rested gal wrote:
<snip>
Wrong. Hitting "max PS" (max press sup) does not stop IPAP from continuing to go up if need be. If the difference between Max IPAP and Min EPAP setting is more than 8 cm H2O, IPAP can continue moving up higher than the max press sup setting on the Respironics BiPAP Auto is set for. In that case, IPAP doesn't stop there. It simply drags EPAP up, too.
<snip>
RG,
I agree that the 10 shd be 8 - now if you had asked me DSM didn't you really mean 8, I would have said 'yes of course'. As I have posted here several times over the years 8 is the max.
You mention in your reply that Ipap will keep going up after it hits the end of the maxPS gap & will drag Epap with it if there is room. Now this is of interest to me as, it is what I used to think happened at first but the more I worked with this machine the more I became convinced (perhaps wrongly) that Epap will hold Ipap back rather than Ipap drag Epap up. There are to me good therapy reasons why Epap would override Ipap so I am wondering where you got the info that says Ipap will keep rising & take Epap with it ?.
Here are the links to the various manuals, Ive looked for it but wasn't able to find it - based on the unusual 'tone' of your post I am certain you must be very confident of that detail & so am sure you can point us all to it.
Thanks kindly - DSM
http://bipapautomseries.respironics.com/
http://cpaphelp.internetage.ws/manuals/ ... Manual.pdf
http://cpaphelp.internetage.ws/manuals/ ... Manual.pdf
The Bipap Auto Rules:
==================
NOTE: The settings for Min EPAP, Max IPAP, and Max PS are highly interdependent.
Consider the range of
allowable pressures as bounded at the top by the Max
IPAP setting and at the bottom by the Min EPAP setting. While setting up one
parameter, another may change according to the following set of rules:
• The device will not allow a Max PS setting outside the range of 3.0 to
8.0 cm H2O, and the device will prohibit other settings to ensure that
this rule is not violated.
• Within the range of 3.0 to 8.0 cm H2O, any setting can be chosen for
Max PS, but it is further subject to the following rule: The Max PS setting
cannot exceed Max IPAP setting minus Min EPAP setting.
• If the Min EPAP setting is incremented greater than Max IPAP setting
minus Max PS setting, it will force the Max PS to decrease. The Max PS
setting will only decrease down to the minimum setting of 3.0 cm H2O,
so Min EPAP setting cannot be incremented greater than Max IPAP
setting minus 3.0.
• If the Max IPAP setting is decremented less than Min EPAP setting plus
Max PS setting, it will force the Max PS to decrease. The Max PS setting
will only decrease down to the minimum setting of 3.0 cm H 2O, so
Max IPAP setting cannot be decremented less than Min EPAP setting
plus 3.0.