Posted: Wed Nov 28, 2007 8:39 pm
Eh, no rubbing or lashing intended.
That is not an optional "as needed" wash. And if you friend needs better filtering - he should add the white, fine particle filter. He would get cleaner air adding that filter, and changing it as frequently as needed than he will by buying a new gray filter once a month.7.2 Cleaning or Replacing the Filters
Caution: Operating the device with a dirty filter may keep the system from working properly and may damage the device.
Under normal usage, you should clean the gray foam filter at least once every two weeks and replace it with a new one every six months. The white ultra-fine filter is disposable and should be replaced after 30 nights of use or sooner if it appears dirty. DO NOT clean the ultra-fine filter.
Caution: Dirty inlet filters may cause high operating temperatures that may affect device performance. Regularly examine the inlet filters as needed for integrity and cleanliness.
<snip>
4. Wash the gray foam filter in warm water with a mild detergent. Rinse thoroughly to remove all detergent residue. Allow the filter to air dry completely before reinstalling it. If the foam filter is torn, replace it.