Looking for Some Help With My OSCAR Data

Post by Selectstart » Sun Nov 01, 2026 6:10 pm

Hey everyone!

I'm fairly new to OSCAR and was hoping someone more experienced could take a look at my data and offer suggestions.

I've been using a ResMed AirSense 11 AutoSet with an AirFit N30i mask. I tried CPAP several years ago but couldn't tolerate it. Several months later, I had surgery to correct a deviated septum. I was also fitted for a dental device, which I used for around two years. I tested pillow masks and found that they provide good therapy, but my nose is too sensitive. They caused irritation and cuts on my nose

I went back to my ENT for a follow-up earlier this year because I was still having sleep issues. He recommended trying CPAP again and also testing combination therapy (CPAP + dental device).

My settings are:

- Pressure: 5–20
- EPR: 2, Full Time
- Humidity: 4
- Tube temperature: 84°F

I was originally diagnosed with mild obstructive sleep apnea. I also use a mandibular advancement device (MAD) every night along with my CPAP. The MAD hasn't been adjusted since I started combination therapy.

The frustrating part is that I don't seem to be feeling any better since starting CPAP again three months ago.

I'm consistently getting 8–9 hours of usage, but I still:

- Wake up multiple times during the night.
- Often wake around 4–5 AM and stay awake for 30–60 minutes before falling back asleep.
- Feel tired and sleepy in the morning.
- Experience occasional morning headaches and brain fog.
- Sometimes need a nap during the day.

I've noticed quite a few clear-airway events showing up in my data, which is something I'm trying to understand.

Here are two recent nights:

**October 30**
- AHI: 4.94
- Clear-airway index: 4.44
- Obstructive index: 0.12
- 95% pressure: 6.70
- 95% leak: 6.0

**October 31**
- AHI: 1.87
- Clear-airway index: 1.76
- Obstructive index: 0
- 95% pressure: 6.82
- 95% leak: 2.4

MyAir has also been showing around 38 central events per night over the past week.

I did an overnight oxygen recording while using CPAP, and my average oxygen was 96%, with a low of 90% and no time below 90%.

One thing I noticed is that I tightened my N30i mask a little more last night, and it seemed to help somewhat. My leak numbers were lower, but I still had clear-airway events.

I'm wondering if anyone sees anything interesting in these OSCAR screenshots.

A few things I'd love some feedback on:

- Do the clear-airway events look unusual?
- Do my pressure settings look reasonable?
- Is EPR 2 something worth looking into?
- Does anything stand out that might explain why I'm waking up so much?

I'm also following up with my sleep doctors, but I'd really appreciate some additional eyes on my data.

Thanks!
Image
Image

User avatar
Pugsy
Posts: 65429
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Looking for Some Help With My OSCAR Data

Post by Pugsy » Sun Nov 01, 2026 6:33 pm

Did you sleep more soundly Oct 31 compared to the 30th?

I suspect that the centrals are more related to the wake ups than anything else. The machine can't tell if you are asleep or awake and often when we wake up during the night for some reason our breathing is more irregular and the machine flags awake breathing changes by mistake.
I call them false positives. LOL I see lots of them myself because I wake often from other health issues.

Look at this thread and I think you will understand what I mean.

viewtopic.php?f=1&t=187767&p=1451526#p1451526

Otherwise......do you take any medications of any kind even OTC...if so what?
And do you have any other physical or mental health issues that can or might mess with your sleep?
I may have to RISE but I refuse to SHINE.

Selectstart
Posts: 5
Joined: Sat Oct 31, 2026 5:51 pm

Re: Looking for Some Help With My OSCAR Data

Post by Selectstart » Sun Nov 01, 2026 6:48 pm

Pugsy wrote: ↑
Sun Nov 01, 2026 6:33 pm
Did you sleep more soundly Oct 31 compared to the 30th?

I suspect that the centrals are more related to the wake ups than anything else. The machine can't tell if you are asleep or awake and often when we wake up during the night for some reason our breathing is more irregular and the machine flags awake breathing changes by mistake.
I call them false positives. LOL I see lots of them myself because I wake often from other health issues.

Look at this thread and I think you will understand what I mean.

viewtopic.php?f=1&t=187767&p=1451526#p1451526

Otherwise......do you take any medications of any kind even OTC...if so what?
And do you have any other physical or mental health issues that can or might mess with your sleep?
okay thanks! I'll check out that other thread. Yes, I take Vyvanse 50mg and Lexapro 5mg for ADHD and Anxiety. I usually take it in the morning. I also take a multivitamin, omega 3, and magnesium glycinate. It did seem like I got slightly better sleep on the 31st. However, I still woke up several times.

User avatar
Pugsy
Posts: 65429
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Looking for Some Help With My OSCAR Data

Post by Pugsy » Sun Nov 01, 2026 6:57 pm

Vyvanse 50mg and Lexapro.....look up the common side effects for both ......both are well known for causing the side effects you mentioned.

I really suspect that your symptoms are more likely a side effect of poor sleep quality and not the cause of the poor sleep.
We see that often....meds causing unwanted symptoms and the wake ups are related to medications and then people see elevated central count and blame or worry about the centrals. Remember...if you aren't sound asleep anything flagged when you aren't asleep doesn't count for anything except to document poor sleep quality.
I may have to RISE but I refuse to SHINE.

Selectstart
Posts: 5
Joined: Sat Oct 31, 2026 5:51 pm

Re: Looking for Some Help With My OSCAR Data

Post by Selectstart » Sun Nov 01, 2026 7:27 pm

Pugsy wrote: ↑
Sun Nov 01, 2026 6:57 pm
Vyvanse 50mg and Lexapro.....look up the common side effects for both ......both are well known for causing the side effects you mentioned.

I really suspect that your symptoms are more likely a side effect of poor sleep quality and not the cause of the poor sleep.
We see that often....meds causing unwanted symptoms and the wake ups are related to medications and then people see elevated central count and blame or worry about the centrals. Remember...if you aren't sound asleep anything flagged when you aren't asleep doesn't count for anything except to document poor sleep quality.
Thank you, I can definitely talk with my doctor about that. I didnt realize the medications could be contributing to it.

Another note about Oct 31st, I use the Fitbit air for some additional tracking. A common theme I've noticed is getting deep sleep during the first half of the night, then the fragmented wake ups, falling back to sleep, and not getting any additional deep sleep. I had my usual fragmented night last night, but I managed to get deep sleep afterwards. 77min of deep sleep last night compared to typical 30-45min over the last 30 days.

Selectstart
Posts: 5
Joined: Sat Oct 31, 2026 5:51 pm

Re: Looking for Some Help With My OSCAR Data

Post by Selectstart » Tue Nov 03, 2026 1:22 pm

Pugsy wrote: ↑
Sun Nov 01, 2026 6:57 pm
Vyvanse 50mg and Lexapro.....look up the common side effects for both ......both are well known for causing the side effects you mentioned.

I really suspect that your symptoms are more likely a side effect of poor sleep quality and not the cause of the poor sleep.
We see that often....meds causing unwanted symptoms and the wake ups are related to medications and then people see elevated central count and blame or worry about the centrals. Remember...if you aren't sound asleep anything flagged when you aren't asleep doesn't count for anything except to document poor sleep quality.

Update, tightening my mask and taking my medication as soon as possible after I wake up has made a difference in my sleep quality and results. Thanks for pointing that out. I also slightly adjusted my humidity and tube temp from 4 to 3 and 84 to 82.

Here's last night:

Image

User avatar
Pugsy
Posts: 65429
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Looking for Some Help With My OSCAR Data

Post by Pugsy » Tue Nov 03, 2026 5:23 pm

If you are sleeping a little better and feeling a little better then that's a big positive. I still think that the centrals you see flagged are very likely related to wake ups and we may just have to accept that you will see a few of those.
Even if they were real asleep centrals the doctors don't get excited until you start having over 5 per hour all night long because an occasional real asleep central is actually normal every now and then. There are what we call sleep onset or sleep transition centrals and aren't of concern unless you have a truckload of them bouncing you out of sleep.

Now you take a big patience pill and give it some time. :lol: :lol:
I may have to RISE but I refuse to SHINE.

Selectstart
Posts: 5
Joined: Sat Oct 31, 2026 5:51 pm

Re: Looking for Some Help With My OSCAR Data

Post by Selectstart » Thu Nov 05, 2026 8:34 am

Pugsy wrote: ↑
Tue Nov 03, 2026 5:23 pm
If you are sleeping a little better and feeling a little better then that's a big positive. I still think that the centrals you see flagged are very likely related to wake ups and we may just have to accept that you will see a few of those.
Even if they were real asleep centrals the doctors don't get excited until you start having over 5 per hour all night long because an occasional real asleep central is actually normal every now and then. There are what we call sleep onset or sleep transition centrals and aren't of concern unless you have a truckload of them bouncing you out of sleep.

Now you take a big patience pill and give it some time. :lol: :lol:
Yes, I agree! The sleep has been much better over the last few days. I am hopeful that these changes have me back on track. Even if I wake up, I'm getting better sleep overall and waking up feeling more refreshed. Thank you again for taking the time to review and troubleshoot. My sleep doctor was thinking of scheduling a titration study but now that im getting better results, we are going to put that on hold and stick with this for now.

An optimization test I'm considering is lowering EPR from 2 to 1. My 95% is around 6.22 and breathing out has me at 4.22. Even though OA is 0 like this, I wonder if it could possibly help things on the CA side.

User avatar
Pugsy
Posts: 65429
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Looking for Some Help With My OSCAR Data

Post by Pugsy » Thu Nov 05, 2026 10:15 am

It may be worth trying at some point....just remember it is unlikely to make much of a change in CA events UNLESS the EPR happens to be related to the use of EPR (how much exhale relief) itself. Some people see a reduction of CA events with a reduction in EPR and some people don't. EPR use can trigger centrals in some people but the number of people that it triggers centrals is actually very low. It's not near as common as people might think.
If you try reducing (or even turning EPR off) and you don't see much of a change in central events then you know that the use of EPR isn't causing centrals.
Also remember that a few centrals are actually normal and you just have to accept it they are real asleep centrals.
I may have to RISE but I refuse to SHINE.