Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
-
Charlieboy UK
- Posts: 45
- Joined: Wed Jan 07, 2026 10:39 am
Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
I thought a RERA (Respiratory Effort-Related Arousal) was when you have 'subtler limitations in breathing' of 10 seconds or more... (which can cause significant sleep fragmentation if you have a lot of them in a night) ... but OSCAR and SleepHQ start recording them at under 10 secs... mine are mostly 7 secs, and on this night I only had one over 10 secs. https://sleephq.com/public/76a358e7-790 ... 9be041822c
- Attachments
-
- Screenshot 2026-01-10 150621.png (11.54 KiB) Viewed 1414 times
- Dog Slobber
- Posts: 4283
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
No.
Apneas and Hypopneas need to last 10 seconds to be marked.
The length of the disturbed breathing that caused the RERA, isn't really significant. An arousal is an arousal.
Apneas and Hypopneas need to last 10 seconds to be marked.
The length of the disturbed breathing that caused the RERA, isn't really significant. An arousal is an arousal.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
-
Sleepzilla
- Posts: 16
- Joined: Fri Nov 28, 2025 3:39 pm
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
RERA events are typically defined as lasting longer than 10 seconds, similar to apneas and hypopneas. (A quick internet search for "RERA definition" shows this repeatedly, without notable exception.)
Unlike apneas or hypopneas, RERAs are uniquely categorized as OSA-related upper airway breathing disruptions that manifest as the combination of:
1) >10s period of decreased nasal breathing combined with increased respiratory effort, and
2) Absence of significant oxygen desaturation, and
3) Result in brief arousals from sleep, which contribute to sleep fragmentation.
The only way to accurately detect RERAs is via sleep lab polysomnography that monitors respiration, oximetry, and brain activity. (The latter being required to detect the actual arousal.)
Some PAP machine, like the ResMed AirSense series, will imply and flag RERA events by monitoring flow via an algorithm for this purpose. Per the Apnea Board RDI wiki, the premise for ResMed devices is that the "RERA is flagged when increasing flow limitation or flow reduction is followed by an increased respiratory flow rate suggesting an arousal or recovery breathing."
Unlike apneas or hypopneas, RERAs are uniquely categorized as OSA-related upper airway breathing disruptions that manifest as the combination of:
1) >10s period of decreased nasal breathing combined with increased respiratory effort, and
2) Absence of significant oxygen desaturation, and
3) Result in brief arousals from sleep, which contribute to sleep fragmentation.
The only way to accurately detect RERAs is via sleep lab polysomnography that monitors respiration, oximetry, and brain activity. (The latter being required to detect the actual arousal.)
Some PAP machine, like the ResMed AirSense series, will imply and flag RERA events by monitoring flow via an algorithm for this purpose. Per the Apnea Board RDI wiki, the premise for ResMed devices is that the "RERA is flagged when increasing flow limitation or flow reduction is followed by an increased respiratory flow rate suggesting an arousal or recovery breathing."
OSCAR and SleepHQ don't record anything. They simply report what the therapy machine has recorded and stored in its memory.Charlieboy UK wrote: ↑Sat Jan 10, 2026 9:09 am... but OSCAR and SleepHQ start recording them at under 10 secs... mine are mostly 7 secs, and on this night I only had one over 10 secs.
-
Charlieboy UK
- Posts: 45
- Joined: Wed Jan 07, 2026 10:39 am
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
I thought they just took the RAW data... but.. whichever way, it does seem to be a bit silly to even show it, if its under 10 secs... it doesn't even seem to really mean much...I guess we can over analyse this stuff.. if AHI is under 5, Sp02 over 95% all night and not getting up in the night, with minimal movement shown when sleeping.. they seem to be the key indicators that PAP is treating you ok.Sleepzilla wrote: ↑Sat Jan 10, 2026 10:40 amRERA events are typically defined as lasting longer than 10 seconds, similar to apneas and hypopneas. (A quick internet search for "RERA definition" shows this repeatedly, without notable exception.)
Unlike apneas or hypopneas, RERAs are uniquely categorized as OSA-related upper airway breathing disruptions that manifest as the combination of:
1) >10s period of decreased nasal breathing combined with increased respiratory effort, and
2) Absence of significant oxygen desaturation, and
3) Result in brief arousals from sleep, which contribute to sleep fragmentation.
The only way to accurately detect RERAs is via sleep lab polysomnography that monitors respiration, oximetry, and brain activity. (The latter being required to detect the actual arousal.)
Some PAP machine, like the ResMed AirSense series, will imply and flag RERA events by monitoring flow via an algorithm for this purpose. Per the Apnea Board RDI wiki, the premise for ResMed devices is that the "RERA is flagged when increasing flow limitation or flow reduction is followed by an increased respiratory flow rate suggesting an arousal or recovery breathing."
OSCAR and SleepHQ don't record anything. They simply report what the therapy machine has recorded and stored in its memory.Charlieboy UK wrote: ↑Sat Jan 10, 2026 9:09 am... but OSCAR and SleepHQ start recording them at under 10 secs... mine are mostly 7 secs, and on this night I only had one over 10 secs.
-
Sleepzilla
- Posts: 16
- Joined: Fri Nov 28, 2025 3:39 pm
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
With respect, I think you're kind of overlooking some key concepts with this line of thinking.Charlieboy UK wrote: ↑Sat Jan 10, 2026 4:37 pmI thought they just took the RAW data... but.. whichever way, it does seem to be a bit silly to even show it, if its under 10 secs... it doesn't even seem to really mean much...I guess we can over analyse this stuff.. if AHI is under 5, Sp02 over 95% all night and not getting up in the night, with minimal movement shown when sleeping.. they seem to be the key indicators that PAP is treating you ok.
AHI is just a simple metric that summarizes the frequency—not the duration or severity—of apnea and hypopnea events on a per-hour basis. There's no direct correlation between AHI scores and arousals and/or quality of sleep. Put differently: Plenty of people have low AHI scores and still suffer from poor sleep due to myriad other non-apnea related reasons, including RERAs.
RERAs on the other hand, are not factored into AHI calculation and include arousal from sleep as a criteria of their definition. This regardless of their duration. So people with low AHI and still poor sleep quality have much reason to be interested in RERA events... even if they do last less than 10 seconds. (Same for coughing spouses, slamming doors, fireworks, gunshots, barking dogs, and kids driving by with loud exhausts etc.)
-
Charlieboy UK
- Posts: 45
- Joined: Wed Jan 07, 2026 10:39 am
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
Well I'm certainly no expert.. I do feel better using my PAP machine, even though I get a number of RERA's...i guess you cant really tell without a proper sleep study, so its more of an indicator to the Doctor, like FL's, eFL, eMO, eS and eSO...of what might be going on... I do think its a bit odd though, when it says for example I have had 8 Hypopnea events.. which should of course mean my Oxygen drops.. but it doesn't, well not below 95% and I use the 02 Ring from Viatom so I know its accurate. Its very complex all this data. Our NHS in the UK only seem to look at AHI and leaks and of course how long you use it for.Sleepzilla wrote: ↑Sat Jan 10, 2026 5:11 pmWith respect, I think you're kind of overlooking some key concepts with this line of thinking.Charlieboy UK wrote: ↑Sat Jan 10, 2026 4:37 pmI thought they just took the RAW data... but.. whichever way, it does seem to be a bit silly to even show it, if its under 10 secs... it doesn't even seem to really mean much...I guess we can over analyse this stuff.. if AHI is under 5, Sp02 over 95% all night and not getting up in the night, with minimal movement shown when sleeping.. they seem to be the key indicators that PAP is treating you ok.
AHI is just a simple metric that summarizes the frequency—not the duration or severity—of apnea and hypopnea events on a per-hour basis. There's no direct correlation between AHI scores and arousals and/or quality of sleep. Put differently: Plenty of people have low AHI scores and still suffer from poor sleep due to myriad other non-apnea related reasons, including RERAs.
RERAs on the other hand, are not factored into AHI calculation and include arousal from sleep as a criteria of their definition. This regardless of their duration. So people with low AHI and still poor sleep quality have much reason to be interested in RERA events... even if they do last less than 10 seconds. (Same for coughing spouses, slamming doors, fireworks, gunshots, barking dogs, and kids driving by with loud exhausts etc.)
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
believe me, the same is true here in the former land of the free. and then they say an ahi of 5 is fine. here in the forum, we look at an ahi of 2 to be fine.Charlieboy UK wrote: ↑Sat Jan 10, 2026 6:06 pmOur NHS in the UK only seem to look at AHI and leaks and of course how long you use it for.
whether UK or US, doctors pretty much just want to make their boat payments.
/off soapbox
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
Sure you can. AAMOF, your last SleepHQ upload shows no RERAs. The machine is lying.Charlieboy UK wrote: ↑Sat Jan 10, 2026 6:06 pmI get a number of RERA's...i guess you cant really tell without a proper sleep study
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
Just like it lied to you about SoftPAP.Nocibur wrote: ↑Sun Jan 11, 2026 3:30 amSure you can. AAMOF, your last SleepHQ upload shows no RERAs. The machine is lying.Charlieboy UK wrote: ↑Sat Jan 10, 2026 6:06 pmI get a number of RERA's...i guess you cant really tell without a proper sleep study
That is one uppity machine. I think you need to KIFA.
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
However, in response to your question:
It appears Lowenstein has elected to tag just the "A" instead of the "RE-R" and "A" together in scoring RERAs.Charlieboy UK wrote: ↑Sat Jan 10, 2026 9:09 amI thought a RERA (Respiratory Effort-Related Arousal) was when you have 'subtler limitations in breathing' of 10 seconds or more... but...mine are mostly 7 secs...
-
Charlieboy UK
- Posts: 45
- Joined: Wed Jan 07, 2026 10:39 am
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
yep it does seem odd, that it is tagging RARA under 10 secs.... I rather like the machine though, the humidifier is really easy to clean, its quiet and easy to change settings etc good old German engineering ...I guess SoftPAP is really meant for higher pressures, as mine is low it didn't really do anything.. apart from causing loads of flow limitations, although they didn't seem to impact AHI etc.Nocibur wrote: ↑Sun Jan 11, 2026 4:24 amHowever, in response to your question:
It appears Lowenstein has elected to tag just the "A" instead of the "RE-R" and "A" together in scoring RERAs.Charlieboy UK wrote: ↑Sat Jan 10, 2026 9:09 amI thought a RERA (Respiratory Effort-Related Arousal) was when you have 'subtler limitations in breathing' of 10 seconds or more... but...mine are mostly 7 secs...
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
A deep dive would be necessary, but I think there's a good chance SoftPAP simply engineers breaths to make them look like FLs, or it occasionally intervenes prematurely. An expiratory adjunct should have NO influence on inspiration.Charlieboy UK wrote: ↑Sun Jan 11, 2026 5:40 am...(SoftPAP causes) loads of flow limitations, although they didn't seem to impact AHI etc.
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
Or somehow SoftPAP FsU the algorithm somehow. I mean, I'm looking at these
and thinking
"On what planet are those FLs??"
and thinking
"On what planet are those FLs??"
-
Charlieboy UK
- Posts: 45
- Joined: Wed Jan 07, 2026 10:39 am
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
well its interesting that since turning off SofatPAP I have had no FLs...but as you say, they were clearly not real FLs anyway.. I think it might simply be the way OSCAR is built...because my understanding is they just take the raw data, really need to see what Lowenstein software says which I guess my hospital has. I only have the app, which just looks at AHI, leaks, deep sleep and total sleep time. My App said i had an AHI today of 3 but OSCAR says almost 5 and the app said no leaks...guess because they are small.
- Attachments
-
- Screenshot 2026-01-11 121711.png (259.74 KiB) Viewed 1201 times
Re: Why does Oscar and SleepHQ hightlight RERA lower than 10 secs?
Something else to consider.
These are from your current D/L:
RERAs need FL reduction to be scored RERAs.
So how did Prisma score RERAs w/o scoring FL?
AAMOF, the areas preceding these RERAs look more FL-y than the FLs scored during SoftPAP!!
Like WTF??
These are from your current D/L:
RERAs need FL reduction to be scored RERAs.
So how did Prisma score RERAs w/o scoring FL?
AAMOF, the areas preceding these RERAs look more FL-y than the FLs scored during SoftPAP!!
Like WTF??