We're glad that you find the webinars useful. We are working on a fix for the audio issues and should have them repaired in the next month. If you have any other suggestions - courses you would like to see presented, etc...please let me know. You can always email me at
Kristin.Mastin@DeVilbisshc.com.
Thanks again for the feedback.
One thing I'd like to see on the reports is the ability to zoom into an hour and see what the events are minute by minute. Also, I'd like to see the airflow line chart that's shown in the Webinars to see the details of the breathing pattern that's being analyzed. After all, now that we're learning how to differentiate an apnea from an hypopnea, it would be nice to see the raw data or line chart.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Sun May 24, 2009 8:56 pm
by ozij
Kit,
According to the webinars, Devilbiss assumes a drop in flow of more than 95% is central, and it will not respond to those apneas. Can you point us to studies supporting those assumptions?
Do you have info about the percentages of false positive and false negatives when that rule is used?
Thank you
O.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Wed Jun 03, 2009 10:01 am
by ozij
Bumping this up in the hope of getting a response to my week old question:
Kit,
According to the webinars, Devilbiss assumes a drop in flow of more than 95% is central, and it will not respond to those apneas. Can you point us to studies supporting those assumptions?
Do you have info about the percentages of false positive and false negatives when that rule is used?
Thank you
O.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Sat Mar 26, 2011 10:56 pm
by natediggidydogg
yardbird wrote:Your data looks almost exactly like mine. Last night I had ZERO apneas recorded and I think that's a first for me. However I had a LOT of hypopnea events and they wake me up so I need to track them down. I also think it's important to know what a particular setting MEANS if we're going to make changes. The way I'm reading the Advanced Menu options seems to be indicating that the percentage setting is is a figure representing how far below normal flow we must fall before registering as an event.
For example (just tossing out easy numbers, not meant to be representative of actual):
Let's say our normal flow is 100
Default Hypopnea% is 50%
I read this as: Respiratory flow must drop to 50% of normal and stay at or below 50% of normal for the time specified in Hypopnea Duration. Default for that is 10 seconds.
So respiratory flow of 50 for 10 seconds registers as a Hypopnea event.
If we change Hypopnea% to 40% and change nothing else.
Respiratory flow must fall 40% from normal for 10 seconds. This means you still HAVE 60% respiratory flow. You've made the machine more sensitive because now your respiratory flow only needs to drop 40% from normal instead of having to drop to 50% below normal.
If we change Hypopnea% to 60% and change nothing else.
Respiratory flow must fall 60% from normal for 10 seconds. This means you still HAVE 40% respiratory flow. You've made the machine less sensitive because now your respiratory flow needs to drop 60% from normal before registering an event.
So the apnea% or the hypopnea% are not percentages of what you HAVE, but rather percentages of how far you have fallen off of the normal.
A hypopnea% setting of 40% would be telling the machine that if I reach a respiratory flow that is down 40% from my normal (not 40% OF my normal), start working on this as a hypopnea event. A hypopnea setting of 50% would ignore that same scenario and wait until I've reached a 50% reduction from normal before reacting.
So LOWERING that percentage number makes the machine more sensitive, and RAISING it makes the machine less sensitive to these events.
Now here's a bigger question...
If the machine pressure is set, such that I have ZERO apnea events recorded, but lots of hypopnea events. Do I want to CHASE the hypopnea events by making these small and granular adjustments or do I do something much simpler and INCREASE my low pressure setting in an effort to keep the airway splinted open and PREVENT the hypopnea events?
BE VERY CAREFUL HERE. YOUR ABOVE CALCULATIONS ARE EXACTLY THE OPPOSITE OF WHAT THEY SHOULD BE.
Changing hypoapnea to 40% makes the machine LESS sensitive to the event.
Changing hypoapnea to 60% makes the machine MORE sensitive to the event
Here's why: look at the "apnea" options. We have 0% - 20%. That means an apnea can be defined by 0% of normal breathing volume (a true apnea) to 20% of normal breathing volume. The same rules apply to hypoapneas: setting at 60% (instead of 50%) means we are adjusting the machine so that it defines hypoapneas as 60% normal breathing volume. We are making the machine more sensitive bc it is now responding to an smaller event.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Sun Mar 27, 2011 4:13 pm
by Bobby269
The new model has a feature called smartflex
http://www.devilbisshealthcare.com/prod ... /smartflex
It should make it better for people who like assistance in breathing when using high pressure.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Mon Mar 28, 2011 2:32 pm
by DeVilbiss Marketing
I am checking with our Clinical Team for studies.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Tue Mar 29, 2011 6:22 pm
by DeVilbiss Marketing
Ozij - Dr Scharf conducted a study of the performance of our device compared to classical hand scoring. The study concluded that the AutoAdjust Algorithm accurately detected respiratory events, including "mixed" (central) events, when compared to a hand scored record during an 8.5 hour evaluation of which 237 events where scored. Manual respiratory event scoring versus device event detection scoring and event durantion correlated extremely well. A second method of statistical analysis revealed an event detection sensitivity and specificity of 86.8% and 98.4% respectively.
Unfortunately, this study is not published.
...
Posted: Wed Mar 30, 2011 12:22 am
by Evette Toepels
Hello. And Bye.
Re: IntelliPAP Auto 19, M Series Auto A-Flex 11
Posted: Sun Apr 17, 2011 3:46 pm
by marylandcpap
natediggidydogg wrote:yardbird wrote:Your data looks almost exactly like mine. Last night I had ZERO apneas recorded and I think that's a first for me. However I had a LOT of hypopnea events and they wake me up so I need to track them down. I also think it's important to know what a particular setting MEANS if we're going to make changes. The way I'm reading the Advanced Menu options seems to be indicating that the percentage setting is is a figure representing how far below normal flow we must fall before registering as an event.
For example (just tossing out easy numbers, not meant to be representative of actual):
Let's say our normal flow is 100
Default Hypopnea% is 50%
I read this as: Respiratory flow must drop to 50% of normal and stay at or below 50% of normal for the time specified in Hypopnea Duration. Default for that is 10 seconds.
So respiratory flow of 50 for 10 seconds registers as a Hypopnea event.
If we change Hypopnea% to 40% and change nothing else.
Respiratory flow must fall 40% from normal for 10 seconds. This means you still HAVE 60% respiratory flow. You've made the machine more sensitive because now your respiratory flow only needs to drop 40% from normal instead of having to drop to 50% below normal.
If we change Hypopnea% to 60% and change nothing else.
Respiratory flow must fall 60% from normal for 10 seconds. This means you still HAVE 40% respiratory flow. You've made the machine less sensitive because now your respiratory flow needs to drop 60% from normal before registering an event.
So the apnea% or the hypopnea% are not percentages of what you HAVE, but rather percentages of how far you have fallen off of the normal.
A hypopnea% setting of 40% would be telling the machine that if I reach a respiratory flow that is down 40% from my normal (not 40% OF my normal), start working on this as a hypopnea event. A hypopnea setting of 50% would ignore that same scenario and wait until I've reached a 50% reduction from normal before reacting.
So LOWERING that percentage number makes the machine more sensitive, and RAISING it makes the machine less sensitive to these events.
Now here's a bigger question...
If the machine pressure is set, such that I have ZERO apnea events recorded, but lots of hypopnea events. Do I want to CHASE the hypopnea events by making these small and granular adjustments or do I do something much simpler and INCREASE my low pressure setting in an effort to keep the airway splinted open and PREVENT the hypopnea events?
BE VERY CAREFUL HERE. YOUR ABOVE CALCULATIONS ARE EXACTLY THE OPPOSITE OF WHAT THEY SHOULD BE.
Changing hypoapnea to 40% makes the machine LESS sensitive to the event.
Changing hypoapnea to 60% makes the machine MORE sensitive to the event
Here's why: look at the "apnea" options. We have 0% - 20%. That means an apnea can be defined by 0% of normal breathing volume (a true apnea) to 20% of normal breathing volume. The same rules apply to hypoapneas: setting at 60% (instead of 50%) means we are adjusting the machine so that it defines hypoapneas as 60% normal breathing volume. We are making the machine more sensitive bc it is now responding to an smaller event.
WHO IS RIGHT HERE? CAN SOMEONE PLEASE TELL ME ?