Latest Version of Resmed Airsense 10 Software
Latest Version of Resmed Airsense 10 Software
Can someone tell me the latest version of Resmed Airsense 10 software. My machine about says SX567-0401, which appears to be a version from 2019 or earlier.
Re: Latest Version of Resmed Airsense 10 Software
as palerider told you-
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34385
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Latest Version of Resmed Airsense 10 Software
Unlike the makers of phones and computers,
Resmed tries to get it right--right off the bat.
Don't you wish everybody cared like that?
Resmed tries to get it right--right off the bat.
Don't you wish everybody cared like that?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Latest Version of Resmed Airsense 10 Software
Apples and oranges.chunkyfrog wrote: ↑Tue May 30, 2023 11:04 pmUnlike the makers of phones and computers,
Resmed tries to get it right--right off the bat.
Don't you wish everybody cared like that?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
Do not regret growing older. It is a privilege denied to many...
- Dog Slobber
- Posts: 3960
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Latest Version of Resmed Airsense 10 Software
The latest firmware I have seen referenced for the Air10's is SX567-0401
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Re: Latest Version of Resmed Airsense 10 Software
I was just wondering what causes gaps in data logging during therapy hours. I have been trying get to the bottom of the problem for the last three months without success. The problem I am having is not that common. That is why a lot of people think it is due to clock setting. But I am convinced it is not. If someone in Resmed sits down and analyzes the data from my machine, an answer is likely to be found.
- Dog Slobber
- Posts: 3960
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Latest Version of Resmed Airsense 10 Software
You have started numerous topics.
People have been responding. You then ignore the responses and disappear.
People have been asking you to post graphs. You haven't.
People have advised you about the forum image capacity and work around. You ignore it.
Don't bother asking questions, if you're going to ignore the responses.
People have been responding. You then ignore the responses and disappear.
People have been asking you to post graphs. You haven't.
People have advised you about the forum image capacity and work around. You ignore it.
Don't bother asking questions, if you're going to ignore the responses.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
- chunkyfrog
- Posts: 34385
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Latest Version of Resmed Airsense 10 Software
How?Conrad wrote: ↑Wed May 31, 2023 6:44 amApples and oranges.chunkyfrog wrote: ↑Tue May 30, 2023 11:04 pmUnlike the makers of phones and computers,
Resmed tries to get it right--right off the bat.
Don't you wish everybody cared like that?
All consumer products.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Latest Version of Resmed Airsense 10 Software
I have tried multiple times to attach images without success. When did 8 become numerous. I read all replies and respond when necessary. Here is another attempt
. Can someone explain the breaks in the graphs?Re: Latest Version of Resmed Airsense 10 Software
numerous=many.
can't explain the gaps. someone with greater knowledge will take a swing at that.
what concerns me is that you are at a fixed pressure of nine. this isn't doing your therapy any favors. see how your pressure pressure raises to 9 and then can't go any higher? this is okay for your sleep before 2 am. but at 2 am, you start having obstructive events. that's what the OA means on the left side of the top events graph.
pressure needs to go higher but cannot because it's set to a fixed, immovable pressure of 9. if i were you, i'd set the machine to apap mode and set the lower pressure to at least 7 (if not 9) and the higher pressure at 20. (it won't go to 20 unless it needs to to control your events.)
does what i just said make sense? i'm not talking down to you. i merely want to make sure that i'm communicating with you.
if you have any questions, please ask.
good luck!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Latest Version of Resmed Airsense 10 Software
This looks like something we have seen in the past with a Respironics machine. We just called it the multiple session bug and we never could figure out what was causing it. I haven't seen it with a ResMed machine though.
Look at all the sessions....numerous sleep sessions and most being 1 to 3 minutes in duration.
So...either a major power problem at the electric outlet causing multiple outages or something wrong with the machine and/or internal software.
I would be returning the machine....what you see happening isn't normal...we don't know what is causing it and we can't fix it either.. I returned a Respironics machine for doing this very thing.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Latest Version of Resmed Airsense 10 Software
My provider did change my machine once already. The problem with the first machine has become worse with the replacement. The break is not due to power outage because in that case each break would be followed by a ramp. Note that there is a ramp when we turn off the machine to go to the bathroom in the middle of the night.Pugsy wrote: ↑Wed May 31, 2023 6:56 pmThis looks like something we have seen in the past with a Respironics machine. We just called it the multiple session bug and we never could figure out what was causing it. I haven't seen it with a ResMed machine though.
Look at all the sessions....numerous sleep sessions and most being 1 to 3 minutes in duration.
So...either a major power problem at the electric outlet causing multiple outages or something wrong with the machine and/or internal software.
I would be returning the machine....what you see happening isn't normal...we don't know what is causing it and we can't fix it either.. I returned a Respironics machine for doing this very thing.
Re: Latest Version of Resmed Airsense 10 Software
There is something wrong with this machine.
It needs to be returned under warranty issues.
You can't fix this problem.
It needs to be returned under warranty issues.
You can't fix this problem.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Latest Version of Resmed Airsense 10 Software
just for my own education, are you saying that because of these "breaks" showing up, we can't trust the data that oscar is presenting?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Latest Version of Resmed Airsense 10 Software
Are you using the same SD card in both machines? might be a bad card.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.