Re: Why doesn't APAP respond to apneas?

Posted: Sun Oct 19, 2008 10:56 pm
by rested gal
Songbird wrote:NOTICE: Please file the following thought, which has occurred to me several times during this particular discussion, under HAVING fun, not MAKING fun; GOOD-HEARTED TEASING, not COMPLAINT; the kind of fun you'll only find in a tight-knit group of highly cherished people, be it family or friends: https://www.youtube.com/watch?v=1_47KVJV8DU

Carry on, boys.
Marsha
Perfect!!!!!!!! Perfect song at the perfect time, Marsha.

What's got me laughing so much (besides the song) is that just one minute after you posted, a "carry on" did happen!!

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 2:00 am
by dsm
Test results
Image

I need to come up with a better test approach as the results were erratic & changed
with different settings in the early tests.

But when I finally set the machine with epap=8, ipapMin=12 & IpapMax=30 & BPM=10 it
started to behave as predicted. With this setup the 1st ipap cycle the machine went
to 12 CMs from 8 CMs a few times then it went (in a short burst) 2 CMs higher (to 14
CMs). Each new Ipap cycle It boosted by 2 CMs delivering a 'puff' of air until the ipap
reached approx 10 CMs above ipapMin (22-23 CMs). Upon reaching 21-23 all future ipap
cycles remained at 21-23 CMs.

The conclusions I reached are that if there is zero flow, the machine will cycle the
epap-ipap as any timed bilevel will but the SV on each successive ipap cycle adds
2 CMs pressure & does so as a 'puff' of air. The increases appear to then top out at
approx 10 CMs above IpapMin.

Streaming Video (it took a short while to stop the leaks from the mask, you can hear
this). Epap=8, ipap=12, ipapMax=30 and BPM=10 (with INSP=1.7).

(high quality streaming video)
http://www.internetage.ws/cpapdata/dsm- ... papsv2.wmv
DSM

Fixing A Watch With A Chainsaw...

Posted: Mon Oct 20, 2008 5:20 am
by StillAnotherGuest
-SWS wrote:Re: SAG's rhetorical question about events seen at BPM rates > 6... That > 6 time frame gets into the 10 s definition of hypopnea. But if you don't see it because it's been ventilated, then theoretically you missed seeing it because it's presumed to be a successfully treated hypopnea.
Right. And without seeing arousals or desaturations, it can only be "presumed" that they are successfully treated vs simply resetting the timer so the event doesn't register (although admittedly, I don't know how that reporting mechanism works. Does one breath negate the respiratory event? Two? 8,420?

But with that thing pegged at IPAPmax for about 5.5 hours out of a total about 9.5 hours (I don't think it's completely pegged, short-cycle events should have at least some IPAPmin time. I mean, the thing can't establish a baseline at IPAPmax, so it certainly seems real busy attacking something or other). However, this can't be determined due to the coarseness of the reporting mechanism. And why it is essential to qualify the underlying disease process.

Since Younes demonstrated that you can create "central dysregulation" in almost anybody

Susceptibility to periodic breathing with assisted ventilation during sleep in normal subjects

it is now impossible to determine whether this entire approach is correcting or creating "central dysregulation".

As previously noted, this machine will "automatically" address "central dysregulation", but obstructive events need to be manually addressed. And if events cannot be accurately identified, how do know (1) when you've fixed them; or (2) if they're even there?

So again, if it is obstructive disease, then you need an obstructive algorithm (APAP or BiPAP Auto). If it's central disease, then you can use the central algorithm (AdaptSV or BiPAP AutoSV). If you've created central disease with a central algorithm (which I think will be nearly impossible to determine given the limited amount of tools) and/or are trying to fix Wake/Stage 1 transition, then there's gonna be an awful lot of Manual adjustments on a machine that is being touted as Automatic.

SAG

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 5:24 am
by OutaSync
Report from last night. Gotta go to work now, I'll try to check in from there, but real busy today. Still have a terrible headache.
Image

Bev

Re: Are You Seeing What You're Seeing?

Posted: Mon Oct 20, 2008 6:44 am
by jnk
StillAnotherGuest wrote:. . . any symptoms of narcolepsy . . . ? SAG
Page 8 of this discussion:
OutaSync on Sun Oct 12, 2008 12:24 pm wrote:. . . I used to wake up a lot thinking someone was in the room because I had woken to talking or animal like noises. It was just me. . . .
jeff (only trying to be helpful this time )

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 7:43 am
by OutaSync
SAG,

No symptoms of nacolepsy. Sometimes in the morning before I wake up, I feel as though I'm lying there awake and cannot move. I can see everything in my room but can't move. Then I wake up. That hasn't happened since I've been on CPAP. Don't know if that means anything.

Bev

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 7:43 am
by -SWS
StillAnotherGuest wrote:But with that thing pegged at IPAPmax for about 5.5 hours out of a total about 9.5 hours (I don't think it's completely pegged, short-cycle events should have at least some IPAPmin time. I mean, the thing can't establish a baseline at IPAPmax, so it certainly seems real busy attacking something or other). However, this can't be determined due to the coarseness of the reporting mechanism. And why it is essential to qualify the underlying disease process.

Since Younes demonstrated that you can create "central dysregulation" in almost anybody

Susceptibility to periodic breathing with assisted ventilation during sleep in normal subjects

it is now impossible to determine whether this entire approach is correcting or creating "central dysregulation".
SAG, I agree with this assessment. We can see what is either outstanding or induced central dysregulation in Bev's Encore charts. Encore's scoring of PB is the strongest clue. The other revealing Encore data trace is the IPAP peak line. It shows the algorithm being very busy, sending IPAP peak all the way up to IPAP max for grotesquely extended periods. When SV successfully stabilizes periodic breathing, on an as-needed basis, there is absolutely no algorithmic basis to hold that IPAP max value for extended periods as we see on Bev's charts.

I'm also inclined to think that this machine's adaptive pressure response may be creating more problems for Bev than it is solving. Younes, et al cite that tendency in PAV, short of any temporal analysis the BiPAP autoSV algorithm may employ that experimental PAV circa 1998 did not employ. Despite the fact that the earlier implementation of PAV studied by Younes et al was not initially designed to address PB for lack of temporal variable analysis---and the fact that the Respironics implementation of autoSV claims to succesfully treat PB----there is absolutely no doubt in my mind whatsoever that the Respironics (and even Resmed) implementations of auto/adaptive SV can and will induce central dysregulation in some patients for lack of necessary algorithmic and especially temporal-variable based controls. Unfortunately today's "automated" SDB technology is far from perfect---far from being any kind of SDB treatment panacea. Here, SV technology is extremely shortsighted in that it sorely lacks temporal variable analysis IMO.

Regarding Bev's tendency toward periodic breathing. My strong personal hunch is that Bev, like so many other patients, probably has a tendency toward (stimuli) "responsive" central dysregulation. I think those are likely the patients who, ironically, have an inordinately difficult time with SV adaptation during the wake/stage-1 transition----based on highly dynamic adaptive/maladaptive receptor sensitivities that might be theoretically inherent to certain autonomic defensive characteristics (but one possible genetic phenotype entailing viable evolutionary survival value). Any patient entailing highly characteristic/deterministic respiratory controller loop/gain dyscontrol should be much better suited for auto/adaptive SV in my own way of analyzing. Conversely, any patient with neurologically defensive adaptive/maladaptive receptor sensitivities are probably going to be extremely difficult for any temporally short-sighted auto/adaptive SV algorithm to resolve. Those chemoreceptor sensitivities are theoretically going to be a moving target. These latter patients may be the ones having an inordinately difficult time making that wake/stage-1 transition. Then off to high-strung and defensive autonomic breathing they eventually go, where they still may have a difficult time defensively and adaptively attempting to survive that highly dynamic stimuli presented by SV pressure therapy.

If this neurologically defensive genetic phenotype even exists, my hunch is that central dysregulation can be methodologically induced by presenting nothing more than external sensory stimuli---much more so than in any control set.
SAG wrote:As previously noted, this machine will "automatically" address "central dysregulation", but obstructive events need to be manually addressed. And if events cannot be accurately identified, how do know (1) when you've fixed them; or (2) if they're even there?
Another extremely valid point IMO. As always, thanks for your professionally seasoned and very wise input, SAG.

-----------------------------------------------------------------------------------

Bev, I'm still of the highly-speculative impression that you're a "tightly strung" or "neurologically defensive" sleeper. I think any set of disruptive stimuli is inclined to devastate your sleep architecture. That means GERD, pain, aerophagia, unfamiliar environmental input (new surroundings, new sound, etc.), and now the pressure pulsing of auto SV can account for deteriorated sleep and even defensively-dysregulated breathing at times.

If auto SV's highly dynamic pressure response is even capable of offsetting bifurcated central dysregulation in your hypothetical case, it's only going to be because you have come to consciously and subconsciously associate that same pressure stimuli as being familiar and friendly rather than being potentially threatening. And in my own view it's highly questionable whether any human can carry that adaptive association into sleep and autonomic breathing. I strongly suspect that may be an impossibility.

I also recommend finding a different sleep doctor in the upcoming insurance cycle---one who is open to the idea that your daytime cognitive issues and EDS are problems that need to be solved. Short of that I would suggest trying fixed BiLevel. There is a chance that fixed BiLevel may be a suitable modality for you. And if it turns out that you eventually become extremely comfortable with fixed BiLevel, then the auto SV algorithm may be less-inclined to incessantly push at IPAP max as it currently does.

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 7:54 am
by Snoredog
Thanks SAG, your journal paper answered my question on Periodic Breathing,
In the absence of a mandatory back-up rate, neither method is capable of producing sustained arrest of respiratory efforts; if ventilation increases excessively and PCO2 decreases below the apnea point, no triggering and, hence, no ventilation occur until PCO2 rises again enough to reestablish spontaneous efforts. The result of excessive assist would, therefore, be periodic breathing (PB). It has been reported that some normal subjects develop PB during sleep while receiving PSV (18).
SWS: Maybe what we are seeing is the same train wreck that was seen on her Bilevel titration at 12/8 pressure, it also has the lowest desaturations seen on that PSG.

Believe that was the 200712006-1.jpg image Bev posted.

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 8:20 am
by Snoredog
OutaSync wrote:SAG,

No symptoms of nacolepsy. Sometimes in the morning before I wake up, I feel as though I'm lying there awake and cannot move. I can see everything in my room but can't move. Then I wake up. That hasn't happened since I've been on CPAP. Don't know if that means anything.

Bev
That is called Sleep Paralysis, thought associated with early stages of narcolepsy. Has your doc ever tried you on Xyrem?

It is a controlled substance but sometimes given for more energy during the day so you sleep better at night.

For the Sleep Paralysis, sometimes understanding what it is (neuro switch hasn't turned full muscle control back on yet and you became conscious or awake first). The fix is fall back asleep and next time you awaken you should awaken normally.

Unless SWS makes a suggestion, I would go back on the Aflex at prior settings for a night or two so you can catch up on your sleep.

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 8:24 am
by -SWS
Snoredog wrote: Unless SWS makes a suggestion, I would go back on the Aflex at prior settings for a night or two so you can catch up on your sleep.
Catch up on sleep with the old familiar machine. Absolutely. Then entertain whether you would like to try fixed BiLevel modality some weekend.

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 8:44 am
by Snoredog
-SWS wrote:
Snoredog wrote: Unless SWS makes a suggestion, I would go back on the Aflex at prior settings for a night or two so you can catch up on your sleep.
Catch up on sleep with the old familiar machine. Absolutely. Then entertain whether you would like to try fixed BiLevel modality some weekend.
Out of what I've seen so far;

1. No mention of Periodic Breathing on all of her PSG's, so I'm inclined to agree with what SAG said.
2. I really don't like the PB seen or the "agressiveness" seen with current settings.
3. While attempting the SV modality, only 1 cm increase made things fall apart in a hurry, IMO.
4. I see last night's report actually worse and I don't like her having headaches, that means desaturation to me and this was what I think DSM was also thinking with his comment.

Me thinks we need to think this out more before Bilevel. Reason I say that is we seen her bilevel titrations and the hope was with the SV having the pressure support that it may change things. Currently, we are still CPAP +SV. And well she does better on CPAP from what I have seen to date.

IS the machine overly responsive because of backup mode being Auto? Of course it is Bev's decision, I would be in favor of supporting the following settings before tossing in the towel on the SV:

EPAP=14
IPAP Min=14
IPAP Max=18

Fixed backup BPM=10 and IT=1.2.

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 8:56 am
by jnk
. . . Sleep Paralysis, thought associated with early stages of narcolepsy. . . .
Yes. Associated. Although it can occur occasionally in anyone, and especially when a person is sleep-deprived. Same with the sensation of hearing noises or seeming to feel a presence in the room when waking up. Non-narcoleptics can experience those too. Still such details can be worth noting.

(This post is directed at the many following this thread.)

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 9:15 am
by Snoredog
jnk wrote:
. . . Sleep Paralysis, thought associated with early stages of narcolepsy. . . .
Yes. Associated. Although it can occur occasionally in anyone, and especially when a person is sleep-deprived. Same with the sensation of hearing noises or seeming to feel a presence in the room when waking up. Non-narcoleptics can experience those too. Still such details can be worth noting.

(This post is directed at the many following this thread.)
Exactly, severe sleep deprivation can bring it on also.

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 9:19 am
by -SWS
Snoredog, CPAP + SV would entail IPAP min also being set at EPAP's value.

If it were me I would also eventually go back to trialing SV---but only after learning how to sleep and feel extremely comfortably with the small, fixed pressure cycling of ordinary BiLevel (thus with IPAP min and IPAP max both set only 3 cm above EPAP). That gets away from wondering which events might be SV induced. It also gives Bev a chance to find the best fixed IPAP and EPAP values that address her obstructive components.

Then once I was very comfortable with that fixed BiLevel pressure cycling, I would re-trial SV---first with auto BPM, and then with manual backup and IT settings. I would use the above determined EPAP as SV's EPAP. And I would use the above determined IPAP as SV's new IPAP min. I would experimentally set IPAP max at no higher than 10 cm above the EPAP value. Again, I'd give auto BPM another shot before heading off to manual BPM and IT settings. I'd at least initially employ whatever rise time I found most comfortable while running at fixed BiLevel.

Absolutely positively time to catch up on sleep with the old APAP machine first IMO.

Never did find out what rise time Bev is using.

Re: Why doesn't APAP respond to apneas?

Posted: Mon Oct 20, 2008 9:34 am
by rested gal
-SWS wrote:I also recommend finding a different sleep doctor in the upcoming insurance cycle---one who is open to the idea that your daytime cognitive issues and EDS are problems that need to be solved. Short of that I would suggest trying fixed BiLevel. There is a chance that fixed BiLevel may be a suitable modality for you.
I think that would be the best thing to try, if another doctor/sleep study is out of the question.
-SWS wrote:I would suggest initially turning SV completely off while you first learn to sleep with ordinary BiLevel. When it comes to BiLevel, some people may conceivably need to learn how to crawl (with low constant IPAP pressures) before they learn to run (with wildly or dynamically fluctuating IPAP pressures). If you would like to spend a few nights learning how to sleep with basic BiLevel before attempting to try SV modality, then just set IPAP max as low as you have your IPAP min. When IPAP max equals IPAP min, you have SV turned off and you are running at ordinary Bilevel. See if you can learn to sleep with simple BiLevel first, collecting important baseline data along the way.
-SWS wrote: viewtopic.php?f=1&t=35298&start=120#p304114

Bev very simply stated that 14 to 17 yielded her best AHI.
Yes, and since Bev was speaking of her autopap results in that thread, perhaps BiPAP with EPAP 13, IPAP 17, Biflex at 2 or 3 might be a good way to see if things could settle down with some time spent on "just bilevel". The BIPAP Auto SV doesn't have bi-flex however, so...I've had a thought

If Bev could get hold of a bilevel machine, and I think she can....