- Vpap S9.png (101.72 KiB) Viewed 12396 times
You old machine was not an Automatic Machine.
You did have an EPAP of 10 and an IPAP of 18
The majority of your apneas were central apneas
Contrary to what you think, you never used VAuto before.
VPAP is not VAUTO
https://www.resmed.com/en-us/sleep-apne ... -and-cpap/
The short answer: An abbreviation of variable positive airway pressure, VPAP™ is the ResMed brand name for a specialized form of positive airway pressure (PAP) that provides two continuous levels of pressure. It’s called “variable” because it varies between two different levels to better meet the needs of people with sleep apnea. Because of this double-level approach, VPAP is also commonly known as “bilevel” PAP. Again, bilevel literally refers to two levels of air pressure: one for exhaling, and one for inhaling.
VPAP belongs somewhere between CPAP and APAP on the PAP scale. It doesn’t adjust airway pressure automatically like an APAP device does, but it offers a greater variety of customization than standard CPAP technology.
I don't think your new machine was
ever set up to be identical to the old one.
According to last nights OSCAR,
you're still letting your machine go lower that the 18 IPAP you need - it rarely stays long beneath that 18, and every time it drops too low, you snore, probably wake up, and the pressure obligingly return to 18.
So let start with making sure
minimum IPAP pressure at 18, no less than that.
Do
not let your IPAP go lower than 18.
The question is how to do that.
I've never touched - let alone set up - a VPAP machine of any kind, but I have a vague memory of reading you set up EPAP and then PS.
The first thing I would do, for the sake of trouble shooting it to set up the new machine
identical to the old one. Which mean to forgo auto mode for a week, try min EPAP= Eax IPAP =10 and PS=8.
Until we know if it was the old machine's age, or if the old prescription no longer fits you, we won't know enough to get you feeling better.
Of course, if you immediately start having many CA's, there's no need to stick to that setting for a week.
Who decided your PS should be 8, and when was that determined?
Did anything in your life or health change at about the time you became dissatisfied with your bi-level therapy?
Bottom right of the OSCAR overview lets you choose the time period for the overview. Can you see in the overview screen when you started getting all those CA's? was it sudden or gradual?