Resmed 10 Heat Plate

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
shaneo
Posts: 6
Joined: Mon Apr 17, 2023 3:39 am

Resmed 10 Heat Plate

Post by shaneo » Mon Apr 17, 2023 4:07 am

Hi my heat plate does not heat but everything else works -no alert on screen-if the fuse is blown should a fault be shown on screen-first l will test fuse easy bit first-if fuse is ok what should l check next and how-one last thing-if its the heat plate and l change it-is it plug and play or is there some calibration needed.Any help would be greatly appreciated-thanks

User avatar
LSAT
Posts: 13413
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resmed 10 Heat Plate

Post by LSAT » Mon Apr 17, 2023 7:43 am

Is the humidity turned on?

User avatar
Dog Slobber
Posts: 4339
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed 10 Heat Plate

Post by Dog Slobber » Mon Apr 17, 2023 8:26 am

I don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.

Remove all water from the the tank, put the tank back in, set the humidity to 8, and then run it for 10-15 minutes. Turn it off, remove tank and feel heater plate. Put a thermometer on it, and/or feel the plate.

If it's not hot, then yes, somethings wrong. might be the fuse.
Battery Backup: EcoFlow Delta 2

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 10 Heat Plate

Post by palerider » Mon Apr 17, 2023 4:53 pm

Dog Slobber wrote: ↑
Mon Apr 17, 2023 8:26 am
I don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.
Just what we need, a check engine light on the CPAP! :lol: :lol: :lol:
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11507
Joined: Fri Jun 19, 2015 4:36 pm

Re: Resmed 10 Heat Plate

Post by zonker » Mon Apr 17, 2023 5:25 pm

palerider wrote: ↑
Mon Apr 17, 2023 4:53 pm
Dog Slobber wrote: ↑
Mon Apr 17, 2023 8:26 am
I don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.
Just what we need, a check engine light on the CPAP! :lol: :lol: :lol:
maybe with an a-oo-gah! horn as well?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

shaneo
Posts: 6
Joined: Mon Apr 17, 2023 3:39 am

Re: Resmed 10 Heat Plate

Post by shaneo » Tue Apr 18, 2023 3:50 am

Dog Slobber wrote: ↑
Mon Apr 17, 2023 8:26 am
I don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.

Remove all water from the the tank, put the tank back in, set the humidity to 8, and then run it for 10-15 minutes. Turn it off, remove tank and feel heater plate. Put a thermometer on it, and/or feel the plate.

If it's not hot, then yes, somethings wrong. might be the fuse.
Hi fuse checked=ok - humidity was on one put to 4 plate heats up thanks for your advice

shaneo
Posts: 6
Joined: Mon Apr 17, 2023 3:39 am

Re: Resmed 10 Heat Plate

Post by shaneo » Tue Apr 18, 2023 4:05 am

Well had fuse checked=ok- humidity was on 1 put it on 4 plate heats up-thanks to all for your suggestions- one more thing if l need to replace the heat plate can l unplug the old plate and plug new plate in- without disturbing the calibration-thanks

User avatar
Dog Slobber
Posts: 4339
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed 10 Heat Plate

Post by Dog Slobber » Tue Apr 18, 2023 9:13 am

palerider wrote: ↑
Mon Apr 17, 2023 4:53 pm
Dog Slobber wrote: ↑
Mon Apr 17, 2023 8:26 am
I don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.
Just what we need, a check engine light on the CPAP! :lol: :lol: :lol:
Unfortunately for the OP, the Check Engine Light feature wasn't introduced until the Air 11.

Image
Battery Backup: EcoFlow Delta 2

Aretino
Posts: 109
Joined: Sun Mar 15, 2020 2:07 pm

Re: Resmed 10 Heat Plate

Post by Aretino » Tue Apr 18, 2023 9:29 am

The Air 11 need a code scanner as well?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed 10 Heat Plate

Post by palerider » Tue Apr 18, 2023 10:03 pm

Dog Slobber wrote: ↑
Tue Apr 18, 2023 9:13 am
palerider wrote: ↑
Mon Apr 17, 2023 4:53 pm
Dog Slobber wrote: ↑
Mon Apr 17, 2023 8:26 am
I don't think there is any kind of onscreen diagnostic displayed. Don't recall seeing anything reported.
Just what we need, a check engine light on the CPAP! :lol: :lol: :lol:
Unfortunately for the OP, the Check Engine Light feature wasn't introduced until the Air 11.

Image
:lol: :lol: :lol:
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Resmed 10 Heat Plate

Post by chunkyfrog » Wed Apr 19, 2023 3:37 pm

Sure beats pulling out the planchet and Oija Board.

Bugsmasher737
Posts: 1
Joined: Sun Nov 01, 2026 8:56 am

Re: Resmed 10 Heat Plate

Post by Bugsmasher737 » Sun Nov 01, 2026 9:03 am

I can't remember what it said but I'm pretty sure my Resmed 10 gave me a fault message when the heater plate went. I took it apart and could see that water over the years had been getting in and finally killed it. I've been looking for an economical replacement. But from what I can see the price of the part is too expensive to make it worth fixing the old machine. If anyone knows where it can get a R370-791 cheap please let me know.