Because these tests are so extensive I'll post different tests as I do them & can then gather the data into spreadsheets later. Also they will be easier to digest.
Test 1 using a Invocare Polaris ISP 9800 (flying Saucer) Cpap (worked out how to change the pressure again) ...
http://www.cpapwholesale.com/invacare-polaris-cpap.htm
(PS - this machine acts like it has a pressure transducer & I see in the above link they say it has one so this adds a variation from using the Healthdyne Quest that I don't believe does have a transducer. Also, notice below that the CMS did not jump when I put the mask on (no breathing), but stayed the same compared to just blocking the hose).
*********************************************************
Cpap set to 16 CMS
6' hose - blocked -
Manometer at the machine end - cpap led says 16, manometer says 16.
Manometer at the mask end - cpap led says 16, manometer says 16.
12' hose (2 x 6')
Manometer at the machine end - cpap led says 16, manometer says 16.
Manometer at the mask end - cpap led says 16, manometer says 16.
SUMMARY 1:
Static (no airflow) pressure shows pressure equal at all points.
********************************************************
With UMFF Std mask attached - no breathing (but allowing fixed leak venting)
Cpap set to 16 CMS - 6' hose - no breathing
Manometer at cpap end - Cpap led = 16 - Manometer says 15.75 cms
Manometer at mask end - Cpap led = 16 - Manometer says 15.25 cms
Cpap set to 16 CMS - 6' hose - breathing in
Manometer at cpap end - Cpap led = 15 - Manometer says 15 cms
Manometer at mask end - Cpap led = 15 - Manometer says 14 cms
Cpap set to 16 CMS - 6' hose - breathing out
Manometer at cpap end - Cpap led = 16 (spikes to 17 ) - Manometer says 16 cms
Manometer at mask end - Cpap led = 16 (spikes to 17 ) - Manometer says 15.5 cms
SUMMARY 2:
The 1st figure (15.25 & 15.75 ) indicate a friction loss of .5 due to venting alone.
When breathing in, the friction loss grows due to faster air flow (& greater friction)
and this is even more obvious in the 12' tube below.
================================
Cpap set to 16 CMS - 12' hose - no breathing
Manometer at cpap end - Cpap led = 16 - Manometer says 15.5 cms
Manometer at mask end - Cpap led = 16 - Manometer says 15.0 cms
Cpap set to 16 CMS - 12' hose - breathing in
Manometer at cpap end - Cpap led = 15 - Manometer says 15 cms
Manometer at mask end - Cpap led = 15 - Manometer says 13* cms
*(this figure of 13 varies greatly depending on how I breathe - if I nose breathe it
only goes to 14, if I mouth breathe at what I believe is a slow rate (sleep) it goes
to 13, if I breathe at a modest deep rate it drops as far as 12, if I breathe really
deeply it drops to 11. I believe reflects the variation in pressure loss due to the
friction in 12' of tubing & due to faster air flow as breathed air volume increases)
Cpap set to 16 CMS - 12' hose - breathing out (modest rate)
Manometer at cpap end - Cpap led = 16 (spikes to 17 ) - Manometer says 16.0 cms
Manometer at mask end - Cpap led = 16 (spikes to 17 ) - Manometer says 15.5 cms
SUMMARY 3:
Breathing out doesn't seem to vary the manometer very much - typically only by 0.5
even if I vary the way I breathe out (faster & slower). It is breathing in that has
the most notable impact.
DSM
More tests later
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition):
hose,
tubing,
manometer,
CPAP
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition):
hose,
tubing,
manometer,
CPAP
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition):
hose,
tubing,
manometer,
CPAP