How should my 90 % pressure relate to apap setting?
How should my 90 % pressure relate to apap setting?
Hi, all. Ok. a few weeks now on ap and even took it while travelling. I have managed to get myself down to the 2.0-4.0 AHI range.. keeping the pressure steady now from 9-14 for 2 weeks, it shows that my 90% pressure is around 12.. Do I change my settings? Not sure how to use the 90% data, although i understand what it means statistically. some stats below:
9 is used for 30% of night
10 is used for 25% of night
11, 12, 14 each 15% of night
9 is used for 30% of night
10 is used for 25% of night
11, 12, 14 each 15% of night
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
What did you put in you magic air box, DS?DreamStalker wrote:Your stats only add up to 70%.
25 + 30 + 3*15 = 100
That still doesn't make 12 a 90% recommendation; and what happened to 13?
O.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
It must not have been the good stuff ...ozij wrote:What did you put in you magic air box, DS?DreamStalker wrote:Your stats only add up to 70%.
25 + 30 + 3*15 = 100
That still doesn't make 12 a 90% recommendation; and what happened to 13?
O.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
0% for 13?ozij wrote:What did you put in you magic air box, DS?DreamStalker wrote:Your stats only add up to 70%.
25 + 30 + 3*15 = 100
That still doesn't make 12 a 90% recommendation; and what happened to 13?
O.
You didn't get the good stuff either ...
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
Re: How should my 90 % pressure relate to apap setting?
Sorry, back to your question ...neillebo wrote:Hi, all. Ok. a few weeks now on ap and even took it while travelling. I have managed to get myself down to the 2.0-4.0 AHI range.. keeping the pressure steady now from 9-14 for 2 weeks, it shows that my 90% pressure is around 12.. Do I change my settings? Not sure how to use the 90% data, although i understand what it means statistically. some stats below:
9 is used for 30% of night
10 is used for 25% of night
11, 12, 14 each 15% of night
Assumming you have mouth leaks under control (you do don't you?) ... you could try narrowing your range by increasing your min pressure setting to 10 cm.
Yes ... sleepie beat me to it ... what sleepie said.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
sorry. I meant 15 at 11-12 & 13. I also rounded, i have a small % at 14, but not much. I was thinking of raising to 10, but my titration in the lab was 9.. 9 has been my min and i get plenty of events above it, so that is why i have been cautious. If all concur, i like the idea of raising to 10 to see what happens... of courese, that still doesnt answer, if you dont mind, what the 90% pressure means. Even if i didnt have that stat, i would consider raising the band to 10 since i still get an ahi of 2+ and many event at 9... does the 90% pressure tell me to keep the band closer to 12 (bear in mind, for the first week or so, 90% was 13.
Yes,. leaks are ok.. average around 33 each night.. how would a mouth leak show up, btw? does it show as apnea since i wont be breathing through my nose? does it show as a leak? not sure how that works out.
Yes,. leaks are ok.. average around 33 each night.. how would a mouth leak show up, btw? does it show as apnea since i wont be breathing through my nose? does it show as a leak? not sure how that works out.
- DreamStalker
- Posts: 7503
- Joined: Mon Aug 07, 2006 9:58 am
- Location: Nowhere & Everywhere At Once
90% means that is the pressure for which you spent 90% of your time at or below. If you were going to set it to straight CPAP mode, that would be a good number to set it at. I was titrated at 10 but a year's worth of data suggests 11.5 or 12 based on my 90% stat.
If you had a mouth leak, you would have a leak rate higher than what is desiganted for your mask.
Apneas may be correlated to high leak rates on your nightly data using the software ... however, it is possible to have apneas without it being associated with a high leak too.
If you had a mouth leak, you would have a leak rate higher than what is desiganted for your mask.
Apneas may be correlated to high leak rates on your nightly data using the software ... however, it is possible to have apneas without it being associated with a high leak too.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.
I'm not sure i follow how a mouth leak will be evidenced by higher leak than my mask is rated. My mask is rated around 30 and i leak around 33.. I have been assuming that i am NOT mouth leaking and that this is a leak around my nose when i turn over, etc.... I konw a mouth leak would also cause the pressure leak to rise, but I am trying to figure out if there is a way to tell if i really have any mouth leaks??
Sleepy, when you say you run it 1 point over, is that in cpap mode or apap? meaning you put your MAX pressure at 1 level above the 90%? where do you put min in relation to 90%?Posted: Sun Dec 02, 2007 9:30 am Post subject: Delete Spam
--------------------------------------------------------------------------------
i run mine 1 point over my 90%---seems to work for me----the extra point for those hard to knock out apneas
_________________
-
Guest
-
Sleepdeprived
- Posts: 325
- Joined: Mon May 14, 2007 3:07 pm
- Location: Florida
Re: How should my 90 % pressure relate to apap setting?
Hi
I was titrated at 9 cm
My 90 % was around 13
by best results are from straight cpap at 9
Apap and 90 % pressure does not always give the you
the best therapy
I am now down to .4 AHI for the week on cpap when my apap was over 5
It was Snoredog who first suggested to me I might be better off on straight cpap several months ago
Experimentation is often needed to find ones sweet spot
I experienced improved therappy not only with cpap vs apap but also in going from the Bravo to the Swift to the Nassal Aire II
Also with poligrip strips, no humidity vs humidity (I live in a humid place,) pantyhose tights , pursleep, melatonin and am now experimenting with the pappillow
Find your best results and if it is consistently at one pressure you might want to do straight cpap
and then tweak you therapy by other factors
Sleepdeprived
[quote="neillebo"]Hi, all. Ok. a few weeks now on ap and even took it while travelling. I have managed to get myself down to the 2.0-4.0 AHI range.. keeping the pressure steady now from 9-14 for 2 weeks, it shows that my 90% pressure is around 12.. Do I change my settings? Not sure how to use the 90% data, although i understand what it means statistically. some stats below:
9 is used for 30% of night
10 is used for 25% of night
11, 12, 14 each 15% of night
I was titrated at 9 cm
My 90 % was around 13
by best results are from straight cpap at 9
Apap and 90 % pressure does not always give the you
the best therapy
I am now down to .4 AHI for the week on cpap when my apap was over 5
It was Snoredog who first suggested to me I might be better off on straight cpap several months ago
Experimentation is often needed to find ones sweet spot
I experienced improved therappy not only with cpap vs apap but also in going from the Bravo to the Swift to the Nassal Aire II
Also with poligrip strips, no humidity vs humidity (I live in a humid place,) pantyhose tights , pursleep, melatonin and am now experimenting with the pappillow
Find your best results and if it is consistently at one pressure you might want to do straight cpap
and then tweak you therapy by other factors
Sleepdeprived
[quote="neillebo"]Hi, all. Ok. a few weeks now on ap and even took it while travelling. I have managed to get myself down to the 2.0-4.0 AHI range.. keeping the pressure steady now from 9-14 for 2 weeks, it shows that my 90% pressure is around 12.. Do I change my settings? Not sure how to use the 90% data, although i understand what it means statistically. some stats below:
9 is used for 30% of night
10 is used for 25% of night
11, 12, 14 each 15% of night