Re: Hah?
Posted: Thu Jul 05, 2007 9:05 pm
Ok, I have asked our sleep lab manager why we have classified 'normal' RDI as <15. I must make an apology as I had assumed this was the criteria for all labs in Australia. However, this is the criteria set for our lab only. Other labs have not adopted it at this stage. The reason why we classify normal as being <15 is due to a study our lab ran which looked at what normal was in a group of asymptomatic normal subjects. This is also at the bottom of our sleep reports
Normative data: In an Australian study1 of randomly-selected asymptomatic normal subjects aged 30 - 70 years respiratory disturbance index (apneas + hypopnoeas/ hr of sleep) was 6.1 + 7.2 (Mean + 2SD). i.e. RDI upper limit of "normal" in this population was approximately 13. Clinicians should, however, note that RDI was averaged for the whole study population and RDI is known to increase with age. Further, epidemiological studies suggest that increased risk of hypertension, cardiovascular disease and neurobehavioural disturbance may be present at RDIs as low as 5-10/ hr. Thus, when advising an individual patient, judgement is required in assessing what constitutes a "normal" RDI
Current versus historical RDI values.
This along with the use of pressure transducers (information below) resulted in our lab's current criteria for assessing OSA severity
In accordance with recent internationally agreed recommendations2, the AISH laboratories transferred in 2002 from using oronasal thermistors to using nasal pressure transducers to measure airflow and at the same time adopted a new standard for scoring apneas and hypopnoeas2. Nasal pressure is a more sensitive measure of airflow and as a consequence, average AISH reported RDI values are now higher for the same severity sleep disordered breathing. The best estimate is that RDI has increased by approximately 40.%. This is relevant when interpreting current patient results with respect to our earlier studies of normal subjects1 and other previous epidemiological studies, which relied on thermistors to measure airflow.
Severity of sleep disordered breathing. Bearing in mind the above, AISH consultants have adopted the following consensus classification of severity of sleep disordered breathing
RDI: <15 Normal (see above caveats)
>15 - <30 Mild
>30 - <45 Moderate
>45 Severe
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): Titration, Arousal, RDI, Hypopnea, APAP
Normative data: In an Australian study1 of randomly-selected asymptomatic normal subjects aged 30 - 70 years respiratory disturbance index (apneas + hypopnoeas/ hr of sleep) was 6.1 + 7.2 (Mean + 2SD). i.e. RDI upper limit of "normal" in this population was approximately 13. Clinicians should, however, note that RDI was averaged for the whole study population and RDI is known to increase with age. Further, epidemiological studies suggest that increased risk of hypertension, cardiovascular disease and neurobehavioural disturbance may be present at RDIs as low as 5-10/ hr. Thus, when advising an individual patient, judgement is required in assessing what constitutes a "normal" RDI
Current versus historical RDI values.
This along with the use of pressure transducers (information below) resulted in our lab's current criteria for assessing OSA severity
In accordance with recent internationally agreed recommendations2, the AISH laboratories transferred in 2002 from using oronasal thermistors to using nasal pressure transducers to measure airflow and at the same time adopted a new standard for scoring apneas and hypopnoeas2. Nasal pressure is a more sensitive measure of airflow and as a consequence, average AISH reported RDI values are now higher for the same severity sleep disordered breathing. The best estimate is that RDI has increased by approximately 40.%. This is relevant when interpreting current patient results with respect to our earlier studies of normal subjects1 and other previous epidemiological studies, which relied on thermistors to measure airflow.
Severity of sleep disordered breathing. Bearing in mind the above, AISH consultants have adopted the following consensus classification of severity of sleep disordered breathing
RDI: <15 Normal (see above caveats)
>15 - <30 Mild
>30 - <45 Moderate
>45 Severe
No, we try to abolish respiratory events. But this isn't achievable in all cases.StillAnotherGuest wrote: Hmmm, now SAG is greatly confused. Regardless of the diagnostic airflow methodology used, during intervention the waveform from the PAP machine is used to score events, and that is a pressure transducer. Are you then saying that RDI during titration and APAP is normal as long as it is <15?
Yes, I'm pretty sure what you have in bold are the components of RDIStillAnotherGuest wrote:Also, what are your components of RDI? Are they apneas, hypopneas and RERAs? RERAs and hypopneas need to have an associated event, be it an arousal and/or a desaturation. And >5 events per hour of those events are abnormal.
I think the confusion to why our RDI is so high is due to our classification of what a hypopnea is. Our hypopnea rule is "a discernable decrease in amplitude (flow) accompanied by a 3% or greater desaturation." I.e. a 50% decrease in flow is not usedStillAnotherGuest wrote:Or put another way, the diagnostic device selected is really academic. Either device (thermistor or pressure transducer) will identify an apnea, but beyond that, a respiratory event without an associated desaturation or at least an arousal is nothing.
SAG
Or put another, another way, let's say a patient has 14 respiratory events, all of which cause desaturations. Under the thermistor rule, that would be a problem. Are you saying that under pressure transducer, that's no longer a problem?
OK, let's say instead, there were 14 arousals that were associated with thermistor-defined events. Are you saying that with pressure transducer technology, they are no longer considered to be respiratory-associated events? Especially if the waveforms have a flattened appearance? And if not, then what do you call them?
The whole idea behind pressure transducer is that it picks up events that thermistor may miss. And it's not the respiratory event per se that's the issue, it's the associated desaturation, arousal and/or increased effort (i.e., the flattened waveform seen in UARS). And that number is almost universally accepted as being 5.
SAG
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): Titration, Arousal, RDI, Hypopnea, APAP