Resmed 11 wakes me up with strong pressure and leaks
Resmed 11 wakes me up with strong pressure and leaks
Hello everyone, hope you can shed some lights on my case. I was waked up by my Resmed 11 a couple of times a night, and each time when I waked up by it, I found the air pressure is high (14), and there is some leaks around my mask. AHI is 6.
If I setup the max pressure lower (12), above issue is almost gone, but my AHI goes up (10).
Any advice will be greatly appreciated, especially on how to prevent Resmed from waking me up during night with high pressure.
Jon
If I setup the max pressure lower (12), above issue is almost gone, but my AHI goes up (10).
Any advice will be greatly appreciated, especially on how to prevent Resmed from waking me up during night with high pressure.
Jon
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
Re: Resmed 11 wakes me up with strong pressure and leaks
Where is your minimum pressure set at?
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: Resmed 11 wakes me up with strong pressure and leaks
I set min pressure at 5.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
Re: Resmed 11 wakes me up with strong pressure and leaks
Without any data, I'd guess your min should be 10, maybe more.
PS, 14 is far from 'strong', by lowering your max, you're preventing the machine from trying to stop your apneas, which is why your AHI is so bad.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 44
- Joined: Mon Feb 17, 2014 7:04 pm
Re: Resmed 11 wakes me up with strong pressure and leaks
a leak strap will seriously reduce your leaks
Max Pressure 14
Min Pressure 7
Min Pressure 7
Re: Resmed 11 wakes me up with strong pressure and leaks
Whatever a "leak strap" is, it probably won't.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Resmed 11 wakes me up with strong pressure and leaks
As suggested, last night I turned up min pressure from 5 to 10, and max from 14 to 16, and attached is my Oscar diagram, not pretty. (somehow this board won't let me attach the diagram)
Maybe I should further change the min and max?
Maybe I should further change the min and max?
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
Re: Resmed 11 wakes me up with strong pressure and leaks
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Resmed 11 wakes me up with strong pressure and leaks
I have the same setup with recent upgrade to Airsense 10 to Aisense 11. I use RemZzzs Full Face Cpap Mask Liners to reduce the leaks. They are effective\ for me, but costly. When set a 5 min and 18 high ( recommended by new pulmonologist) I have leak issues. My median pressure is 12 and always used on Aircurve 10. If I set 12 min and 18 high leaks are minimum and my AHI is less than 2. At 5 -18 my AHI varies from 5 - 9 with excessive leaks. With Oscar you can see we what your median pressure looks like using the dotted line “med” … I then set my min at that or close to it. The constant upsurge to prevent the OA/CA tends to induce leaks … perhaps I move around more because of the constantly changing pressure. Pressure surges fro 5 to 12.. 15 the recedes back toward 5 then ramps up repeatedly. Set at 12 the pressure swings are only ~3 vs 10. I would give a higher min pressure a try… perhaps “sneekup” on it 8, then 10, then 12. Also try the Pressure relief of 1,2, or 3. You can change settings yourself.
I have observed an issue with the Airsense 11 that I had not seen with the 10. The F20 mask valve tends to open and shut with every breath when the min pressure is below 8-10… make a noise and makes the mask move up/down on my face. Seems a function of the valve on the f20 mask since I never had the Aircurve 10 at at pressure level lower than 10
I have observed an issue with the Airsense 11 that I had not seen with the 10. The F20 mask valve tends to open and shut with every breath when the min pressure is below 8-10… make a noise and makes the mask move up/down on my face. Seems a function of the valve on the f20 mask since I never had the Aircurve 10 at at pressure level lower than 10
_________________
Machine: AirSense 11 Autoset |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
Additional Comments: Jan 17, 2016 Split Sleep study results w/o CPAP AHI=83, W/CPAP AHI=29. Pressure in APAP 15-20 cm |
The only way to keep your health is to eat what you don't want, drink what you don't like, and do what you'd rather not. Mark Twain
Was Mark Twain a CPAP user?
Was Mark Twain a CPAP user?
Re: Resmed 11 wakes me up with strong pressure and leaks
very helpful information, and thanks. I have already set at 3 for inhale/exhale difference, and believe the same that,upsurge of air pressure, not constant high pressure, are likely causing mask leaks, which in turn causes even higher pressure. Will bump up minimum pressure from 10 to 12, max at 18 (my median aire pressure from Oscar map is , and see how it works.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
Re: Resmed 11 wakes me up with strong pressure and leaks
What the sellers don't tell you is that you can just chuck 'em in the wash after use, and reuse them over and over and over.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Resmed 11 wakes me up with strong pressure and leaks
I don't know where you got the idea that leaks cause pressure to increase, because that is absolutely false.
Leaks don't affect the pressure the machine delivers until they get absolutely ludicrously large, at which point the machines will *REDUCE* the set pressure to try and get the mask to seal.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 44
- Joined: Mon Feb 17, 2014 7:04 pm
Re: Resmed 11 wakes me up with strong pressure and leaks
A leak strap really reduced my leaks.
https://www.padacheek.com/anti-leak-and ... ity-straps
Max Pressure 14
Min Pressure 7
Min Pressure 7
Re: Resmed 11 wakes me up with strong pressure and leaks
Well, if you'd SAID that you were talking about something Karen made instead of a generic "chin strap", then I'd probably said you might just be right.I've Been Tired wrote: ↑Wed May 03, 2023 8:12 pmA leak strap really reduced my leaks.
https://www.padacheek.com/anti-leak-and ... ity-straps
She makes good stuff.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 44
- Joined: Mon Feb 17, 2014 7:04 pm
Re: Resmed 11 wakes me up with strong pressure and leaks
You have 32,000 posts on here and you don't know what a leak strap is?!?! Good lord.palerider wrote: ↑Wed May 03, 2023 10:22 pmWell, if you'd SAID that you were talking about something Karen made instead of a generic "chin strap", then I'd probably said you might just be right.I've Been Tired wrote: ↑Wed May 03, 2023 8:12 pmA leak strap really reduced my leaks.
https://www.padacheek.com/anti-leak-and ... ity-straps
She makes good stuff.
Max Pressure 14
Min Pressure 7
Min Pressure 7