Hello
I just got a new phillips respironics dream station bipap and I am wondering if it can run on 12 volts?
thank you
new phillips respironics dream station bipap...12v?
new phillips respironics dream station bipap...12v?
_________________
Mask: Swift™ LT Nasal Pillow CPAP Mask with Headgear |
Additional Comments: auto bipap bi-flex & humidifier are respironics/phillips. |
Re: new phillips respironics dream station bipap...12v?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
but that's enough about them.
Re: new phillips respironics dream station bipap...12v?
Look at the brick that powers your machine. What does it say on it? Somewhere you'll see the output voltage listed.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
Do not regret growing older. It is a privilege denied to many...
Re: new phillips respironics dream station bipap...12v?
If this is like my Philips Respironics Dream Station APAPs it can run directly on 12 volts if you use the proprietary DC connector cord. Note that the same cord works for both Dreamstation 1 and 2 machines. Confirm this with your instruction manual or find a detailed manual online. The connectors typically run around $30 each. They plug into a standard 12v cigarette lighter socket so depending on your battery set up you may need an additional adapter with one of those.
- chunkyfrog
- Posts: 34220
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: In the flood plain--waiting for swamp.
Re: new phillips respironics dream station bipap...12v?
You might need to turn off the heat on the humidifier,
if you are running off a battery--heat uses a lot of juice.
if you are running off a battery--heat uses a lot of juice.