S9 hard to breathe out?
S9 hard to breathe out?
One of my S9 units is set as my others but it is hard to breathe out. Any sugestions, is this comon?
- Dog Slobber
- Posts: 4158
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: S9 hard to breathe out?
Verify that it actually is set identically to the others, including EPR.
You can compare it's pressure to your other devices by building a simple manometer. Essentially, putting the end of the tube into a deep water bath, turning it on and then measureing how far the tube goes down before the air stop coming out. if the devices are all set the same, they should all measure the same.
You can compare it's pressure to your other devices by building a simple manometer. Essentially, putting the end of the tube into a deep water bath, turning it on and then measureing how far the tube goes down before the air stop coming out. if the devices are all set the same, they should all measure the same.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |