Using CPAP for years, always tired (oscar analysis?)
- Miss Emerita
- Posts: 3445
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Stick with 13.5 a few more nights, then bump it up to 14.5.
I'm also curious: do you like having the ramp on? Your starting ramp pressure of 4 is very low -- for many people, so low as to be uncomfortable. If you like the ramp feature, then keep it, but you might experiment with a higher starting ramp pressure.
It makes sense to me that you'd have more CAs with Flex turned on. The little boost in pressure that you get when you shift from exhalation to inhalation can wash out just enough CO2 to delay your "breathe now" neurochemical signalling. If the delay is 10 seconds or more, that will be scored as a CA. But I don't know why turning Flex off would allow your OAs to increase. Possibly just a coincidence?
I'm also curious: do you like having the ramp on? Your starting ramp pressure of 4 is very low -- for many people, so low as to be uncomfortable. If you like the ramp feature, then keep it, but you might experiment with a higher starting ramp pressure.
It makes sense to me that you'd have more CAs with Flex turned on. The little boost in pressure that you get when you shift from exhalation to inhalation can wash out just enough CO2 to delay your "breathe now" neurochemical signalling. If the delay is 10 seconds or more, that will be scored as a CA. But I don't know why turning Flex off would allow your OAs to increase. Possibly just a coincidence?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Thanks for the replies, I'll keep it at 13.5 tonight, then boost to 14.5 tomorrow.Miss Emerita wrote: ↑Tue May 30, 2023 11:21 amStick with 13.5 a few more nights, then bump it up to 14.5.
I'm also curious: do you like having the ramp on? Your starting ramp pressure of 4 is very low -- for many people, so low as to be uncomfortable. If you like the ramp feature, then keep it, but you might experiment with a higher starting ramp pressure.
It makes sense to me that you'd have more CAs with Flex turned on. The little boost in pressure that you get when you shift from exhalation to inhalation can wash out just enough CO2 to delay your "breathe now" neurochemical signalling. If the delay is 10 seconds or more, that will be scored as a CA. But I don't know why turning Flex off would allow your OAs to increase. Possibly just a coincidence?
I've never had an issue with the 4 ramp being too low, I can breathe just fine. I like it nice and low when I lay down and settle in.
As for the OAs with Flex off, it's definitely not coincidence. Going back to March, I had a run of OAs with Flex off. The OAs were 19 events, 16 events, 25, 31 events. The next night I had turned Aflex 2 on and got OA: 1 event, 3, 1, 5 events.
I think I'm gonna go back to Aflex 2, Cflex isn't doing me any good.
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
- Miss Emerita
- Posts: 3445
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Yes, ugh. Have you tried the suggestions made earlier about the leaks? As for AHI: I would suggest you make just one change at a time. So try one more night with 13.5, leaving Flex unchanged.
Next time you post a chart, could you turn off the calendar? That way we can see the stuff toward the bottom of the left panel, which includes Flex information, as I recall.
Next time you post a chart, could you turn off the calendar? That way we can see the stuff toward the bottom of the left panel, which includes Flex information, as I recall.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
I reread all the suggestions for leaks and none of them apply for my situation. My pillow is flat, I sleep on my side, that website doesn't sell nasal pillow liners and I looked up videos on how to reduce leaks on an airfit p10 and there aren't any. One thing I will try though is sizing down my nasal pillow from M to S. I also tightened both my straps.Miss Emerita wrote: ↑Wed May 31, 2023 11:35 amYes, ugh. Have you tried the suggestions made earlier about the leaks? As for AHI: I would suggest you make just one change at a time. So try one more night with 13.5, leaving Flex unchanged.
Next time you post a chart, could you turn off the calendar? That way we can see the stuff toward the bottom of the left panel, which includes Flex information, as I recall.
I'll post tomorrow with calendar off, I was looking at previous results and forgot to switch it off. Thanks for your help.
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Forgot to post last night's.
https://imgur.com/a/hwe7CuR
Leak was better. Woke up at 3:45am and couldn't get back to sleep. This happens a couple times a month.
https://imgur.com/a/hwe7CuR
Leak was better. Woke up at 3:45am and couldn't get back to sleep. This happens a couple times a month.
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
How'd you sleep last night?danissleepy wrote: ↑Fri Jun 02, 2023 6:35 amThur night:
https://imgur.com/a/0Ag3Uig
Bumped up to 14.5 minimum
Did the leaks wake you up?
What about after the last big leak around 4:15 and on...were you asleep?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Slept fine, minimal wakeups. No leaks woke me up. Been very tired all day. I woke up around 5am and by 10am was really tired. That's pretty standard for me though. Haven't had a restful night of sleep in 25 years at least.Pugsy wrote: ↑Fri Jun 02, 2023 6:38 amHow'd you sleep last night?danissleepy wrote: ↑Fri Jun 02, 2023 6:35 amThur night:
https://imgur.com/a/0Ag3Uig
Bumped up to 14.5 minimum
Did the leaks wake you up?
What about after the last big leak around 4:15 and on...were you asleep?
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Whoa! One of my lowest AHIs ever. Not sure what happened but hopefully it keeps up. I definitely feel more awake than usual.
https://imgur.com/a/CWwy148
https://imgur.com/a/CWwy148
- Miss Emerita
- Posts: 3445
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Using CPAP for years, always tired (oscar analysis?)
It's very good to hear you're feeling more rested today! Stick with these settings and keep working on the leaks. I'll be curious whether the smaller P10 pillows help. If they don't, you may need a mask with a little more structure. (One that I liked was the Aloha, but of course that doesn't mean it'd work for you.)
Fingers crossed you're on a roll here.
Fingers crossed you're on a roll here.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Thanks! Really hoping this isn't a fluke but we'll see. I'll check out the aloha, I haven't heard of it.Miss Emerita wrote: ↑Sat Jun 03, 2023 11:15 amIt's very good to hear you're feeling more rested today! Stick with these settings and keep working on the leaks. I'll be curious whether the smaller P10 pillows help. If they don't, you may need a mask with a little more structure. (One that I liked was the Aloha, but of course that doesn't mean it'd work for you.)
Fingers crossed you're on a roll here.
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
OOF
https://imgur.com/a/11jadep
Dunno what to do with these leaks. I tried a 30i mask and they don't work for me. I have a thick beard that makes it hard to get a good seal. One time I tried it without the beard and the latex or whatever irritated my skin and made me break out really bad. Not to mention trying to sleep with the thing on. It's pretty much the airfit p10 or nothing so I've got to make it work.
Edit: I've not tried the Aloha and it looks like something I could use. Can't really afford it at the moment though.
https://imgur.com/a/11jadep
Dunno what to do with these leaks. I tried a 30i mask and they don't work for me. I have a thick beard that makes it hard to get a good seal. One time I tried it without the beard and the latex or whatever irritated my skin and made me break out really bad. Not to mention trying to sleep with the thing on. It's pretty much the airfit p10 or nothing so I've got to make it work.
Edit: I've not tried the Aloha and it looks like something I could use. Can't really afford it at the moment though.
- Miss Emerita
- Posts: 3445
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Using CPAP for years, always tired (oscar analysis?)
CAs are consistently inconsistent. They may just reflect a lot of arousals from sleep, and it's possible the arousals are connected with the major leakage. Hard to be sure. How are you feeling today?
I'm wondering whether you're having some mouth leaks. If you are, you might check out the Jay/Zonker Scunci method:
viewtopic/t184804/Full-face-mask-option ... lit=scunci
I use a Scunci over my mouth tape and keep it in place with two bobby-pins in back. It's very comfortable and completely eliminates mouth leaks.
For the P10, can you shave a tiny bit right under your nostrils to get a better seal? Or would that lead to the skin irritation?
I'm wondering whether you're having some mouth leaks. If you are, you might check out the Jay/Zonker Scunci method:
viewtopic/t184804/Full-face-mask-option ... lit=scunci
I use a Scunci over my mouth tape and keep it in place with two bobby-pins in back. It's very comfortable and completely eliminates mouth leaks.
For the P10, can you shave a tiny bit right under your nostrils to get a better seal? Or would that lead to the skin irritation?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
- Posts: 17
- Joined: Thu Apr 20, 2023 8:45 pm
Re: Using CPAP for years, always tired (oscar analysis?)
Yesterday felt pretty average. I'll check out that scunci method. I can't imagine having mouth leaks, I'm using that hypafix tape and it seals up super tight. I'm pretty sure the leak is just the seal from the pillow to my nostrils. When I wake up randomly in the night I can feel it leaking out of there. I think whether or not I have leaks just depends on how I shift around in the night and knock the mask pillow askew enough to leak.