Has anyone figured out how to make an AirSense 11 stop popping with every breath?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Lyndmc
Posts: 25
Joined: Sun Apr 22, 2018 2:41 pm
Location: Finleyville, Pa

Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Lyndmc » Fri Jan 30, 2026 5:54 pm

My AirSense 11 is making me crazy! I called Resmed, they sent me to my equipment supplier, my supplier said yeh, they do that! 😡

So far I’ve replaced the tank, I’ve removed and cleaned the outlet seal, I’ve cleaned the mating surfaces in front of the tank, and I’ve cleaned the hose. I’m waking up 2-3 times a night, pulling the tank loose, and putting it back in. Sometimes that fixes it for the night, mostly not. If I push down on the top of the tank, it usually stops clicking, but that doesn’t seem to work as well as pulling the tank loose.

I’m really frustrated with this thing! Any suggestions?

User avatar
tlohse
Posts: 267
Joined: Fri Jun 27, 2014 5:32 pm
Location: West Bend, Wisconsin
Contact:

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by tlohse » Fri Jan 30, 2026 8:22 pm

Hmm. I never had experience with a AirSense 11 as I have a 10 currently. I suppose you could change out parts. I have really have no idea.

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack
Additional Comments: Machine is Autoset.
Thomas Lohse
Machine ResMed AirSense 11 Auto Set
Mask: ResMed AirFit N30i Standard Frame with Medium cushion
CPAP USER SINCE JUNE 2013

Lyndmc
Posts: 25
Joined: Sun Apr 22, 2018 2:41 pm
Location: Finleyville, Pa

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Lyndmc » Fri Jan 30, 2026 10:15 pm

I also have an AirSense 10 and it works great. It uses a lot less water and it doesn’t make that noise. I currently use that unit as my traveling in-unit.

User avatar
LSAT
Posts: 13410
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by LSAT » Sat Jan 31, 2026 9:18 am

I don't understand the posts telling about the Airsense 10 when the poster explicitly asked a question about his Airsense 11.

Sleepzilla
Posts: 64
Joined: Fri Nov 28, 2025 3:39 pm

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Sleepzilla » Sat Jan 31, 2026 2:09 pm

.
Last edited by Sleepzilla on Tue Feb 10, 2026 8:49 am, edited 2 times in total.

Lyndmc
Posts: 25
Joined: Sun Apr 22, 2018 2:41 pm
Location: Finleyville, Pa

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Lyndmc » Sat Jan 31, 2026 2:32 pm

As an update, and this is very preliminary, I put a rubber band around the water tank way out on the end away from the heater pad. I figured if I can squeeze the tank while in operation and the popping stops, that may be where the pop is coming from.

After night 1 with the rubber band, two times last night I heard 1 single pop, and honestly, I slept better. Tonight I have 2 rubber bands around the tank. This tank isn’t very old, and I actually swapped the tank because of the pop.

Here’s what I think is happening. When I slam my tank in, it tightly clamps the two halves of the tank down, you can’t squeeze it and get any flex. Over the evening either air pressure, especially if I sneeze, the tank backs out ever so slightly. This allows the slightest spring in the tank gasket that you can notice when you squeeze the two halves of the tank. Breathing in and out may be pressuring and then depressurizing the tank. As the tank pressure drops, the machine ramps up and inflates the tank making a pop. If I’m correct, with 2 rubber banks installed, I should have zero pops tonight. I will update soon.

Lyndmc
Posts: 25
Joined: Sun Apr 22, 2018 2:41 pm
Location: Finleyville, Pa

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Lyndmc » Sun Feb 01, 2026 11:45 am

2 nights in and no more popping! My rubber band trick seems to be working! It has to be the 2 halves of the tank popping as the pressure changes.
Attachments
IMG_2363.jpeg
IMG_2363.jpeg (851.6 KiB) Viewed 13950 times

Pachetto
Posts: 3
Joined: Sat Jul 25, 2026 1:45 am

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Pachetto » Sat Jul 25, 2026 2:18 am

I have used a Yuwell CPAP for a while and I experience poppings especially in the morning, they are so disruptive of my sleep that I ceased using the machine for a while. I have gone to my sleep doctor who did not solve the problem and just said that the machine was working well, which it does not.
I am intrigued by the fact that this problem could be caused by the tank. Actually my Yuwell looks very similar to the Resmed shown in the picture. So I will try the solution suggested and then post an update on the results.

Pachetto
Posts: 3
Joined: Sat Jul 25, 2026 1:45 am

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Pachetto » Sun Jul 26, 2026 1:09 am

Wrapping the water tank with some tape seems to have solved the popping problem that has bothered me for so many months! I sincerely thank the person who has found out this simple and effective solution. Anyway I will go to the seller of my machine and pretend that he changes it since it is obviously faulty.

User avatar
ChicagoGranny
Posts: 15527
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by ChicagoGranny » Tue Jul 28, 2026 2:37 pm

Pachetto wrote:
Sat Jul 25, 2026 2:18 am
I have used a Yuwell CPAP for a while and I experience poppings especially in the morning, they are so disruptive of my sleep that I ceased using the machine for a while. I have gone to my sleep doctor who did not solve the problem and just said that the machine was working well, which it does not.
I am intrigued by the fact that this problem could be caused by the tank. Actually my Yuwell looks very similar to the Resmed shown in the picture. So I will try the solution suggested and then post an update on the results.
Water (condensation from the humidifier) can cause a popping sound when it builds up in the hose. Do you need some tips on how to prevent this? (It's called rainout.)
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

Pachetto
Posts: 3
Joined: Sat Jul 25, 2026 1:45 am

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Pachetto » Wed Jul 29, 2026 6:34 am

I don't put water in the water tank. When I did it it popped in the morning. I then put the machine on the floor and it stopped, but it continued to pop from time to time and now that I have seen the post by the guy who had the same problem and realized that it was caused by the tank I have adopted the same solution with good results.

User avatar
ChicagoGranny
Posts: 15527
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by ChicagoGranny » Wed Jul 29, 2026 11:23 am

Pachetto wrote:
Wed Jul 29, 2026 6:34 am
I don't put water in the water tank.
Neither do I. This side cover is nice:
The ResMed AirSense 11 side cover (part number 39226, priced around $18.40) replaces the water chamber to let you run your machine without humidification. It automatically disables the heater, reduces bulk for travel, and keeps dust out of the internal port.
When traveling, I like that my machine is smaller.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

ozij
Posts: 10542
Joined: Fri Mar 18, 2005 11:52 pm

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by ozij » Sat Aug 01, 2026 1:45 am

Anyone who doesn't put water in the humidifier should just use the side cover.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Sleepzilla
Posts: 64
Joined: Fri Nov 28, 2025 3:39 pm

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by Sleepzilla » Sat Aug 01, 2026 5:10 pm

ozij wrote:
Sat Aug 01, 2026 1:45 am
Anyone who doesn't put water in the humidifier should just use the side cover.
Why can't everyone just do whatever they want?

The humidifier water tank is included with the machine at no incremental cost.
The optional side cover must be purchased for ~$20 USD.
When switching between humidified and non-humidified therapy, such as when travelling, you can do both with the humidifier tank without having to carry a side cover with you too.
There's no difference in therapy efficacy or experience when either are used with humidification disabled in the machine settings.
The side cover shortens the Air10 length by 20% (2" or 55mm) and the Air11 length by a mere 9% (0.9" or 23mm.)

I have a side cover for my Air10, which I sometimes use on long-haul flights where available space to set up my machine is limited and humidifier use is discouraged due to the possibility of air turbulence etc. while I'm sleeping. Sure, the slightly smaller machine footprint is more convenient in that situation. But it's not exactly a game-changer. <shrugs>

alancalan
Posts: 169
Joined: Wed Dec 27, 2006 4:27 am

Re: Has anyone figured out how to make an AirSense 11 stop popping with every breath?

Post by alancalan » Sun Aug 02, 2026 11:52 am

Edited, dictation is dangerous.

I had the same problem with popping and I thought there was a leak someplace. Then my regular internist asked me where the aircurve 11 is in proximity to my head. I told him it was behind me on a shelf that's about 6 inches above my head. He said, the directions, right in the beginning of the manual, say your device should be below your head. Then he said, I should put it on a night table. Since then I never had another problem with popping and I fill the water compartment full every night.

_________________
MachineMaskHumidifier
Additional Comments: & P30i
Last edited by alancalan on Thu Aug 06, 2026 10:19 am, edited 1 time in total.