-SWS wrote: Respironics clearly refers to their "SV" mode as "automatic pressure support". They clearly allow for that "automatic PS" to be applied not only to BiPAP mode (they call it "BiPAP + PS") but they also intend it to be applied to CPAP mode should etiology require (they call it "CPAP + PS"). And I explained that etiology would be a CSDB/CompSA patient who tends to destabilize less with CPAP modality than BiLevel modality.
No one is disputing
Pressure Support, it is not even a consideration from my point of view at this stage. I thought you asked for suggested CPAP and BIPAP pressure(s) to transfer over to the SV and for discussion. Out of all the data I seen 9 cm was the value I would select to start with for EPAP. I based that on her PSG result, not Encore. Past Encore reports are in the toilet for setting up the SV in my opinion.
My suggestions for Initial Settings for Bev's SV in SV mode, I see no reason to set it up in any other mode:
1. When setting up the SV, you need to first provide the CPAP/EPAP pressure which eliminates obstructive apnea. After looking at her PSG titrations, Encore reports of which I discount the latter in providing the correct information, I resort back to the PSG titration data and since I want to start lower than current, that is 9 cm found on her 10/16/07 PSG. Zero events, it is the highest pressure seen before the train wrecks.
EPAP=9.0 cm
2. Next, I want to know if that EPAP of 9 is going to take care of her OA's, so IPAP doesn't influence that finding, I set IPAP Min to the same or 1 cm higher than EPAP maintaining a 1 cm delta (Respironics recommendation from their titration guide).
IPAP Min=10 cm
3. Next, I want to give the machine a broad range to work during this initial setup period, I don't want to hinder its auto operation at all, so I use the recommended 10 cm Pressure Support spread, that is 10 cm. So as a result, I set IPAP Max to:
IPAP Max=19 cm (EPAP +10 cm)
4. At this point I don't know what her Respiratory Rate or BPM is. None of her prior studies gave me that info, so I need to set the machine in Auto mode and let it find that automatically. SO BPM=Auto.
BPM=Auto
5. Backup mode. If the Auto mode fails, I need to set up some fail-safe parameters until I have the spontaneous data, so backup mode is set to default BPM=10, IT=1.2 sec, Rise Time=2 or 3. Rise time is a comfort setting.
The Auto SV is going to have full control over IPAP. If it needs to add pressure support, working IPAP will rise off the Minimum setting to what ever it needs. It may even rise to 19 cm and bump into the Maximum. Now if she has a central apnea or periodic breathing while in that Auto SV mode it will switch to the backup mode where BPM will be 10 and Inspiration Time will be 1.2 seconds. That should stop any CA where once breathing is stabilized she will cycle back to spontaneous mode. If the machine is having to take over with the Timed backup values, that means the static settings are not set correctly in Spontaneous mode. If reports shows OA higher than wanted, EPAP gets increased. If I see zero OA on her report, I may even drop EPAP by 1 cm. I would hope to see 2 to 6 OA from a session. I want that EPAP pressure just where it eliminates the OA's and no more. Next, I expect the AutoSV mode to completely eliminate all Hypopnea. I'm wanting to see 99% or higher of User initiated breathing. Because if the machine is reporting that high a User Initiated breathing I know she isn't having any centrals or periodic breathing. Once I know what her avg. BPM is in the Spontaneous mode I input that value into the backup BPM field minus 2. You don't want it the same as the Spontaneous mode, you want it slower so it causes slight discomfort where the patient is encouraged to breathe on their own. The machine will automatically predict target volume half way through inspiration, if it doesn't see the target volume being met it will increase and extend inspiration (usually in 2 to 6 breaths).
So starting off I'm giving the SV a nice big range to work in. Once parameters are known values can be fine tuned for best therapy. You don't have to worry about Pressure Support at all. I'd want to see how she did after the initial settings before I'd go jacking her up to those higher pressures, she already has a aerophagia problem.
I also have the clinical and titration guide if you need it Bev let me know, it has easy to understand pictures and yes/no decision trees, I promise it is a whole lot easier to understand than this discussion.
FYI for SWS: You need to update your Marketing data dude, it goes to 30 cm, no wait 25 cm, see below, as pasted right out of their PDF guide
Some definitions
EPAP End Expiratory Positive Airway Pressure
IPAPmin Minimum Inspiratory Pressure
IPAPmax Maximum Inspiratory Pressure
BPM Back-Up Rate
Ti Inspiratory Time
Settings
EPAP: 4 to 25 cm H2O
- Should be adjusted to treat Obstructive
Component
IPAPmin: EPAP to 30 cm H2O
- Equal to EPAP for CPAP to treat OSA
- Add 2–3 cm above EPAP for comfort
IPAPmax: IPAPmin to 30 cm H2O
RATE - OFF
- Auto
- 4-30 bpm
DEFAULT SETTINGS - EPAP: 5 cm H2O
- IPAPmin: 5 cm H2O
- IPAPmax: 15 cm H2O
- RATE: Auto
Set to treat Obstructive
Component
- EPAP = IPAPmin: CPAP
- EPAP < IPAPmin: BiPAP®
Automatic Pressure
Support Maximum
Allows for clinician adjustments
to back-up rate
SETTINGS