alemon wrote: ↑Tue Sep 24, 2024 11:16 am
After a lot of struggling and bad nights again, I booked a consult with a respiratory therapist and they recommended bilevel. I've done a great job of getting my AHI events relatively low, but was still having very poor sleep. I'm giving bilevel a try, and having a few issues, mostly around aerophagia - my HR values look much better on this, though. Previously I was on CPAP-EPR-3 at 9.8cms, and that seemed to be doing all right, I've been trying to translate that over to bilevel.
What are HR values????
Sorry to hear about the aerophagia. More on that in a bit.
This data is for Sept. 21. Therapy settings are Min EPAP = 9.0, Max IPAP = 15, PS = 4.8. All those are significantly higher than your old CPAP settings of fixed pressure at 9.8 and EPR = 3. How did you decide to use these settings?
Given the drastic increase in pressure settings it's no wonder you gave yourself terrible aerophagia. And the aerophagia all by itself would explain the frequency of the spontaneous arousals during the night.
Yesterday I'd tried 3.8/8.6-14, just trying to cut down on the pressures a little, again woken by terrible aerophagia in the early AM and swapped back to the CPAP - for that, I felt pretty good! I slept past my alarm completely.
https://sleephq.com/account/teams/vVNLB ... _id=JzoXBj
This data won't load for me. So I can't say anything about it. If I interpret your statement of "I'd tried 3.8/8.6-14", I would guess that you mean you tried Min EPAP = 8.6, Max IPAP = 14, PS = 3.8. Still way higher than your old settings on the CPAP were. Again, why did you decide to make such a drastic jump in your pressure settings when you went from plain CPAP at 9.8 with EPR = 3 to these settings on the Vauto?
Today I tried cutting the settings down, 7EPAP 11 IPAP PS 3. I definitely slept quite poorly, even though my numbers look great, with tons of wakeups in the early morning - not sure if it was pressure swings or if it was breathing, but I know pressure swings were wrecking my sleep on APAP, so I'm starting to suspect the same. I do not feel nearly as rested - though of course, I recognize it's the first full night. Still seeing lots of that periodic breathing.
*Addendum, I saw it recommended I not put on the mask til I'm about to sleep, and by doing so last night I kept the pressure from swinging way high to try to fix my awake-breathing before I was sleeping. That also may have helped.
https://sleephq.com/public/651c85b6-52a ... e6ca2e5002
This data is from Sept. 23, but it still represents a pressure increase from your APAP settings of min pressure = max pressure = 9.8 with EPR = 3.
Yes, the pressure swings do seem to be triggering a lot of spontaneous arousals. What's triggering those pressure swings are breathing patterns the machine has decided are "flow limitations". But some of those "flow limitations" just might be part of your normal sleep transitional breathing. In other words, as your brain tried to had the control of the respiration off to the autonomous part of the nervous system and change the CO2 trigger for "inhale now", the machine took the change in flow rate to be a "flow limitation" and bumped the pressure up, which then woke you back up.
So you might be better off using fixed bi-level pressures rather than an auto bi-level range. There are several ways to do that with your current AirCurve VAuto. More on that in a bit.
What you are thinking is "periodic breathing" is not significant enough to worry about in my opinion. It's not being labeled that way by the machine because the variation between the largest inhalations and the smallest inhalations is not large enough to be true periodic breathing.
In other words, what you still need to do is to fix the spontaneous arousals. And that's a whole lot harder than just "tweaking" the settings.
What I'm trying to understand is how to translate the 9.8cms/EPR-3 to a useful bilevel analogue without getting horrific aerophagia, (I think EPAP 9.8 might be too much for me with any amount of PS) and how to potentially set relatively more-stable pressures.
As Miss Emerita points out: The corresponding settings to fixed 9.8 with EPR = 3 on the AirSense AutoSet would be to use: Min EPAP = 6.8, Max IPAP = 9.8, PS = 3.
Since aerophagia continues to plague you and since your obstructive stuff was well controlled on the AutoSet at 9.8 with EPR = 3, I would offer these settings as potentially useful settings to try on the AirCurve for the next 3-5 days:
Min EPAP = 6, Max IPAP = 10, PS = 4
The additional PS may make exhalation easier, which can help reduce the aerophagia. Choosing these settings will reduce the exhale pressure a bit over what you were used to on the AutoSet and increase the inhalation pressure a bit. But when PS = Max IPAP - Min EPAP, that will prevent the AirCurve from increasing the pressure in response to events, snoring, or flow limitations.
Now the small increase in PS along with the small decrease in min EPAP (from your AirSense AutoSet),
may increase the number of CAs scored. But if the CAI remains well below 5.0 and the AHI remains below 5.0, the tradeoff of more CAs with less aerophagia may increase your comfort during the night. And increased comfort typically means less aerophagia and fewer spontaneous arousals and a better overall quality of sleep.
But the important thing is to stick with the settings for at least a few days. Don't let a sudden increase in the CAI from 0.23ish to 2.0ish spook you if it happens on the first day or two of using new settings with a larger PS.
I know sometimes aerophagia can just require some time to adjust, but until I'm adjusted I need my sleep! I'm not trying to settings-surf, just find something that seems it is remaining effective through the night without the major wakeups by pressure-swings. I was thinking tonight I'll keep the 8 EPAP, seems to be a bit more effective, IPAP 11~12, PS 3 or 4? Grateful for any insights at all.
Again, dial wingin' (or settings-surfing) seldom fixes things unless you have a specific problem in mind.
Since aerophagia
IS an issue, some dial wingin' to fix the aerophagia is in order. Since your obstructive stuff was well controlled on the AutoSet with a pressure of 9.8 and EPR = 3, that means you can set the min EPAP to 9.8-3 = 6.8 and get the same pressure level you had on the AutoSet. But again: Since your obstructive stuff is so well controlled with an effective EPAP = 6.8 and aerophagia is a problem and you have the AirCurve, it is worth trying:
Min EPAP = 6
Max IPAP = 10
PS = 4
for a few nights. This will give your stomach more exhalation relief (less aerophagia) and the increased PS may make everything feel more comfortable. By setting PS = Max IPAP - Min EPAP, you prevent the AirCurve from doing the pressure swings.
Good luck