Re: My first night on ASV
Posted: Fri Nov 28, 2008 4:50 pm
By comparison to your pressure referenced, my suggestion would be 19/13/9 for comparison. Note: You do NOT want to increase EPAP to 12. You either set that to 9.0 or 10 cm and you leave it, it does NOT change. You only change EPAP when there are obstructive apnea, it is assumed when going to this machine you have already titrated for that. EPAP pressure is fixed it does NOT move on this machine at all. The "single" AP tic you are looking at on last night's report is a Central apnea not a obstructive.jskinner wrote:Seems like we are thinking along the same lines. I had already set my machine to 20/19/12 for tonight. I will go with 13 IPAPmin instead as you suggest. Will leave BPM=Auto for tonight as I don't like to change to many settings as once as it becomes hard to tell what variable change produced the results.Snoredog wrote: I still think your EPAP needs to be lower at 9.0 cm.
EPAP=9.0 cm
IPAP Min=13 cm
IPAP Max=19 cm
Question: Does the AutoVS use square wave transitions from IPAP->EPAP->IPAP or is it more rounded like their Biflex provides?
Note: The EPAP=9 setting is critical, you cannot have that too high with your phenotype, this is the intolerance of CPAP with CSDB. IF you lower EPAP=10.0 to EPAP=9.0 and it is too low, IPAP working pressure would immediately be bumping into IPAP Max ceiling, we are compensating for that happening by increasing IPAP Min=13 or 4 cm higher than EPAP pressure. I don't know how else to stress how critical that EPAP pressure is. You have showed us a history with CPAP that 9.0 cm pressure produced the lowest AHI, and that is what you want the pressure that eliminates the obstructive apnea. Also understand that increasing EPAP by 1 cm is enough to send your breathing into dysregulation. We cannot expect to have the machine correct central dysregulation because EPAP is too high, it simply won't work.