New ASV. Oscar pics.
- ShallowBreather
- Posts: 73
- Joined: Thu Aug 19, 2021 11:34 am
New ASV. Oscar pics.
Link to my thread for my Aircurve 10 ST: viewtopic/t184243/Shallow-Breather-Osca ... shots.html
I have centrals, obstructives, and hypopneas. After a year on the ST machine, I still not getting the rest that I feel I need to get, so I went back to the my sleep doctor and asked if I could "pretty please" be considered for the ASV machine? Here are the titration results:
Total Sleep Time was 288 minutes resulting in a sleep efficiency of 67.6% with Total Recording Time of 426 minutes. 39 awakenings. Sleep latency 14.5 minutes and REM latency was 100 minutes. 2.8% supine. Sleep stages 4.3% N1, 77.3% N2, 5.6% N3, 12.8% REM. 150 arousals, resulting in arousal index of 31.3. 115 stage shifts.
ASV was initiated at EPAP min 3, EPAP max 15 and PS of 5. Snoring was eliminated at all pressures. 21 respiratory events: 0 apneas (0 obs, 0 mixed, 0 central) and 21 hypopneas. Mean O2 was 95% and min sat of 91%. 0% below 90% O2 sat. No Cheyne-Stokes. P10 with pillows and mouth tape.
EPAP PS RATE TST %Sup SupAHI REM RAHI CAI HI AHI ArIDX Nadr O2avg
3/15 5 0 106.5min -- -- 10.5min 5.7 0 2.8 2.8 29.9 91% 95%
3/15 6 0 25.0 min -- -- 0min -- 0 7.2 7.2 52.8 92% 96%
3/15 7 0 61.0 min -- -- 12.5min 0 0 4.9 4.9 19.7 93% 95%
3/15 8 0 57.0 min 14% 0 8min 7.5 0 3.2 3.2 15.8 92% 95%
3/15 9 0 38.5 min -- -- 6min 10 0 7.8 7.8 23.4 935 95%
No abnormal behavior noted. 11 periodic limb movement for a PLM index of 2.3. Heartrate was 71 beats per minute average, range 63-87. No arrhythmias.
Diagnosis: Obstructive Sleep Apnea; Treatment Emergent Central Apnea
Impression:
1. ResMed ASV at EPAP min 3, EPAP max pressure support of 5 to 9, auto-rate was used.
2. At ASV of EPAP min 3, EPAP max 15, PS 5 the apnea-hypopnea and arousal indices normalized, snoring eliminated, O2 maintained above 91%.
Recommendations:
1. ResMed ASV at EPAP min 6 and max 15, pressure support 5-8, and auto-rate. Follow-up in 1 month
*********
So, there it is. I was surprised to see the recommended min jump to 6 when we only tested 3 during the study. I was happy to see 0 obstructives and 0 centrals, but I'm a little concerned that I had so many hypopneas. Is the min of 6 a response to the hypopneas?
Thanks for looking,
SB
I have centrals, obstructives, and hypopneas. After a year on the ST machine, I still not getting the rest that I feel I need to get, so I went back to the my sleep doctor and asked if I could "pretty please" be considered for the ASV machine? Here are the titration results:
Total Sleep Time was 288 minutes resulting in a sleep efficiency of 67.6% with Total Recording Time of 426 minutes. 39 awakenings. Sleep latency 14.5 minutes and REM latency was 100 minutes. 2.8% supine. Sleep stages 4.3% N1, 77.3% N2, 5.6% N3, 12.8% REM. 150 arousals, resulting in arousal index of 31.3. 115 stage shifts.
ASV was initiated at EPAP min 3, EPAP max 15 and PS of 5. Snoring was eliminated at all pressures. 21 respiratory events: 0 apneas (0 obs, 0 mixed, 0 central) and 21 hypopneas. Mean O2 was 95% and min sat of 91%. 0% below 90% O2 sat. No Cheyne-Stokes. P10 with pillows and mouth tape.
EPAP PS RATE TST %Sup SupAHI REM RAHI CAI HI AHI ArIDX Nadr O2avg
3/15 5 0 106.5min -- -- 10.5min 5.7 0 2.8 2.8 29.9 91% 95%
3/15 6 0 25.0 min -- -- 0min -- 0 7.2 7.2 52.8 92% 96%
3/15 7 0 61.0 min -- -- 12.5min 0 0 4.9 4.9 19.7 93% 95%
3/15 8 0 57.0 min 14% 0 8min 7.5 0 3.2 3.2 15.8 92% 95%
3/15 9 0 38.5 min -- -- 6min 10 0 7.8 7.8 23.4 935 95%
No abnormal behavior noted. 11 periodic limb movement for a PLM index of 2.3. Heartrate was 71 beats per minute average, range 63-87. No arrhythmias.
Diagnosis: Obstructive Sleep Apnea; Treatment Emergent Central Apnea
Impression:
1. ResMed ASV at EPAP min 3, EPAP max pressure support of 5 to 9, auto-rate was used.
2. At ASV of EPAP min 3, EPAP max 15, PS 5 the apnea-hypopnea and arousal indices normalized, snoring eliminated, O2 maintained above 91%.
Recommendations:
1. ResMed ASV at EPAP min 6 and max 15, pressure support 5-8, and auto-rate. Follow-up in 1 month
*********
So, there it is. I was surprised to see the recommended min jump to 6 when we only tested 3 during the study. I was happy to see 0 obstructives and 0 centrals, but I'm a little concerned that I had so many hypopneas. Is the min of 6 a response to the hypopneas?
Thanks for looking,
SB
Last edited by ShallowBreather on Fri May 26, 2023 10:12 am, edited 1 time in total.
AirCurve 10 ASV
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
- ShallowBreather
- Posts: 73
- Joined: Thu Aug 19, 2021 11:34 am
Re: New Titration, Getting an ASV
Do you think that the ASB will do a good job handling all of those hypopneas?
AirCurve 10 ASV
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
- Respirator99
- Posts: 341
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: New Titration, Getting an ASV
Yes, a properly configured ASV will kill all apneas and hypopneas.
I suspect you'll need to change those settings once you have the machine. Resmed recommend the PSmax should be 15 or more. But try it out and see what works for you.
I suspect you'll need to change those settings once you have the machine. Resmed recommend the PSmax should be 15 or more. But try it out and see what works for you.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
- ShallowBreather
- Posts: 73
- Joined: Thu Aug 19, 2021 11:34 am
Re: New Titration, Getting an ASV
Thank you
AirCurve 10 ASV
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
- ShallowBreather
- Posts: 73
- Joined: Thu Aug 19, 2021 11:34 am
Re: New ASV, OSCAR pics. How does it look?
Good morning. I hope you all slept wonderfully last night.
I'm adding three nights of Oscar pics because I'm curious to know what the experts here on the board see.
As a non-expert, I'm wondering if my top pressure is high enough. It seems to be topping out at 16, but I thought my Rx max should be 15 top + 9 PS max or 24 as the absolute max pressure.
I'm loving that I'm not seeing any centrals, obstructives or even though nasty hypopneas. And I'm dreaming like there's no tomorrow. The best thing about the ASV versus the ST machine is that it doesn't feel like it's forcing you to breathe. The ST always felt like an iron lung. "You will breathe or else." The ASV is more subtle. It let's me breathe in, and then add just a little push to keep me out of hypopnea. It's a gentler experience. I told my wife, ASV is more like a good dance partner. ST was a drill sergeant.
Thanks for the insights in advance. If I need to zoom in on anything, I'm happy to do it.
Have a blessed day!!
I'm adding three nights of Oscar pics because I'm curious to know what the experts here on the board see.
As a non-expert, I'm wondering if my top pressure is high enough. It seems to be topping out at 16, but I thought my Rx max should be 15 top + 9 PS max or 24 as the absolute max pressure.
I'm loving that I'm not seeing any centrals, obstructives or even though nasty hypopneas. And I'm dreaming like there's no tomorrow. The best thing about the ASV versus the ST machine is that it doesn't feel like it's forcing you to breathe. The ST always felt like an iron lung. "You will breathe or else." The ASV is more subtle. It let's me breathe in, and then add just a little push to keep me out of hypopnea. It's a gentler experience. I told my wife, ASV is more like a good dance partner. ST was a drill sergeant.
Thanks for the insights in advance. If I need to zoom in on anything, I'm happy to do it.
Have a blessed day!!
- Attachments
-
- 5-25-2023.png (135.96 KiB) Viewed 2546 times
-
- 5-24-2023.png (140.27 KiB) Viewed 2546 times
-
- 5-23-2023.png (135.99 KiB) Viewed 2546 times
AirCurve 10 ASV
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
- Respirator99
- Posts: 341
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: New ASV. Oscar pics.
Looking good!
Your EPAP (inhale pressure) is fixed at 6.0, while pressure support can vary between 5 and 10. This will give an inhale pressure (IPAP) of 11 - 16. (IPAP = EPAP + PS). So the max IPAP of 16 is correct. Whether it's high enough is a question answered by another question - are you having many apneas and hypopneas? Quite clearly you're not, so yes, the pressure is high enough.
Then there is the most important question - how do you feel? Seems you're feeling pretty good, so all is well. Yes, the ASV is a fantastic machine - pure magic.
Your EPAP (inhale pressure) is fixed at 6.0, while pressure support can vary between 5 and 10. This will give an inhale pressure (IPAP) of 11 - 16. (IPAP = EPAP + PS). So the max IPAP of 16 is correct. Whether it's high enough is a question answered by another question - are you having many apneas and hypopneas? Quite clearly you're not, so yes, the pressure is high enough.
Then there is the most important question - how do you feel? Seems you're feeling pretty good, so all is well. Yes, the ASV is a fantastic machine - pure magic.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: New ASV. Oscar pics.
Pertinent question.Respirator99 wrote: ↑Fri May 26, 2023 8:30 pmThen there is the most important question - how do you feel?
This kinda caught my eye:
TWA.Total Sleep Time was 288 minutes resulting in a sleep efficiency of 67.6% with Total Recording Time of 426 minutes. 39 awakenings. Sleep latency 14.5 minutes and REM latency was 100 minutes. 2.8% supine. Sleep stages 4.3% N1, 77.3% N2, 5.6% N3, 12.8% REM. 150 arousals, resulting in arousal index of 31.3. 115 stage shifts.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: New ASV. Oscar pics.
I call it The Big Lie, but potato, potahto.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
- ShallowBreather
- Posts: 73
- Joined: Thu Aug 19, 2021 11:34 am
Re: New ASV. Oscar pics.
Thank youRespirator99 wrote: ↑Fri May 26, 2023 8:30 pmLooking good!
Your EPAP (inhale pressure) is fixed at 6.0, while pressure support can vary between 5 and 10. This will give an inhale pressure (IPAP) of 11 - 16. (IPAP = EPAP + PS). So the max IPAP of 16 is correct. Whether it's high enough is a question answered by another question - are you having many apneas and hypopneas? Quite clearly you're not, so yes, the pressure is high enough.
Then there is the most important question - how do you feel? Seems you're feeling pretty good, so all is well. Yes, the ASV is a fantastic machine - pure magic.
AirCurve 10 ASV
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
- ShallowBreather
- Posts: 73
- Joined: Thu Aug 19, 2021 11:34 am
Re: New ASV. Oscar pics.
TWA? Today was awful? This was amazing?
I think I was told before that I have Hiroshima sleep here on the boards. As in train wreck? In my defense, nobody could’ve slept in that place. The study was done in a second floor motel room. A parking lot was literally outside the window. Kids running in the hallway. Music. Ugh.
But, yeah I’m feeling a lot better…I think. I’ll give it some time. But it’s not like I can’t have it.
AirCurve 10 ASV
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Re: New ASV. Oscar pics.
All 4 of them?ShallowBreather wrote: ↑Sat May 27, 2023 4:55 pmIn my defense, nobody could’ve slept in that place. The study was done in a second floor motel room. A parking lot was literally outside the window. Kids running in the hallway. Music. Ugh.
The guy at the desk wasn't Georgie Santayana, was it.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: New ASV. Oscar pics.
"It?"
You mean "Hiroshima Sleep"?
OK.I’ll give it some time.
Meanwhile, I shall ponder parameters other than The Number:
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
- ShallowBreather
- Posts: 73
- Joined: Thu Aug 19, 2021 11:34 am
Re: New ASV. Oscar pics.
No, I mean the machine.
I’m don’t know much about this sleep stuff. But I know that I stop breathing when I’m asleep. And the ST machine was terrible. It had a good Number, but I didn’t feel as good in the morning.
The Number? I’m do check it, I’ll admit, but I’m more enjoying the dreams I get from the ASV. I didn’t get many with the ST.
AirCurve 10 ASV
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Min PS 5.0 Max Ps 10.0 EPAP 6.0
P10 pillows
Re: New ASV. Oscar pics.
Dream recall is generally not a good thing, as it's often (usually?) a sign that REM is fragmented and you're going REM > Wake.ShallowBreather wrote: ↑Sun May 28, 2023 4:11 pm... I’m more enjoying the dreams I get from the ASV. I didn’t get many with the ST.
Last edited by Rubicon on Mon May 29, 2023 3:37 am, edited 1 time in total.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.