Krickets?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Sun Jan 25, 2026 3:46 am

Anyway, it looks like this is the Patent Application Publication:

https://patentimages.storage.googleapis ... 2733A1.pdf

So it would appear that the system will have a VCom in the circuit with additional IPAP drop from the machine:
krick2.jpg
krick2.jpg (25.68 KiB) Viewed 2429 times
I wonder if these reduced flows affect the mask intentional leak rates, and if so, that's why TECSA is reduced? Little more CO2 floating around there?

Seems to me all the mask companies are watching this closely. Maybe SleepRes will have their own mask line as all the other companies won't support their products if used with a Kricket.

Or maybe they thought about that already and have Plan B.

At any rate, I would assume having a VCom in the circuit is the only way to do this as a machine on it's own wouldn't be able to identify and react to an inspiration fast enough.

Again, now all of a sudden the "perfect" 2.0 cmH2O IPAP drop isn't good enough?

Why not offer a Variety Pak of VComs with different pressure drops so users could pick what works best for them? They could raise the price a little for like a 4-pack (cause most of the cost of an inexpensive product is in the packaging and getting to end user).

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Sun Jan 25, 2026 3:59 am

Nocibur wrote:
Sun Jan 25, 2026 3:46 am

At any rate, I would assume having a VCom in the circuit is the only way to do this as a machine on it's own wouldn't be able to identify and react to an inspiration fast enough.
IOWs, Kricket is a workaround, and there may be a more efficient IPAPreduction algorithm in the works.

By the other companies.

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Sun Jan 25, 2026 4:22 am

But IMO there's a question that needs to be answered before KPAP takes over the planet.

If KPAP is a comfort modality; and

comfort only applies to when you're awake; and

specifically, the hypothesis is that lower inspiratory pressures are more comfortable; then

why not just mash the Ramp Button?

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Sun Jan 25, 2026 4:35 am

McSleepy wrote:
Fri Jan 23, 2026 5:10 pm
... but how about bilevel ...
Or that.

If it turns out that IPAPreduction is the next Big Thing IIWM I'd go back and dig up the old Proportional Assist Ventilation algorithms.

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Sun Jan 25, 2026 4:38 am

Or fiddle with Rise Time.

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Sun Jan 25, 2026 4:45 am

Or run a comparison of VCom alone to Kricket.

McSleepy
Posts: 499
Joined: Fri Nov 26, 2010 10:32 pm
Location: USA

Re: Krickets?

Post by McSleepy » Sun Jan 25, 2026 3:36 pm

Boy, I wish I had as much time for this as you do! (Truly! And respectfully! :) ) I'll just say this:

1) The entire purpose of bi-level CPAP is to improve patient compliance. (Unless you have a different idea about what "compliance" is.) And it has done that admirably. It may not apply to all patients, and it may only help a smaller percentage of CPAP users, but that doesn't mean it doesn't hold the key to addressing the general problem. What is a fact, though, is that bi-levels are more expensive, and more complex, as they need to cover the needs of a smaller population (e.g., higher pressures). It saved my life - the fact that 23 years ago, a pulmonologist knew enough to switch my constant-level CPAP to a bi-level. And that the particular machine he prescribed (an early Respironics BiPAP Pro with Bi-Flex) happened to have some hardcoded algorithms that suited my needs. "Compliance" or something else, it made the difference between me abandoning CPAP and not being around by now, and being a vigorous, healthy individual today.

2) I love simple arithmetic, but it doesn't tell the whole story when it comes to the complex physiology and psychology of a sapient human. Vital capacity is more than the sum of those three volumes, their proportionality to the total notwithstanding (and good enough argument by itself). One aspect of it is that - yes, tidal volume is important, but when that one is larger than usual (most people), then the CPAP machine has a harder task in the case of people with larger such capacity (and I've posted about that recently). Just like the typical CPAP user has a hard time explaining to someone without OSA why the machine is so important to them, the non-typical OSA+UARS+Central+WhoKnowsWhatElse patient has hard times explaining why such a convoluted airflow requirements is so crucial to them. More on that in the next paragraph.

3) I'll try to explain my situation, without having any idea what percentage of CPAP users are like me (and to what extent). Let's start with the simple case: soft tissue in the pharynx collapses when lying down and unconscious, so we apply some continuous pressure to keep it open - case closed! If the airways are relatively wide, and the soft tissue is "compliant" enough to be easily pushed back (i.e., with lower pressures), then it is indeed that simple. But, how about if the airways are really narrow and tight (to take it in a sequential order - from machine to lungs)? Now the pressure in the pharynx (which is all you care about) will differ significantly from the pressure at the source (where the machine sensor is) - simple Bernoulli's principle. How does the CPAP account for that? It has no sensors inside our body to close a feedback loop, so - heuristics? Or - not at all. Then, what do we do? Increase the pressure. With all of its downsides. It's where the bi-level comes, but not only. I was not able to use the Respironics System One bi-level. The reason - its gradual transition from EPAP to IPAP. Luckily, in some ResMed machines, it's adjustable (rise time). And here is where most people are shocked: to me, the shorter the rise time - the better; it's more comfortable. I don't need reduction of pressure as it rises, I need an increase! And before you think that I'm a rare animal - there's a reason those machines are designed with the ability to reduce rise time to as fast as the blower will go - they knew something! Now, as to why that is - here is where vital capacity comes in. Years of untreated OSA, while being a big, athletic man, has trained my mind to have the ability, to need to "know" that if it asks for enough air to fill the full capacity - regardless of the fact that while sleeping (or trying to!) I "should" only be aiming to fill my tidal volume - it can do that. My data shows exactly that - flow increases significantly as I'm trying to get extra air, and then... I awake. It may be at the root of all UARS cases, but it certainty is in mine. So, that's what I meant when saying that I wish there weas a CPAP machine that takes into account the pressure drop after all restrictions, and the flow demand of the patient. It is impossible to account for all that just by looking at the pressure sensor inside the machine, but pending some invention where we can place pressure sensors and flow meters inside our airways, we may just need some powerful heuristic algorithms; and more adjustability by the patients (which we all know how much the CPAP manufacturers love doing!)

4) You are obviously a very smart and compassionate person, and I don't mean to ask you to solve any problems of mine, or in general. I know I am not providing nearly enough detail and information, so - please, don't worry about it. Maybe one day I'll have enough time to actually delve deep into this, but knowing that laws and regulations will never allow CPAP to be open-srouced, I don't know that I would. But maybe bringing some awareness is good enough at this time?

McSleepy

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Previous machine: ResMed S9 VPAP Auto 25 BiLevel. Mask: Breeze with dilator pillows. Software: ResScan ver. 5.1
ResMed AirCurve 10 VAuto; Puritan-Bennett Breeze nasal pillow mask; healthy, active, middle-aged man; tall, athletic build; stomach sleeper; on CPAP since 2003; lives @ 5000 ft; surgically-corrected deviated septum and turbinates; regular nasal washes

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Sun Jan 25, 2026 4:16 pm

McSleepy wrote:
Sun Jan 25, 2026 3:36 pm
Boy, I wish I had as much time for this as you do!
Funny you should say that...

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Mon Jan 26, 2026 2:31 am

McSleepy wrote:
Sun Jan 25, 2026 3:36 pm
I love simple arithmetic
Great, I just happen to have a formula to fiddle with.
people with larger such capacity (and I've posted about that recently).
Yeah I remember. The minute ventilation thing. The vital capacity of ~7.0 liters.

So back to our formula. Many labs use Cherniack 1972 to calculate pulmonary function. he says:
VC in liters for White males 15-79 years () = (0.12102 * (height in inches)) - (0.01357 * (age in years)) - 3.18373
Given your love of adds, subtracts and gazintas, you'll recognize the key component of Cherniak (and all the prediction formulas) is height.

Anyway, working the prediction formula backwards, in order to have a VC of ~7.0 liters:

vital.jpg
vital.jpg (33.55 KiB) Viewed 2280 times
you'd have to be about 7'6" tall.
You are obviously a very smart and compassionate person
Smart?

I get by.

Compassionate?

Nope.

I'm actually a misanthrope.

But I'm certainly smart enough to not FAFO with Godzilla.

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Mon Jan 26, 2026 4:21 am

McSleepy wrote:
Sun Jan 25, 2026 3:36 pm
- simple Bernoulli's principle.
Yeah I saw how "simple" the Bernoulli Principle can be in your exchange with PR.

I am amused by the irony of that exchange in that Bernoulli is a "Conservation of Energy" principle.

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Mon Jan 26, 2026 4:27 am

Nocibur wrote:
Sun Jan 25, 2026 4:16 pm
McSleepy wrote:
Sun Jan 25, 2026 3:36 pm
Boy, I wish I had as much time for this as you do!
Funny you should say that...
...cause doing "this" is how I kept kishka on the table.

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Mon Jan 26, 2026 4:36 am

I'm also an ASPS, and I tell ya, DST can't get here fast enough for me...

McSleepy
Posts: 499
Joined: Fri Nov 26, 2010 10:32 pm
Location: USA

Re: Krickets?

Post by McSleepy » Mon Jan 26, 2026 1:30 pm

You definitely have too much time on your hands! :lol: So, if you're a physician, you certainly would understand that:

Image

I've measured higher.

Maybe that'll shake your confidence in "simple arithmetic" (sorry you missed the sarcasm!) and Internet "wisdom", and trusting that guy or another?
I made (and still do) my "kishka" by being educated, but also by practicing critical thinking (and not doubting other people's knowledge). I was hoping you'd be a good resource since you seem to stay on top of new technologies, but if you're going to be a self-professed "misanthrope", and aspire to be PL's replacement here, then I have nothing further, your honor.

McSleepy

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Previous machine: ResMed S9 VPAP Auto 25 BiLevel. Mask: Breeze with dilator pillows. Software: ResScan ver. 5.1
ResMed AirCurve 10 VAuto; Puritan-Bennett Breeze nasal pillow mask; healthy, active, middle-aged man; tall, athletic build; stomach sleeper; on CPAP since 2003; lives @ 5000 ft; surgically-corrected deviated septum and turbinates; regular nasal washes

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Mon Jan 26, 2026 2:17 pm

McSleepy wrote:
Mon Jan 26, 2026 1:30 pm
... and aspire to be PL's replacement here...
RUFKM?

I was the resident misanthrope LONG before PR made his initial appearance.
I was hoping you'd be a good resource
No wait! Please stay!

I was really looking forward to serving you!

User avatar
Nocibur
Posts: 429
Joined: Sat Jan 25, 2025 6:08 am

Re: Krickets?

Post by Nocibur » Tue Jan 27, 2026 4:42 am

McSleepy wrote:
Mon Jan 26, 2026 1:30 pm
you certainly would understand that:
I do.

I need to comment in case posting that (old) PFT was a test to see if I would say

"HF!! You breathe with that PFT!?"