Posted: Mon Jun 16, 2008 5:12 pm
This is a rather old thread, and I being relatively new, just happened to stumble on it now By now, socknitster might already have her baby ??
I have been thinking about this issue on and off for the last 4 weeks. I have used the headrest about half the time. While oxygen sat tests are good way of finding it for a particular person, there may still be differences between masks for their CO2 wash out efficiency. Even though I had close to moderate/severe apnea, during the initial non titration study my O2 sat never went below 94% or so, so that by itself is not good enough. One wants to get the best possible given all of one's own non-changeable circumstances, IMO.
Here is my analysis (I have a scientific background, but not in CPAP machines. I have designed and got FDA approval for medical instruments.)
The wash out of the breathed out CO2 happens because of the constant pressure, and the constant flow through the out port, as people have pointed out. However much of these gets washed out depends on a lot of factors.
1. If your pressure is high, more of CO2 gets washed out.
2. If you breath out slowly more of CO2 gets pushed out.
3. The more the inherent built in leak through the out port for a given pressure, more of the CO2 gets pushed out. You can check the leak rate for a particular mask from its spec sheets, usually. The Headrest spec is 15LPM at 4cmH2O and 21LPM at 8cmH2O.
4. Now how far away the outport is from the nose tip, and the volume of air in the tube from the nose to the outport is the trapped a CO2 volume, only part of it will get pushed out. Some manufacturers may list this in the spec as the 'Seal Dead Space', which for the Headrest is 35cc --- I have a guess this may be higher for the Headrest than others.
Now here are some factors which may not be that obvious.
5. When you breath out, the qty you breath out does not simply go out immediately through the out port. The air in the tube leading to the machine will momentarily compress and some of the out breath will go into that tube, and get mixed in with the incoming air, the concentration of the CO2 near the face will be higher, decreasing as you go towards the machine.
6. The longer the tube, there is more take up of this used air into the tube!
7. If you have Cflex or Aflex, the pressure and hence the outport flow, as well as the resistance for the used air to enter the long tube, is DECREASED while breathing out. So Cflex and Aflex will decrease the CO2 push out compared to fixed pressure, for the same pressure setting.
Now you can do this simple experiment to get a feel of how long it takes to push out your breath thru the outport. Wearing the mask or nasal pillow, disconnect it from the main hose (so you have a relative small tube remaining). Do a couple of regular breathing with the tube open. Now close the open end of the tube tightly with your palm, and see how long it takes to push out your inhaled air thru the normal outport. You would be surprised how longer this is compared to your normal breath out time. Remember the actual time needed to push out all of your CO2 is going to be even longer than this, since the air that is pushed out is mixed with some good incoming air.
So what we need to know, CO2 wash out is rather complex to simply figure out. There need to be a specification for CO2 wash out efficiency, which is after some time of breathing, when steady state conditions are archived, what percentage of the CO2 is getting continuously removed. Your CO2 removal will be reduced by this percentage when you are using the cpap as opposed normal breathing. I have not seen any specification in my limited experience with any of couple of masks I have looked at.
Just becausue FDA approves something does not mean, they have this issue well covered. It depends on a lot of factors other than just the mask. Perhaps there is already a measure established. I would image this has already been done well for scuba diving gear, since the O2/CO2 levels are much more critical for that. If anybody has direct knowledge or experience kindly teach the rest of us.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): hose, nasal pillow, cflex, Titration, CPAP
I have been thinking about this issue on and off for the last 4 weeks. I have used the headrest about half the time. While oxygen sat tests are good way of finding it for a particular person, there may still be differences between masks for their CO2 wash out efficiency. Even though I had close to moderate/severe apnea, during the initial non titration study my O2 sat never went below 94% or so, so that by itself is not good enough. One wants to get the best possible given all of one's own non-changeable circumstances, IMO.
Here is my analysis (I have a scientific background, but not in CPAP machines. I have designed and got FDA approval for medical instruments.)
The wash out of the breathed out CO2 happens because of the constant pressure, and the constant flow through the out port, as people have pointed out. However much of these gets washed out depends on a lot of factors.
1. If your pressure is high, more of CO2 gets washed out.
2. If you breath out slowly more of CO2 gets pushed out.
3. The more the inherent built in leak through the out port for a given pressure, more of the CO2 gets pushed out. You can check the leak rate for a particular mask from its spec sheets, usually. The Headrest spec is 15LPM at 4cmH2O and 21LPM at 8cmH2O.
4. Now how far away the outport is from the nose tip, and the volume of air in the tube from the nose to the outport is the trapped a CO2 volume, only part of it will get pushed out. Some manufacturers may list this in the spec as the 'Seal Dead Space', which for the Headrest is 35cc --- I have a guess this may be higher for the Headrest than others.
Now here are some factors which may not be that obvious.
5. When you breath out, the qty you breath out does not simply go out immediately through the out port. The air in the tube leading to the machine will momentarily compress and some of the out breath will go into that tube, and get mixed in with the incoming air, the concentration of the CO2 near the face will be higher, decreasing as you go towards the machine.
6. The longer the tube, there is more take up of this used air into the tube!
7. If you have Cflex or Aflex, the pressure and hence the outport flow, as well as the resistance for the used air to enter the long tube, is DECREASED while breathing out. So Cflex and Aflex will decrease the CO2 push out compared to fixed pressure, for the same pressure setting.
Now you can do this simple experiment to get a feel of how long it takes to push out your breath thru the outport. Wearing the mask or nasal pillow, disconnect it from the main hose (so you have a relative small tube remaining). Do a couple of regular breathing with the tube open. Now close the open end of the tube tightly with your palm, and see how long it takes to push out your inhaled air thru the normal outport. You would be surprised how longer this is compared to your normal breath out time. Remember the actual time needed to push out all of your CO2 is going to be even longer than this, since the air that is pushed out is mixed with some good incoming air.
So what we need to know, CO2 wash out is rather complex to simply figure out. There need to be a specification for CO2 wash out efficiency, which is after some time of breathing, when steady state conditions are archived, what percentage of the CO2 is getting continuously removed. Your CO2 removal will be reduced by this percentage when you are using the cpap as opposed normal breathing. I have not seen any specification in my limited experience with any of couple of masks I have looked at.
Just becausue FDA approves something does not mean, they have this issue well covered. It depends on a lot of factors other than just the mask. Perhaps there is already a measure established. I would image this has already been done well for scuba diving gear, since the O2/CO2 levels are much more critical for that. If anybody has direct knowledge or experience kindly teach the rest of us.