Banned wrote:
"The "recommended minimum IPAP-EPAP is 4 cmH20 and the maximum IPAP-EPAP differential is 10 cmH2O (Consensus)", refers to a recommended 4-10 cm gap between EPAP and IPAP or IPAP Min.
That recommended EPAP/IPAP/IPAP Min 4-10 cm gap has nothing to do with properly setting IPAP MAx.
Banned
Actually is does. The recommendation is for fixed bi-level but the same guidelines should apply. This is Pressure Support we are talking about here. That is the difference between EPAP and IPAP pressure.
-On the Minimum side for Bi-level they suggest a Minimum EPAP-to-IPAP of 4 cm.
-On the Maximum side for Bi-level they suggest a Maximum EPAP-to-IPAP separation of 10 cm.
This has NOTHING to do with any IPAP Min, it is NOT a 10 cm working pressure range from IPAP Min to IPAP Max. Respironics
says Set EPAP then set IPAP Max 10 cm higher. Doing so keeps the Maximum pressure suppport allowed to 10 cm and within
task force guide lines.
But the bottom line is, if you already have 10 cm allowable pressure support and you are bumping into any IPAP max consistently then EPAP needs to be increased by 1 cm so you do stay within that 10 cm pressure support maximum.
But when you look at Adapt SV therapy the only difference is IPAP is responding much faster or rapidly then what you would see on even a Bipap Auto, but it is still Bipap therapy. On this machine, that IPAP working pressure can increase much more rapidly to meet peak target values.
Hard to tell from 1 night's settings, but I'm surprised to hear that breathing against BPM=12 seemed fast for him but I suspect the avg. BPM we have been looking at in the reports has been weighted by some
rapid breathing leading to the PB seen. That did drop ever so slightly from 1% to .8% which is what I was wanting to see, certainly didn't increase, so we are going the right direction if that one night is an indicator. He probably is going to need 10 before its over. But when using fixed lowering by only 1 BPM can have a big impact on respiration as it changes the duration of the breath, my new calculations for BPM=11 are:
Pressure Settings:
===========================================================
EPAP=10.0 (he seems to like that setting)
IPAP Min=12 (I would move this to 13 cm, that is where it has been averaging anyway)
IPAP Max=20 (I'd still follow protocol and general guidelines, no sense in over inflating his lungs when it is not needed).
New BPM=Fixed settings I would try:
=============================================================
BPM=11
(60 / 11 = 5.45 secs per breath, for IT; 5.45/2=IT=2.72 seconds)
IT=2.7
(if you can dial that in, should be able to dial in 2.5)
Rise Time=3 or 4
(I would set this while awake for most comfort).
While I'm concerned about the increase seen in Machine Triggered Breaths, the rest looks pretty good and we did see
-Periodic Breathing reduced slightly
-No real change in HI seen
-Peak and Tidal volumes are up
-He reports he feels better
We have to toss out the first 2.5 hours of therapy as those settings didn't work, then you have to toss out the first 1/2 hour of concurrent sleep seen because peak and tidals seen are up with any settings used. I would try sticking to fine tuning the backup BPM mode instead of making any drastic pressure changes.
When you look at what are we trying to fix, it is the residual PB and central H's seen, both of those kind of go together in either case. We have no AP's so there are no obstructive apnea showing up, I consider that taken care of. Now we are trying to correct the PB and hopefully land at settings which produce more stable sleep across the night. I don't consider those rapid pressure swings of IPAP to be an indicator of that, allowing more of that to be seen with a higher IPAP max is the opposite of what you want to see.
Every time you make a change to EPAP you have to
start over with IPAP settings, that is why that value is so critical once you find it you set it and forget it.
If IPAP Min feels more comfortable at 12 cm then 11 then that is where I would set it. IF IPAP Min feels better at 13 or 14 that is where I would set it. Machine is going to 13 to 16 cm range when EPAP=10 anyway. But it appears BPM=Fixed has a big impact on what is seen so we don't want to limit where IPAP working pressure (SV) is thinking it should land.