I have to say I am mighty impressed with the content of this thread! I have only skimmed over it but most of it is very interesting. I'll have to take some time to read it in more detail.
Velbor wrote:
So .... if not radially (assuming that anything I have expressed above makes any sense at all), does CPAP work by exerting its pressure "axially," along the length of the upper airway? Here, at least, I can find pressure differentials, with respect to the flexible-walled chest and the lungs. Does CPAP work, then, by its pressure differential (with respect to the negative pressure within the thorax on inspiration) "pushing axially downward" (i.e., toward the lungs)?
Velbor: You raise some interesting points. I'll focus on the above point as it kind of relates to my work. While CPAP is thought to maintain airway patency by "splinting" the airway via positive pressure, axial tension is also likely to prevent collapse. The airway is connected to several intrathoracic structures. These structures are pulled downwards during inspiration along with the diaphragm. This action helps to stretch the airway, reducing airway collapsibility.
There have been several studies conducted in animals showing an inverse relationship between the degree of axial tension on the airway and airway collapsibility. This effect is independent of the influence of airway dilator muscles.
Van de Graaff also demonstrated the importance of axial tension (known as caudal traction) on upper airway resistance. Resistance is related to the radius of the "tube." The smaller the radius, the higher the resistance (actually resistance is inversely proportional to the radius^4). In this study, Van de Graaff showed that airway resistance went down during inspiration, indicating that the airway was getting bigger. It was throught that the increase in airway size was due to the actions of the dilator muscles. Therefore, Van de Graaff hyperventilated the dogs to apnea, thus "switching" off the dilator muscles. The dogs were then mechanically ventilated. Again, airway resistance decreased during inspiration, thus indicating that the dilator muscles were in fact not solely involved in increasing airway size. Van de Graaff then went on to cutting all cervical structures which linked the thorax to the airway. This resulted in airway collapse. These data suggest that axial tension generated by the descending diaphragm helps to maintain airway patency during inspiration.
Van de Graaff WB.
Department of Medicine, Stritch School of Medicine, Loyola University, Maywood, Illinois 60153.
Patency of the upper airway (UA) is usually considered to be maintained by the activity of muscles in the head and neck. These include cervical muscles that provide caudal traction on the UA. The thorax also applies caudal traction to the UA. To observe whether this thoracic traction can also improve UA patency, we measured resistance of the UA (RUA) during breathing in the presence and absence of UA muscle activity. Fifteen anesthetized dogs breathed through tracheostomy tubes. RUA was calculated from the pressure drop of a constant flow through the isolated UA. RUA decreased 31 +/- 5% (SEM) during inspiration. After hyperventilating seven of these dogs to apnea, we maximally stimulated the phrenic nerves to produce paced diaphragmatic breathing. Despite absence of UA muscle activity, RUA fell 51 +/- 11% during inspiration. Graded changes were produced by reduced stimulation. In six other dogs we denervated all UA muscles. RUA still fell 25 +/- 7% with inspiration in these spontaneously breathing animals. When all caudal ventrolateral cervical structures mechanically linking the thorax to the UA were severed, RUA increased and respiratory fluctuations ceased. These findings indicate that tonic and phasic forces generated by the thorax can improve UA patency. Inspiratory increases in UA patency cannot be attributed solely to activity of UA muscles.
So, CPAP certainly may help to prevent airway collapse by stretching the airway as speculated by yourself. This axial tension is likely due to the caudal displacement of the diaphragm and other intrathoracic structures caused by CPAP.
Another process which helps prevent airway collapse is lung inflation. This process is likely to work via axial tension forces. Heinzer et al undertook two studies looking at the effect of lung volume on AHI and also CPAP requirements to prevent flow limitation.
Effect of increased lung volume on sleep disordered breathing in patients with sleep apnoea.Heinzer RC, Stanchina ML, Malhotra A, Jordan AS, Patel SR, Lo YL, Wellman A, Schory K, Dover L, White DP.
Sleep Medicine Division, Brigham and Women's Hospital, Harvard Medical School, Boston, MA, USA.
rheinzer@post.harvard.edu
BACKGROUND: Previous studies have shown that changes in lung volume influence upper airway size and resistance, particularly in patients with obstructive sleep apnoea (OSA), and that continuous positive airway pressure (CPAP) requirements decrease when the lung volume is increased. We sought to determine the effect of a constant lung volume increase on sleep disordered breathing during non-REM sleep. METHODS: Twelve subjects with OSA were studied during non-REM sleep in a rigid head-out shell equipped with a positive/negative pressure attachment for manipulation of extrathoracic pressure. The increase in lung volume due to CPAP (at a therapeutic level) was determined with four magnetometer coils placed on the chest wall and abdomen. CPAP was then stopped and the subjects were studied for 1 hour in three conditions (in random order): (1) no treatment (baseline); (2) at "CPAP lung volume", with the increased lung volume being reproduced by negative extrathoracic pressure alone (lung volume 1, LV1); and (3) 500 ml above the CPAP lung volume(lung volume 2, LV2). RESULTS:
The mean (SE) apnoea/hypopnoea index (AHI) for baseline, LV1, and LV2, respectively, was 62.3 (10.2), 37.2 (5.0), and 31.2 (6.7) events per hour (p = 0.009); the 3% oxygen desaturation index was 43.0 (10.1), 16.1 (5.4), and 12.3 (5.3) events per hour (p = 0.002); and the mean oxygen saturation was 95.4 (0.3)%, 96.0 (0.2)%, 96.3 (0.3)%, respectively (p = 0.001). CONCLUSION: An increase in lung volume causes a substantial decrease in sleep disordered breathing in patients with OSA during non-REM sleep.
I want to focus on the LV1 result. Basically, the authors measured the change in lung volume caused by therapeutic CPAP. They then increased lung volume up to this volume via negative extrathoracic pressure. The patients were allowed to fall asleep at this lung volume without CPAP. As you can see, AHI decreased from ~62 events down to ~37 events. Not really treated hey? BUT the point was that increasing lung volume does indeed reduce airway collapsibility. The authors went on to state:
Because the subjects we studied were overweight or obese
(like most patients with OSA), we suspect that their
diaphragm was pushed upwards (cranially) when lying on
their back. The negative extrathoracic pressure applied in the
lung almost certainly pulled the diaphragm and trachea to a
more caudal position, thereby increasing the traction on the
upper airway and making it less collapsible.
And this....
In the present study we have shown that an increase in lung volume
alone (without CPAP) is able to decrease the apnoea and
hypopnoea frequency significantly, which supports this
hypothesis. The rest of the effect of CPAP is probably due
to a splinting effect of positive airway pressure on the upper
airway, maintaining a positive transmural pressure throughout
the respiratory cycle.
This is an abstract from their other study:
Lung volume and continuous positive airway pressure requirements in obstructive sleep apnea.Heinzer RC, Stanchina ML, Malhotra A, Fogel RB, Patel SR, Jordan AS, Schory K, White DP.
Division of Sleep Medicine, Sleep Disorders Program @ BI, Brigham and Women's Hospital, 75 Francis Street, Boston, MA 02115, USA.
rheinzer@rics.bwh.harvard.edu
Previous studies have demonstrated that lung volume during wakefulness influences upper airway size and resistance, particularly in patients with sleep apnea. We sought to determine the influence of lung volume on the level of continuous positive airway pressure (CPAP) required to prevent flow limitation during non-REM sleep in subjects with sleep apnea. Seventeen subjects (apnea-hypopnea index, 42.6 +/- 6.2 [SEM]) were studied during stable non-REM sleep in a rigid head-out shell equipped with a positive/negative pressure attachment for manipulation of extrathoracic pressure. An epiglottic pressure catheter plus a mask/pneumotachometer were used to assess flow limitation. When lung volume was increased by 1,035 +/- 22 ml, the CPAP level could be decreased from 11.9 +/- 0.7 to 4.8 +/- 0.7 cm H(2)O (p < 0.001) without flow limitation. The decreased CPAP at the same negative extrathoracic pressure yielded a final lung volume increase of 421 +/- 36 ml above the initial value. Conversely, when lung volume was reduced by 732 +/- 74 ml (n = , the CPAP level had to be increased from 11.9 +/- 0.7 to 17.1 +/- 1.0 cm H(2)O (p < 0.001) to prevent flow limitation, with a final lung volume decrease of 567 +/- 78 ml. These results demonstrate that relatively small changes in lung volume have an important effect on the upper airway in subjects with sleep apnea during non-REM sleep.
Again, this indicates that lung volume, and likely axial tension forces, helps to maintain airway patency.
Snoredog wrote:
I'm sure Split_City would beg to differ with logic and our way of thinking, however I will say with "nasal" pressure being applied it should have enough leverage to dislodge a tongue (e.g. cork as it was referred to), this unless the persons anatomy is so screwed up that doesn't happen.
but an excellent post and all good questions, I refer to that tube as a firehose on the asphalt with not much pressure behind it, cpap cranks up the pressure and that firehose becomes nice and round.
I have no objection with the point about dislodging the tongue. I just say that the tongue is only part of the puzzle. The firehose analogy is acceptible but not complete. The action of the water is predominantly acting radially whereas I believe (and speculated by Velbor) that axial forces are at work as well. My abdominal compressin study and work by Heinzer and colleagues certainly does support this notion.
Another area of interest in this thread. I read up about whether the initial collapse occurs via suction pressures i.e. during inspiration, or passively during expiration. I am currently looking at this in my current study. From what I have seen, collapse predominantly occurs towards late-to-end EXPIRATION and NOT during INSPIRATION. I will attempt to paste in some examples when I get a chance.
This is not surprising given that the activity of the airway dilator muscles decreases during expiration and surrounding tissue pressure increases at the same time. Increased tissue pressure will likely lead to passive collapse of the "flaccid" airway during/at the end of expiration.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition):
CPAP,