Page 5 of 46

Re: Why doesn't APAP respond to apneas?

Posted: Wed Oct 08, 2008 1:20 pm
by -SWS
Snoredog wrote:
-SWS wrote: Once the algorithm determines a present Pcrit value by dropping pressure, the next part of the test is to determine a present Popt value, or "optimal pressure". That's an upward pressure probe, as the algorithm tries to find the lowest pressure that will reduce all signs of flow limitation. So this routine or periodic search for optimum pressure always entails a methodical pressure drop followed by a pressure increase.
Exactly, and that upward pressure probe happens about ever 5 minutes and can be seen on Bev's Encore Report, its the little chair step of pressure seen on her report when there are no events happening. It goes up 1 cm, then another and drops (looks like a chair on the report ever 5 minutes).

I think the probe ends or locks out after an event is seen (period of minutes, think it is 15 minutes) so during those times you won't see the pressure probes and only the baseline pressure.
You'll also note some people never get those algorithmic pressure probes, meaning that: 1) higher-priority control layers are always running in their case or 2) the low end of their prescribed/set pressure range is actually higher than that pressure threshold which would have induced a Pcrit value (induced via that extremely slight transluminal airway collapse).

Image
Pcrit's downward pressure probe (shown at left) commences the search routine.

Re: Why doesn't APAP respond to apneas?

Posted: Wed Oct 08, 2008 2:24 pm
by -SWS
Snoredog wrote:Exactly, and that upward pressure probe happens about ever 5 minutes and can be seen on Bev's Encore Report, its the little chair step of pressure seen on her report when there are no events happening. It goes up 1 cm, then another and drops (looks like a chair on the report ever 5 minutes).
Well, those atypically abundant "chairs" potentially have some interesting implications. I don't think they fit the Pcrit/Popt/Pther graph I have in my post above. The fact that those chairs: 1) start from APAP min pressure, 2) range so very high in pressure magnitude, and 3) are so frequent in occurrence compared to most Encore patient charts (with or without chairs), makes me wonder if we're seeing the algorithm's pressure response to slight ongoing periodic breathing on Bev's part.

Specifically, I'm thinking that even slight periodic breathing is going to make ongoing RemStar flow-signal baselining and the resulting Popt-based search for Pther a fairly dicey proposition. I think that may very well be what we are seeing with those fairly regular pressure-response "chairs". If so, that makes the BiPAP autoSV an even more compelling platform for Bev to try next IMO.

Man, I sure could be wrong about those chairs...

Re: Why doesn't APAP respond to apneas?

Posted: Wed Oct 08, 2008 5:17 pm
by Snoredog
-SWS wrote:
Snoredog wrote:Exactly, and that upward pressure probe happens about ever 5 minutes and can be seen on Bev's Encore Report, its the little chair step of pressure seen on her report when there are no events happening. It goes up 1 cm, then another and drops (looks like a chair on the report ever 5 minutes).
Well, those atypically abundant "chairs" potentially have some interesting implications. I don't think they fit the Pcrit/Popt/Pther graph I have in my post above. The fact that those chairs: 1) start from APAP min pressure, 2) range so very high in pressure magnitude, and 3) are so frequent in occurrence compared to most Encore patient charts (with or without chairs), makes me wonder if we're seeing the algorithm's pressure response to slight ongoing periodic breathing on Bev's part.

Specifically, I'm thinking that even slight periodic breathing is going to make ongoing RemStar flow-signal baselining and the resulting Popt-based search for Pther a fairly dicey proposition. I think that may very well be what we are seeing with those fairly regular pressure-response "chairs". If so, that makes the BiPAP autoSV an even more compelling platform for Bev to try next IMO.

Man, I sure could be wrong about those chairs...
It's in one of the manuals, think I have it on one of my older laptops. I don't have it on this Mac. But it also explains why it goes above the Maximum pressure setting set too. But I'll correct the manual next time I find it

The Remstar Auto's have always done this even the non-flex classic. If you look at Bev's reports, what else is it responding to? there is no snore, there is no flow limitation. Basically her 14 cm Minimum is taking care of most events, it is even avoiding those rogue apnea (ones I suspect are actually central being scored as obstructive). The only reason they are not showing up as NRAH is it takes like 6 of them to trip that. So when there are no events taking priority it gets bored and says put a chair to the left, put a chair to the left, where do you think Beyonce got the lyrics to that hit song from?

I'm looking in the US patent docs also, I think it is in there too. You have to realize, she is at 14 cm pressure. The command on apnea setting for Remstar is 11 cm (paragraph 153 in the patent). It starts at 8 cm and goes to 11 cm where it basically relies on the NRAH from that point on. In Bev's case, she is above the 11 cm threshold with Minimum pressure, that means once the A/H controller is released and the variable breathing controller is released (i.e. not sensing variable breathing) it goes into that probe monitor mode. Basically it is increasing pressure by 1 cm and holding for 30 seconds where it increases again switching to monitor mode looking for the distinct markings of a flow limitation since Respironics think that most apnea are preceded by one. Give me some time, I'll find the exact explanation of it. I don't think you'll find it in that simulation program but it might be there. I could see what you suggest happening if she was above the Minimum pressure, but I don't think it will drop pressure below the Minimum when probing.

Re: Why doesn't APAP respond to apneas?

Posted: Wed Oct 08, 2008 6:34 pm
by deerslayer
also am hoping you start to get more restful sleep Bev. i also don't understand why night before i had an (ahi of 0.4) had 4 events w/20 sec. in apnea for 8 hrs., then last night spent over 2 min. in apnea w/( 2.9 ahi) w/over 3 min in apnea w/ 14 sec duration. same normal leak data. i feel like i had plenty of energy today but the #'s are confusing. maybe when i go in for sleep eval in nov. they will explain-- right, like that will happen.

Re: Why doesn't APAP respond to apneas?

Posted: Wed Oct 08, 2008 7:00 pm
by OutaSync
Wow! I get home from work and look what I missed!

SWS, I have two sleep studies, two months apart. What would you like to see? All the pages? I can scan them in tomorrow whenI get back to the office. Funny thing is my first Sleep study doesn't show any centrals and I never got any Stage 4 or REM. I was given AMbien before both sleep studies.(Thanks, SAG)

Bev

Re: Why doesn't APAP respond to apneas?

Posted: Wed Oct 08, 2008 7:20 pm
by -SWS
Bev, if you don't mind scanning them all in I think that would make for a great discussion. Anybody who knows me well knows that I'm a good-natured opportunist. Sneaky me: I'm thinking that SAG just may spot something/anything. Just about every time SAG looks at sleep charts and comments we all tend to learn. I certainly do.
OutaSync wrote:Funny thing is my first Sleep study doesn't show any centrals and I never got any Stage 4 or REM.
Was that first study without central apneas a split-night study? If so did you manage to sleep on xPAP during the second part of that night?

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 5:12 am
by OutaSync
Yes, my first study was split , I slept all night, barely waking when they came in to put the mask on, and when they woke me up at 6 am, I stumbled to my car, drove straight home and slept until 2pm. My first few months on Cpap, and then Apap were a blur of exhaustion. All I could do was come to this board and read. I don't know how I kept my job.
I have slept through one night this year without remembering waking up. At least now, I can fall back asleep. Before CPAP, I had trouble getting to sleep and staying asleep.

Bev

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 6:41 pm
by -SWS
ozij wrote:
While in Ptherapy mode, if flow limitation occurs during the last 4 breaths or over several minutes, the algorithm begins a Popt search.
My emphasis.

I understand this to mean that after 5 minutes the algrithm actively analyses the data for indications of small degrees for flow limitation.
Ozij, I think your interpretation above is correct. But I think that routine you have highlighted happens on a non-scheduled detection basis---regarding the spontaneous detection of slight FL precursors. I also agree with Snoredog that a scheduled or routine Pcrit/Popt/Pther search happens as well. However, that latter scheduled or routine search is for the purpose of finding the lowest feasible pressure---and it starts with an intentional downward pressure probe to ascertain Pcrit's value. Then it probes upward. The upward pressure search you have highlighted is for the purpose of reacting to FL precursors. In summary, one search wants to see just how low the machine can safely take the patient---and routinely attempts just that by first looking downward for Pcrit. The other extremely similar search is not scheduled, and that pressure-response routine wants to escape spontaneous FL precursors by immediately going to higher pressures in search of Popt. In this case a spontaneously generated Pcrit condition is presented; that spontaneous Pcrit condition obviates the need for the algorithm to first probe for a Pcrit value. The algorithm goes straight to an upward Popt search in this case. Of course, that's only my interpretation.

However, what Ozij has highlighted at the top of my post is what I think is probably occurring in Bev's case. Specifically, I think that subtle preliminary indicators of FL likely occur spontaneously---and that precursor presence triggers this particular upward-pressure sequence. These upward probe-triggering flow signal events are not frank FL events. They are some insufficient subset of the individual FL probability criteria. As an example, slight amplitude reduction is one of the FL probability-based scoring parameters that Respironics tracks. It's one of several.

Let's pretend that Bev shows only slight or marginal periodic breathing. In fact, that flow-amplitude up and down action would be so slight, that it wouldn't even be noticed or scored as periodic breathing in the PSG. Recall CompSA/CSDB was just classified in a 2005 Harvard study. Right? And at that point only full-blown respiratory-controller oscillations were delineated. Since that point in time, slight and moderate CSDB/CompSA diagnostic criteria have been developed, albeit not yet fully explored or delineated.

Anyway, let's say that Bev experiences only slight respiratory-controller oscillation. Fairly steady at times, fairly periodic, but ever so slight regarding the ups and downs of amplitude. Not dramatic enough to score as frank periodic breathing (thus not warranting treatment by Respironics' Variable Breathing part of the algorithm). And not an adequately sustained amplitude reduction that might score as one or more hypopneas. But just wobbly enough for the detection algorithm to think that it sees the makings of a slightly collapsing airway (as opposed to a slightly wobbly respiratory controller). And that's pretty much the only precursor sign of FL that the algorithm sees in Bev's case.

So the algorithm doesn't score FL, for lack of all the other probability components of frank FL scoring criteria. But it decides that it needs to pressure probe upward until that flow wobble goes away. And by, golly, pressure goes up and up quite a ways before that flow-amplitude wobble spontaneously subsides.

In general it would be possible for the algorithm to commence that same upward search routine for any subset of probability-based FL precursors---not just the amplitude parameter. That means slight obstruction or transluminal airway collapse (as opposed to respiratory controller oscillation) is a more likely case across the obstructive patient population. Regardless, Bev's not at all out of the running for a slight respiratory-controller issue with all the biologic discomfort and characteristic cyclic alternating pattern (CAP) that goes with slight CompSA/CSDB in my opinion.

Science is still trying to get a handle on the various complex SDB presentations of one very challenging phenotype IMO.



-----------------------------------------------------------------------------------------


Bev, have you been evaluated for the likes of connective tissue and autoimmune disorders---such as fibromyalgia and chronic fatigue syndrome?

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 9:11 pm
by dsm
Bev,

I know you will get a chuckle from this but here is a blow-your-socks off home ventilator for only $600 includes 6 month warranty - SWS & I will talk ya thru it (SWS ? )
http://www.dotmed.com/listing/550718?mo ... l&cond=all

New the PLV-100 cost over $9,000 (but on discount just under $8,000)
http://www.medexsupply.com/respiratory- ... id-86.html

This machine can pressure ventilate you, volume ventilate you, flow ventilate you, tell you on the dial what it is doing to you & why, will call a nurse (see nurse alarms ) I suspect it will even call ET back on his home planet. It has alarms for everything & a built in backup battery in case you kick the wall plug out or the dog chews the cable

To cap it all off it only goes to a max of 100 CMs now that is ventilation

DSM

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 9:24 pm
by -SWS


Those are volume ventilators... from volume dealerships?

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 9:29 pm
by dsm
-SWS wrote::lol:

Those are volume ventilators... from volume dealerships?
All joking aside, had I been in the US I would grab it just to look inside
hours of fun & experimentation - more fun than a Rolling Stones concert &
probably cheaper

DSM

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 9:35 pm
by dsm
SWS
That unit is more along the lines of what Laryssa needs - pediatric capable.
I guess the problem would be getting it into Brazil.

DSM

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 9:47 pm
by -SWS
That's exactly what I was thinking, Doug... the part about Laryssa. My wife and I always think about her. A week doesn't go by that we don't have at least some conversation about her.

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 10:05 pm
by ozij
-SWS wrote:
ozij wrote:
While in Ptherapy mode, if flow limitation occurs during the last 4 breaths or over several minutes, the algorithm begins a Popt search.
My emphasis.

I understand this to mean that after 5 minutes the algrithm actively analyses the data for indications of small degrees for flow limitation.
Ozij, I think your interpretation above is correct. But I think that routine you have highlighted happens on a non-scheduled detection basis---regarding the spontaneous detection of slight FL precursors. I also agree with Snoredog that a scheduled or routine Pcrit/Popt/Pther search happens as well. However, that latter scheduled or routine search is for the purpose of finding the lowest feasible pressure---and it starts with an intentional downward pressure probe to ascertain Pcrit's value. Then it probes upward. The upward pressure search you have highlighted is for the purpose of reacting to FL precursors. In summary, one search wants to see just how low the machine can safely take the patient---and routinely attempts just that by first looking downward for Pcrit.
Do you assume a search for Pcrit to start at minimum pressure? I wouldn't.
The other extremely similar search is not scheduled, and that pressure-response routine wants to escape spontaneous FL precursors by immediately going to higher pressures in search of Popt. In this case a spontaneously generated Pcrit condition is presented; that spontaneous Pcrit condition obviates the need for the algorithm to first probe for a Pcrit value. The algorithm goes straight to an upward Popt search in this case. Of course, that's only my interpretation.
That's exactly my interpretation of it as well.
However, what Ozij has highlighted at the top of my post is what I think is probably occurring in Bev's case. Specifically, I think that subtle preliminary indicators of FL likely occur spontaneously---and that precursor presence triggers this particular upward-pressure sequence. These upward probe-triggering flow signal events are not frank FL events. They are some insufficient subset of the individual FL probability criteria. As an example, slight amplitude reduction is one of the FL probability-based scoring parameters that Respironics tracks. It's one of several.
Agreed.
So the algorithm doesn't score FL, for lack of all the other probability components of frank FL scoring criteria. But it decides that it needs to pressure probe upward until that flow wobble goes away. And by, golly, pressure goes up and up quite a ways before that flow-amplitude wobble spontaneously subsides. In general it would be possible for the algorithm to commence that same upward search routine for any subset of probability-based FL precursors---not just the amplitude parameter. That means slight obstruction or transluminal airway collapse (as opposed to respiratory controller oscillation) is a more likely case across the obstructive patient population.
Once again, agreed, and this is what I tried to say about the algorithm's response in my posts from Wednesday.

I'm curious to see RG's charts on a RemStar with new firmware. Remember she's one of the people who need to turn off the 420E's response to flow limitation. And the 420E is tagging for more breaths as flow limited than the RemStar. Assuming it's those subliminal (transluminal???) airway collapses that drive her 420E with IFL1=on up and away - I'd expect to find those pressure chairs on a RemStar when she's just a little bit below the minimum she should be at.

Or maybe pressure stools.... without that little upward jog creating the chair's back.

O.

IFL1 = For those who don't know, the 420E's switches controlling its response to Inspiratory Flow Limitations

Re: Why doesn't APAP respond to apneas?

Posted: Thu Oct 09, 2008 10:47 pm
by rested gal
ozij wrote:I'd expect to find those pressure chairs on a RemStar when she's just a little bit below the minimum she should be at.
You'd be right. Enough little chairs on my data to outfit a banquet hall.

It's always been that way for me when I used any Respironics auto including the M auto with A-flex. Fortunately those little up/down pressure changes never seemed to disturb my sleep. Or maybe I got used to them. Or...maybe they do (disturb sleep) and I just don't know it. Anyone got a spare EEG?