Page 2 of 3

Posted: Wed Aug 06, 2008 5:39 pm
by BigEd
Snoredog wrote:
From your ID name, I assume you are a large man? What is your neck size? You have a large tongue? Have an overbite or receding chin? Have you always had difficulty with CPAP therapy? Are you sleeping on your back in the supine position or on your side?
Thanks for taking the time to look at my situation. I'll do my best to answer your questions. Yes I'm a large man. Neck size is 19.5. I don't have a large tongue or overbite or receding chin. Been on CPAP for just over 2 years. No major problems with the therapy, other than occasional large leak (mouth breathing). I loved using the Swift Nasal pillows - just had too many issues recently with mouth breathing so I started playing with trying to solve the problem. I do sleep in the supine position.
Snoredog wrote: Did this just show up when switching to the Swift mask? Personally, if it is what I suspect I don't think that interface will work for you, it cannot deliver the volume of air you need, reason for all the leaks, your nares may not be suited well for that type of interface and that could be another reason for all the leaks. If your nares are "slotted" that interface isn't going to work.
The above reports were created while on the UMFF. I've added a report below (from a week or so ago) while on the nasal pillows. The report below is from one of my "better" nights while taping or using chin-up strips. (I don't recall now.) My nares are not slotted.
Snoredog wrote: What was your pressure when you were on CPAP?
I was only on CPAP for about 10 days after my initial sleep study - and that pressure was 20. Very difficult with nasal pillows. That's why I was able to swap my machine to the current BIPAP Auto.
Snoredog wrote: That is the pressure you should start your Bipap EPAP at even in Auto mode. I would set PS=4. Only PS that appears to be getting used now is the built-in Minimum Pressure Support by the machine which is 2 cm. At therapy hour 6.25 where you apparently had a break there is some separation of IPAP/EPAP what looks to be 3 cm?? Do you have PS at default of 3? If so, you need to increase that to 4 or 5, I wouldn't go over 6 at this point.
My current settings are:

Machine is set in Auto Bi-Level with Bi-Flex mode - ABFLE
Biflex=3
IPAP Max=25
EPAP Min=8
PS=8
Snoredog wrote:Have you been to an ENT for an exam?
No I have not.

Are you suggesting I switch back to the nasal pillows? That would be fine with me if I can control the mouth breathing. Then I could keep the FF for nights when congested.

Hopefully with this additional information you can make a more informed suggestion as to what I should do next.

Thanks again. Here's that report from a week or so ago, while on the Mirage II nasal pillows:
Image


Posted: Wed Aug 06, 2008 9:24 pm
by ozij
Havnin NR show on your report, simply means the apneas are not responding to the pressure raises.

It does not mean they have to be central. It only means that, given the pressures on the machine, the machine can't figure out why the apneas won't respond, so it it instructed to drop everything, and let youi know that some of your apnea do not respond despite three pressure jacks.

I don't understand why the machine is setup to let your IPAP go as low as 10.
It seems to be the machine is give far too much headway, starting at anything below an EPAP of 11 or even 12 looks to me like asking for trouble. The machine simply can't resopnd to your apneas at tha pressure.

.

It seems to me that given the lack of apnea resolution because the pressure is too low , restless sleep may be as much a cause of those leaks you have that its result.

As for the statement that one NR is a problem: that is a pet theory of snoredog's.


O,

Posted: Wed Aug 06, 2008 10:08 pm
by Snoredog
ozij wrote:Havnin NR show on your report, simply means the apneas are not responding to the pressure raises.

It does not mean they have to be central. It only means that, given the pressures on the machine, the machine can't figure out why the apneas won't respond, so it it instructed to drop everything, and let youi know that some of your apnea do not respond despite three pressure jacks.

I don't understand why the machine is setup to let your IPAP go as low as 10.
It seems to be the machine is give far too much headway, starting at anything below an EPAP of 11 or even 12 looks to me like asking for trouble. The machine simply can't resopnd to your apneas at tha pressure.

.

It seems to me that given the lack of apnea resolution because the pressure is too low , restless sleep may be as much a cause of those leaks you have that its result.

As for the statement that one NR is a problem: that is a pet theory of snoredog's.


O,
I think I covered the possible "causes" of "NR" in the first post had you taken the time to read it!

As for the bipap and not understanding how it works, IPAP goes to 10 cm because of the EPAP Minimum pressure setting is 8 cm. PS "Minimum" of 2 cm is hard coded in the machine so IPAP Minimum will always land 2 cm higher than EPAP.



_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): bipap


Posted: Wed Aug 06, 2008 10:22 pm
by Snoredog
BigEd wrote: Are you suggesting I switch back to the nasal pillows? That would be fine with me if I can control the mouth breathing. Then I could keep the FF for nights when congested.
I am suggesting a nasal interface over the Full Face mask only because of the better leverage it has in splinting your tongue out of the way. Much easier for it to do that with a nasal interface as opposed to a full face.

As for the Swift? Are those reports with the Swift or the UMFF? Which one is showing all the leak? You can use the Swift if you can get the leak rate down, but if you are needing >18cm pressure you may not be able to control the leak with that interface or a pillow style interface you cannot apply enough pressure on the pillows to stop the leak at those higher pressures, but if that is all you have use it.

Keep the UMFF, you will always need one of those, but you probably will get better therapy with a nasal.

update: your later report using the Swift looks much better (fewer events) than the UMFF and goes along with what I stated above. Use the Swift until you get things to settle down, work on the leaks when you can, even with all the leak your report looks better using the Swift.


Posted: Wed Aug 06, 2008 10:51 pm
by ozij
I don't understand why the machine is setup to let your IPAP go as low as 10.
It seems to me the machine is given far too much headway, starting at anything below an EPAP of 11 or even 12 looks to me like asking for trouble. The machine simply can't resopnd to your apneas at tha pressure.


For the benefit of anyone who did honestly not understand the above statement:

Setting up the EPAP at 8 and letting the PS go down to 2 means there may be times when IPAP is only 10. Given the data I see, that is too low in my opinion, and I don't undrstand why the machine was set up like that by whoever set it up.

An EPAP of 11 or 12 would of course have kept the IPAP at a minimum of 14. Simple reasoning and mathematics.

O.

Posted: Wed Aug 06, 2008 11:33 pm
by Snoredog
ozij wrote:
I don't understand why the machine is setup to let your IPAP go as low as 10.
It seems to me the machine is given far too much headway, starting at anything below an EPAP of 11 or even 12 looks to me like asking for trouble. The machine simply can't resopnd to your apneas at tha pressure.


For the benefit of anyone who did honestly not understand the above statement:

Setting up the EPAP at 8 and letting the PS go down to 2 means there may be times when IPAP is only 10. Given the data I see, that is too low in my opinion, and I don't undrstand why the machine was set up like that by whoever set it up.

An EPAP of 11 or 12 would of course have kept the IPAP at a minimum of 14. Simple reasoning and mathematics.

O.
You NEED to first understand how the machine functions before you can make sense of that statement.

That 2 cm PS you are talking about is hard-coded in the machines firmware, you are NOT going to change it. So there is no "letting PS go down to 2" it is just there as Respironics feels a minimum of 2 cm PS is always needed in the Auto mode.

You CAN adjust the "Maximum" PS which ranges from 3cm to 8cm, but you cannot change the "minimum" value.

There is also NO IPAP "Minimum" on this machine. That Minimum is controlled by Minimum EPAP setting and the built-in 2 cm Pressure Support. SO if you set EPAP=8 then IPAP Minimum is automatically set 2 cm higher at 10 cm, if you set
EPAP=10, iPAP will be at 12 cm and so on.

That built-in 2 cm PS is like a tennis ball between IPAP and EPAP pressure. It will never get closer than 2 cm. The settable Maximum is a like a string or tether between the two pressures, it can go farther apart based upon the PS "Max" setting used which has a range of 3 to 8 cm.

The best way to describe it is the hard coded PS is a tennis ball between the two pressures, the one you set with a range of 3 to 8 is a string. As events dictate those pressures will move independently of each other until it either runs into the tennis ball or reaches the end of that string.

To understand PS and how it works it is NOT just the difference between IPAP and EPAP, it is like an invisible "band" of pressure in the center of Inhale and Exhale, the farther apart you allow PS the wider or taller that band gets.

While they are both called PS for Pressure Support, there is a Minimum and a Maximum. The Minimum is hard-coded, the Maximum you can change and picks up from the hard coded value at 3 cm and goes to 8 cm Maximum. These are rules the machine follows while in "Auto" mode.

The thing to do is use EPAP Minimum setting to titrate and eliminate OA. Once that is acceptable, you manipulate PS setting (going from 3 to in which the algorithm will automatically take IPAP up in the presence of events that drive up IPAP pressure which is mainly HI, VS, and FL. Increasing PS Max setting allows IPAP to separate higher from EPAP pressure without increasing it until it reaches the end of that proverbial "string" where it then begins pulling up EPAP higher.

Now if all the events are mainly OA as seen in BigEd's case, those apnea will trigger EPAP to increase, when that happens EPAP runs into that hard coded 2 cm PS and begins pushing up IPAP. It is shown right in the reports, count the pressure tics, when you see 2 tic separation between IPAP and EPAP "working" pressure (not the range) that says EPAP is pushing up IPAP based upon current SDB event.

When you see 3 cm or greater separation then that means IPAP is responding independently and if EPAP rises it is because it reached the end of the proverbial string of the settable PS setting and is now pulling up EPAP. This falls then under the little hammer big hammer theory SAG described years ago.

IF you want to allow greater separation or greater PS you increase the PS setting from default=3 to 4 or higher all the way up to 8. 8 is the Maximum that machine will allow. The only reason you want to allow a greater PS is to keep the EPAP pressure down independently. For example if the report shows that SDB events are all HI you might want to allow IPAP to go higher without pulling up EPAP. But this can work against you on the down side. As events subside that "string" works the opposite, as EPAP falls for absence of events it has to reach that PS setting before it pulls IPAP down. If PS is at 4 it will pull that IPAP pressure down before it will at a setting of 8.

It is very important on this machine to allow it a wide range of operation. The higher the PS setting you use the wider that range has to be. That "range" is determined by the EPAP Minimum setting and the IPAP Maximum setting in addition to any PS setting you use.

For example, it does NO good setting EPAP at 12 and IPAP at 16 and using a PS=8 it will NEVER be used as there isn't 8 cm separation allowed in that range.

Eliminate OA first with EPAP setting, then observe reports, if HI persist extend the PS setting. Goal is keeping EPAP at a level where it just eliminates the OA then allow IPAP to run up to clean up the rest. If you are still seeing HI then shortening the PS setting will force the machine to bring out the big hammer and increase EPAP sooner (not faster, sooner).

Those settings are not a set it here and hope it works, you observe OA and increase the one designed to get the job done. Use the legend at the bottom of the Daily report to determine which pressure needs to change to eliminate the event. Notice HI is on IPAP and OA is on EPAP? Adjust the pressure for the event you want to prevent. EPAP being the big hammer eventually takes care of all events just like CPAP pressure does.


Posted: Thu Aug 07, 2008 7:29 am
by BigEd
Thanks for the continued interest, advice and education. Last night I switched back to the Swift II nasal pillows. I changed my EPAP setting to 12 and PS to 4 and ordered SleepGuy's PAP-Cap.

When I first started therapy, I used the "Nose Breathe Mouth Guard". For about 12 months, this mouth guard did a nice job of preventing mouth breathing. I stopped using it because I felt I could continue to nose breathe without the guard. After about 6 months the mouth breathing returned.

So last night, instead of taping, I tried using my mouth guard again. Didn't work so well this time, as far as large leaks go. Guess I need to get used to it again. Had 113.5 min of large leaks, average leak 67.68, no NRs, AHI of 1.1 and 90% pressure readings of IPAP 16.8 and EPAP 13.9. I’ll try taping tonight.

I didn't bother posting last night's report sheet because I wasn't sure it was pertinent with the amount of large leak time affecting the data.

I'm hopeful that once I get the PAP-Cap and test a combination of various additional aids (mouth guard, tape, etc) I'll be able to bring the leaks under control and generate some decent numbers.


Posted: Thu Aug 07, 2008 9:38 am
by Snoredog
[quote="BigEd"]Thanks for the continued interest, advice and education. Last night I switched back to the Swift II nasal pillows. I changed my EPAP setting to 12 and PS to 4 and ordered SleepGuy's PAP-Cap.

When I first started therapy, I used the "Nose Breathe Mouth Guard". For about 12 months, this mouth guard did a nice job of preventing mouth breathing. I stopped using it because I felt I could continue to nose breathe without the guard. After about 6 months the mouth breathing returned.

So last night, instead of taping, I tried using my mouth guard again. Didn't work so well this time, as far as large leaks go. Guess I need to get used to it again. Had 113.5 min of large leaks, average leak 67.68, no NRs, AHI of 1.1 and 90% pressure readings of IPAP 16.8 and EPAP 13.9. I’ll try taping tonight.

I didn't bother posting last night's report sheet because I wasn't sure it was pertinent with the amount of large leak time affecting the data.

I'm hopeful that once I get the PAP-Cap and test a combination of various additional aids (mouth guard, tape, etc) I'll be able to bring the leaks under control and generate some decent numbers.


Why is my AHI so high?

Posted: Thu Aug 07, 2008 9:50 am
by Hawthorne
BigEd - What is a "Nose Breathe Mouth Guard"?
I have not heard of it before.
Thanks.

Posted: Thu Aug 07, 2008 9:52 am
by BigEd
Snoredog wrote:Well that looks like an improvement Ed. Do you happen to have a conventional nasal mask like a ComfortGel or a UltraMirageII to compare to the Swift? Those might be able to handle the pressure better and should tell you if leak is a mouth breathing issue or leaking around the pillows.

The NR looks much better from last night, that means the pressure is able to splint the tongue out of the way, hopefully it resulted in fewer events on the report. Next question would be be do you feel any better from it. AHI of 1.1 looks good.
No. I don't own a conventional nasal mask. My arsenal includes original Swift nasal pillows, Swift II, Hybrid, Ultra Mirage FF. I'm pretty confident that the leaks are from mouth breathing. Dry mouth being one of the results. And yes, I do actually feel better. If I have a night with minimal large leaks, I'll post the results tomorrow. Thanks again for the help.


Re: Why is my AHI so high?

Posted: Thu Aug 07, 2008 9:55 am
by BigEd
Hawthorne wrote:BigEd - What is a "Nose Breathe Mouth Guard"?
I have not heard of it before.
Thanks.
It's a device that helps hold your tongue in place to aid in creating a seal. Here is the link to the site. Worked for me: http://www.nosebreathe.com


Re: Why is my AHI so high?

Posted: Thu Aug 07, 2008 10:48 am
by Snoredog
BigEd wrote:
Hawthorne wrote:BigEd - What is a "Nose Breathe Mouth Guard"?
I have not heard of it before.
Thanks.
It's a device that helps hold your tongue in place to aid in creating a seal. Here is the link to the site. Worked for me: http://www.nosebreathe.com

Re: Why is my AHI so high?

Posted: Thu Aug 07, 2008 11:12 am
by BigEd
Snoredog wrote:Actually, I was going to suggest this one:

http://www.snoremeds.com/

It costs less at $45, is a double unit which marketed as a snore reduction device, it will advance your mandible a tiny bit which opens the airspace up at the back of the throat, this may result in few apnea events seen in addition to reducing snoring. With its double (upper and lower) design it may also work to reduce mouth breathing. But keep in mind you probably will still need to use a chin strap with light vertical pressure it to keep the mandible from dropping when you fully relax. If you have SleepGuy's pap cap on order that will do the job nicely.
Thanks. Looks like it might do the trick. Especially combined with SleepGuy's PAP-Cap. I'll give it a try.

Posted: Thu Aug 07, 2008 11:40 am
by jnk
O. and Snoredog,

Exhanges like the above between you two are what make this forum so valuable to me. And this one definitely shows how much you both want to be helpful to BigEd.

But from where I sat, it sure sounded to me like you were both saying pretty much the same thing: the best way to keep the ipap from going too low on that machine is to make sure the epap isn't set too low. Right?

Or maybe my brain lesions are acting up. If so, never mind.

jnk

AHI so high?

Posted: Thu Aug 07, 2008 11:56 am
by Hawthorne
Thanks BigEd!