Airtouch/Airfit F20 relief holes
- Marcusstephen
- Posts: 122
- Joined: Tue Feb 05, 2019 9:39 pm
- Location: Gold Coast Queensland Australia
Airtouch/Airfit F20 relief holes
I just noticed in my new airtouch attachment
that I couldn’t feel any relief (air) from the mask or the elbow and i think it explains why i’ve been waking feeling
not up to par since using it. I’m thinking that it’s likely i’m retaining too much carbon dioxide. I drilled a series of tiny holes either side of the disposable airtouch and it made therapy feel much more comfortable. I drilled 3 holes either side using the smallest drill bit that comes in a drill set .
Hope this may help someone
any advice welcome
that I couldn’t feel any relief (air) from the mask or the elbow and i think it explains why i’ve been waking feeling
not up to par since using it. I’m thinking that it’s likely i’m retaining too much carbon dioxide. I drilled a series of tiny holes either side of the disposable airtouch and it made therapy feel much more comfortable. I drilled 3 holes either side using the smallest drill bit that comes in a drill set .
Hope this may help someone
any advice welcome
BAD BABY
Re: Airtouch/Airfit F20 relief holes
There are about 20 holes to disburse your exhaled air. They are covered with a plastic cover to deflect the air so it doesn't blow in someone's face. This change was requested by many of the users. This is one of the most widely used full face masks. If there would be a problem with CO2 it would have been corrected. If you feel you have the found a way to make you breathe better...good for you.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... Buckwheat hull pillow |
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Airtouch/Airfit F20 relief holes
Note: if the holes have become clogged with dirt, a water pic can clean them.
I would refrain from adding more holes myself, as my masks work fine.
I would refrain from adding more holes myself, as my masks work fine.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- Marcusstephen
- Posts: 122
- Joined: Tue Feb 05, 2019 9:39 pm
- Location: Gold Coast Queensland Australia
Re: Airtouch/Airfit F20 relief holes
Thanks for the replies
I couldn’t find any release of air anywhere on my mask
The only place with vents is in the elbow and air fails to release from there as the one way valves direct the air straight up to the mask. I would love to know where
the vent holes are that you see in your mask. I’ve been using various masks now for 20 yrs and this is the first time i’ve been unable to find vents releasing air from the mask.
Thanks again i’m very grateful for your help
I couldn’t find any release of air anywhere on my mask
The only place with vents is in the elbow and air fails to release from there as the one way valves direct the air straight up to the mask. I would love to know where
the vent holes are that you see in your mask. I’ve been using various masks now for 20 yrs and this is the first time i’ve been unable to find vents releasing air from the mask.
Thanks again i’m very grateful for your help
BAD BABY
Re: Airtouch/Airfit F20 relief holes
Disconnect the elbow. Look carefully at where the elbow attaches to the mask. Look at it from the mask side.
The vent you're looking for is in the elbow.
Page 22 of the User Guide
The vent you're looking for is in the elbow.
Page 22 of the User Guide
The "vent assembly" is a ring of small holes through which exhaled air is diffused silently out of your mask.The mask must be fitted with the supplied elbow (containing the
valve and vent assembly) to ensure safe and functional usage
unless otherwise specified.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- Marcusstephen
- Posts: 122
- Joined: Tue Feb 05, 2019 9:39 pm
- Location: Gold Coast Queensland Australia
Re: Airtouch/Airfit F20 relief holes
Great thank youozij wrote: ↑Sat Apr 01, 2023 9:38 pmDisconnect the elbow. Look carefully at where the elbow attaches to the mask. Look at it from the mask side.
The vent you're looking for is in the elbow.
Page 22 of the User GuideThe "vent assembly" is a ring of small holes through which exhaled air is diffused silently out of your mask.The mask must be fitted with the supplied elbow (containing the
valve and vent assembly) to ensure safe and functional usage
unless otherwise specified.
fortunately i only drilled holes in the disposable
face cushion
Upon closer inspection i discovered a filter behind the vent holes which was blocking the release of air to some extent so i removed the filter . not as silent as before but therapy feels better
please see image or link
https://ibb.co/JdFyLGv
BAD BABY
- Dog Slobber
- Posts: 3954
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Airtouch/Airfit F20 relief holes
It's not a filter. Why would a mask system be designed to filter air that is constantly flowing out, trapping particles in the system?Marcusstephen wrote: ↑Sat Apr 01, 2023 10:40 pmUpon closer inspection i discovered a filter behind the vent holes which was blocking the release of air to some extent so i removed the filter
It's a diffuser. It's designed to diffuse the air, so it's quiet and doesn't form a stream that can blow on you or a bed partner.
Expected result.
You have been conditioned to associate the noise and feel of non-diffused venting with the venting. You then moved to a silent (diffused) system with venting that isn't easy to feel and you believed the venting was nolonger happening.
It was just fine.
But if you're ok removing the diffusing material, no problem.
With the original diffused venting, if you put the back of your hand *very close* to the vent port, you can barely feel the vented gas, you can even feel a temperature difference between inhaling and exhaling.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Re: Airtouch/Airfit F20 relief holes
Mercus: Would you be able to tell us whether the "leak rate" (as shown in Oscar) went up after you drilled the holes?
- Marcusstephen
- Posts: 122
- Joined: Tue Feb 05, 2019 9:39 pm
- Location: Gold Coast Queensland Australia
Re: Airtouch/Airfit F20 relief holes
Just checking on the AS10 result and it’s showing
7L/min AHI 1.5 That’s with the noise suppressing material removed but no holes
i’ll try the other with drilled holes tomorrow
BAD BABY
- Okie bipap
- Posts: 3550
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Airtouch/Airfit F20 relief holes
If you hold your hand real close to the elbow, ypo can detect a small difference in the temperature. It will be warmer when you exhale and cooler when you inhale.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
Re: Airtouch/Airfit F20 relief holes
This diffuser directs a small stream of air into the eyes. I want to try it from the knee without a diffuser.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Also use a Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
- Marcusstephen
- Posts: 122
- Joined: Tue Feb 05, 2019 9:39 pm
- Location: Gold Coast Queensland Australia
Re: Airtouch/Airfit F20 relief holes
The dirty noise diffuser/filter definitely restricts the air as it was hugely more comfortable without the diffuser but noisy of course. I’ve reinstalled the diffuser and will try tonight with 3 small holes either side of the disposable airtouch .
AHI’s improved also without the noise diffuser for some reason unless a coincidence.
AHI’s improved also without the noise diffuser for some reason unless a coincidence.
Last edited by Marcusstephen on Sun Apr 02, 2023 7:24 pm, edited 1 time in total.
BAD BABY
- Marcusstephen
- Posts: 122
- Joined: Tue Feb 05, 2019 9:39 pm
- Location: Gold Coast Queensland Australia
Re: Airtouch/Airfit F20 relief holes
that’s good to knowOkie bipap wrote: ↑Sun Apr 02, 2023 12:02 pmIf you hold your hand real close to the elbow, ypo can detect a small difference in the temperature. It will be warmer when you exhale and cooler when you inhale.
BAD BABY
Re: Airtouch/Airfit F20 relief holes
Marcus: Where did you drill the holes--through the --plastic? Did you drill them at the bottom (over the lip)? or all around the opening? How many holes? 1/8" or 1/16" perhaps?-can you find out so we can calculate the area? What about the leak rate according to Oscar--did it go up. This is an interesting experiment; I think it would be helpful to get the data.
- Marcusstephen
- Posts: 122
- Joined: Tue Feb 05, 2019 9:39 pm
- Location: Gold Coast Queensland Australia
Re: Airtouch/Airfit F20 relief holes
Yes no problem
The leak rate according to the AS10 was 7 litres per minute.
I ended up putting the noise filter back as it was too noisy without it. I tried with the 3 small holes drilled either side of the bridge of the nose but that also was too noisy . Now the interesting part is that the noise filter was partially clogged as per photo in my first post . I found after reinserting the noise filter material that i could once again feel that small volume of exhaust air coming from the elbow . I do believe that the noise filter became a dirt filter and prevented a reasonable volume of air from being directed away . I now use an Airtouch without the drilled holes and it’s very quiet and certainly more comfortable.
Hope that information is useful to someone.
The leak rate according to the AS10 was 7 litres per minute.
I ended up putting the noise filter back as it was too noisy without it. I tried with the 3 small holes drilled either side of the bridge of the nose but that also was too noisy . Now the interesting part is that the noise filter was partially clogged as per photo in my first post . I found after reinserting the noise filter material that i could once again feel that small volume of exhaust air coming from the elbow . I do believe that the noise filter became a dirt filter and prevented a reasonable volume of air from being directed away . I now use an Airtouch without the drilled holes and it’s very quiet and certainly more comfortable.
Hope that information is useful to someone.
BAD BABY