General Discussion on any topic relating to CPAP and/or Sleep Apnea.
-
Pugsy
- Posts: 64014
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Wed May 31, 2023 8:21 pm
zonker wrote: ↑Wed May 31, 2023 8:12 pm
Pugsy wrote: ↑Wed May 31, 2023 6:56 pm
I would be returning the machine....what you see happening isn't normal...we don't know what is causing it and we can't fix it either.. I returned a Respironics machine for doing this very thing.
just for my own education, are you saying that because of these "breaks" showing up, we can't trust the data that oscar is presenting?
The breaks aren't normal at all. Anytime a machine does stuff that isn't normal or wonky I always hesitate to put much trust in the data shown. It's probably okay but I can't prove it and when I can't prove something.....I don't trust it.
I would maybe try apap mode and see if the multiple sessions are still present but I wouldn't be holding my breath that resolves the issue.
We could maybe try using ResScan but I suspect it will show the same thing. I don't think it is an OSCAR bug.
A test could be done....an awake test...use the machine for a couple of hours while reading or watching TV and then look at the data for those couple of hours and see if there are any multiple sessions and if someone is awake they would know if the machine just up and stopped blowing.
I think it is a bug in the machine's internal something somewhere.
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
-
ozij
- Posts: 10141
- Joined: Fri Mar 18, 2005 11:52 pm
Post
by ozij » Wed May 31, 2023 9:07 pm
palerider wrote: ↑Wed May 31, 2023 8:15 pm
Are you using the same SD card in both machines? might be a bad card.
Makes sense to me too.
I would even try a new profile in OSCAR, and also check whatever it is that is doing the card reading: I have to insert my card into a USB thingy, and would change it if got this crazy data. If your computer has an SD card reader of its own I'd get a usb reader. If not: new USB reader.
This data is simply too crazy.
New SD card
New profile
New way of getting data from card to computer
Did you get a new power supply when you had the machine replaced?
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
mhuq
- Posts: 22
- Joined: Mon May 15, 2023 3:54 pm
Post
by mhuq » Thu Jun 01, 2023 4:14 am
palerider wrote: ↑Wed May 31, 2023 8:15 pm
Are you using the same SD card in both machines? might be a bad card.
Yes I am using the same SD card. But the card is not faulty. Reading on the card agrees with MyAir.
-
ozij
- Posts: 10141
- Joined: Fri Mar 18, 2005 11:52 pm
Post
by ozij » Thu Jun 01, 2023 4:24 am
MyAir shows you the number of session?
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
Respirator99
- Posts: 341
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Post
by Respirator99 » Thu Jun 01, 2023 4:30 am
ozij wrote: ↑Thu Jun 01, 2023 4:24 am
MyAir shows you the number of session?
Yes, it refers to them as "mask on/off events".
-
Conrad
- Posts: 507
- Joined: Mon Jul 30, 2012 9:10 am
- Location: Northern Illinois
Post
by Conrad » Thu Jun 01, 2023 6:50 am
chunkyfrog wrote: ↑Wed May 31, 2023 3:27 pm
Conrad wrote: ↑Wed May 31, 2023 6:44 am
chunkyfrog wrote: ↑Tue May 30, 2023 11:04 pm
Unlike the makers of phones and computers,
Resmed tries to get it right--right off the bat.
Don't you wish everybody cared like that?
Apples and oranges.
How?
All consumer products.
The simple answer. Computers and phones need updates to protect them from hackers. Not too many hackers want to hack your cpap, so...
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
-
Dog Slobber
- Posts: 3960
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Post
by Dog Slobber » Thu Jun 01, 2023 7:18 am
mhuq wrote: ↑Thu Jun 01, 2023 4:14 am
palerider wrote: ↑Wed May 31, 2023 8:15 pm
Are you using the same SD card in both machines? might be a bad card.
Yes I am using the same SD card. But the card is not faulty. Reading on the card agrees with MyAir.
If you encountered the same problem on a different machine, the problem is likely the SD Card.
The "reading on the card" agreeing with MyAir, does not validate the integrity on the card. The display issues appear in the detail data, which isn't used to re-calculate the summary totals.
mhuq wrote: ↑Wed May 31, 2023 6:22 pm
I have tried multiple times to attach images without success.
Because you keep ignoring how to post images properly.
The image posting problem was addressed *numerous* times, and yes you did completely ignore it, and left the thread after ignoring the advice.
Advice given to you, from your other topics, on posting images, including links, that you *completely ignored*:
palerider wrote: ↑Fri May 19, 2023 6:07 pm
Once again, you haven't attached anything.
use imgur.com and read the pinned posts, including the ones that tell you that forum attachments *DO NOT WORK*
Posting multiple threads about the same thing is annoying.
chunkyfrog wrote: ↑Fri May 19, 2023 9:45 pm
The forum is free. The attachment stub has not worked for some time.
We do not expect it to be fixed--the forum is free.
Use an imaging host, like Imgur.
palerider wrote: ↑Sat May 20, 2023 12:11 am
And I told you to READ THE PINNED POSTS.
AND to use imgur.com
palerider wrote: ↑Fri May 19, 2023 6:05 pm
Since you didn't attach anything to the post (use imgur.com) all we can do is say "eh?"
-
Callen
- Posts: 87
- Joined: Tue Apr 25, 2023 8:52 pm
Post
by Callen » Thu Jun 01, 2023 10:06 am
mhuq wrote: ↑Wed May 31, 2023 12:42 pm
I was just wondering what causes gaps in data logging during therapy hours. I have been trying get to the bottom of the problem for the last three months without success. The problem I am having is not that common. That is why a lot of people think it is due to clock setting. But I am convinced it is not. If someone in Resmed sits down and analyzes the data from my machine, an answer is likely to be found.
Have you tried reformatting your SD card? Trying a different SD card? Are you importing files properly? Updating your computer's software? A different computer?
-
Callen
- Posts: 87
- Joined: Tue Apr 25, 2023 8:52 pm
Post
by Callen » Thu Jun 01, 2023 10:37 am
ozij wrote: ↑Wed May 31, 2023 9:07 pm
palerider wrote: ↑Wed May 31, 2023 8:15 pm
Are you using the same SD card in both machines? might be a bad card.
Makes sense to me too.
I would even try a new profile in OSCAR, and also check whatever it is that is doing the card reading: I have to insert my card into a USB thingy, and would change it if got this crazy data. If your computer has an SD card reader of its own I'd get a usb reader. If not: new USB reader.
This data is simply too crazy.
New SD card
New profile
New way of getting data from card to computer
Did you get a new power supply when you had the machine replaced?
Never mind my post - I didn't see these potential issues had already been discussed.
-
Pugsy
- Posts: 64014
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Thu Jun 01, 2023 10:48 am
FWIW....which I admit isn't proof of anything....all of the above ideas I did try with the Respironics machine that was doing the multiple session thing without any success. Remember I had one of those machines in my own personal possession and it did indeed go back to supplier and I was issued a new machine that didn't have that problem.
Is the ResMed similarly affected??? Maybe and all ideas are sure worth trying but I wouldn't be holding my breath here.
Sometimes machines are just plain buggy. Now I admit it sure isn't common but sure isn't impossible for it to be the machine either. Crap does happen.
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
-
ozij
- Posts: 10141
- Joined: Fri Mar 18, 2005 11:52 pm
Post
by ozij » Thu Jun 01, 2023 11:56 am
mhuq wrote: ↑Wed May 31, 2023 7:20 pm
Pugsy wrote: ↑Wed May 31, 2023 6:56 pm
I would be returning the machine....what you see happening isn't normal...we don't know what is causing it and we can't fix it either.. I returned a Respironics machine for doing this very thing.
My provider did change my machine once already. The problem with the first machine has become worse with the replacement. The break is not due to power outage because in that case each break would be followed by a ramp. Note that there is a ramp when we turn off the machine to go to the bathroom in the middle of the night.
Pugsy wrote: ↑Thu Jun 01, 2023 10:48 am
Sometimes machines are just plain buggy. Now I admit it sure isn't common but sure isn't impossible for it to be the machine either. Crap does happen.
mhuq wrote: The problem with the first machine has become worse with the replacement.
ozij wrote: ↑Wed May 31, 2023 9:07 pm
Did you get a new power supply when you had the machine replaced?
Question repeated since it was not answered: Did you get a new power supply when you had the machine replaced?
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
tyrinryan
- Posts: 133
- Joined: Wed Jun 26, 2019 1:46 pm
Post
by tyrinryan » Fri Jun 02, 2023 7:36 am
Is here a "sticky" on how actually to use imgur? I don't want to download it until I know how it works.
Once, I figured out how to post a picture direct from windows I think, but I have forgotten. Perhaps not? Is there a way to post a picture directly from windows?
-
chunkyfrog
- Posts: 34390
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Post
by chunkyfrog » Fri Jun 02, 2023 9:11 am
tyrinryan wrote: ↑Fri Jun 02, 2023 7:36 am
Is here a "sticky" on how actually to use imgur? I don't want to download it until I know how it works.
Once, I figured out how to post a picture direct from windows I think, but I have forgotten. Perhaps not? Is there a way to post a picture directly from windows?
No. Windows is on your computer, and the forum has no room for images from "home".
You have to use the cloud--like on an image hosting service.
Several are free--just do some reading.
I know it's a pain in the butt, but you either spend time--or money.
-
zonker
- Posts: 11048
- Joined: Fri Jun 19, 2015 4:36 pm
Post
by zonker » Fri Jun 02, 2023 11:25 am
tyrinryan wrote: ↑Fri Jun 02, 2023 7:36 am
Is here a "sticky" on how actually to use imgur? I don't want to download it until I know how it works.
unfortunately, no there isn't a sticky.
imgur isn't software, it's a website where you upload your oscar chart images. it stays there and you link to it from here.
please let me know if what i just said is gibberish and i'll try to explain it better.
-
palerider
- Posts: 32300
- Joined: Wed Dec 16, 2009 5:43 pm
- Location: Dallas(ish).
Post
by palerider » Fri Jun 02, 2023 11:49 am
tyrinryan wrote: ↑Fri Jun 02, 2023 7:36 am
Is here a "sticky" on how actually to use imgur? I don't want to download it until I know how it works.
It's a web page, you don't download it.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.