AirCurve 10 ASV pressure settings for pillows vs full face
- MichaelWD64
- Posts: 59
- Joined: Sat Dec 12, 2015 6:40 pm
AirCurve 10 ASV pressure settings for pillows vs full face
EDIT: Added new post with OSCAR and pulse-ox F50 data here:
viewtopic.php?f=1&t=186391&p=1436695#p1436695
Hi,
I've been using full-face for obstructive and central apnea for 7 years. Am well treated, it's working. But sometimes I feel claustrophobic with the full face mask.
My wife got on CPAP recently and she is using nasal pillows, not a full-face mask. I tried hers and like it, so ordered one for me. Her machine is not bilevel, so her settings won't work for me.
I am adept at changing clinical settings on my machine. My full-face mask settings are WAY too high with pillows. What would you recommend I change to to try with pillows? If I get in the ballpark, I can figure out from use what to use exactly. I've changed the setting to Pillows.
Currently for full-face I'm using
Min EPAP: 14
Max EPAP: 15
Min PS: 0
Max PS: 11
Thank you,
Michael
viewtopic.php?f=1&t=186391&p=1436695#p1436695
Hi,
I've been using full-face for obstructive and central apnea for 7 years. Am well treated, it's working. But sometimes I feel claustrophobic with the full face mask.
My wife got on CPAP recently and she is using nasal pillows, not a full-face mask. I tried hers and like it, so ordered one for me. Her machine is not bilevel, so her settings won't work for me.
I am adept at changing clinical settings on my machine. My full-face mask settings are WAY too high with pillows. What would you recommend I change to to try with pillows? If I get in the ballpark, I can figure out from use what to use exactly. I've changed the setting to Pillows.
Currently for full-face I'm using
Min EPAP: 14
Max EPAP: 15
Min PS: 0
Max PS: 11
Thank you,
Michael
_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Additional Comments: Machine: Resmed Aircurve 10 ASVauto bilevel with heated humidifier |
Last edited by MichaelWD64 on Wed Mar 22, 2023 10:38 pm, edited 2 times in total.
Got on CPAP 2015, AHI 147. Much better now. Have central & obstructive apnea. Smoked 30 years, quit when got on O2 & CPAP. I'm not on O2 during day anymore, only when I sleep, into CPAP machine. Ahhhh....oxygen! Refreshing!
- ChicagoGranny
- Posts: 14468
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: AirCurve 10 ASV pressure settings for pillows vs full face
How do you know that? Were you breathing through your nose and mouth when using the FFM?MichaelWD64 wrote: ↑Tue Mar 21, 2023 2:31 amMy full-face mask settings are WAY too high with pillows.
Be sure to change the mask setting to pillows - page 15 - https://www.respshop.com/manuals/37212_ ... n_menu.pdf
OSCAR can be a great help in finding the best pressure settings. Do you use it?
- MichaelWD64
- Posts: 59
- Joined: Sat Dec 12, 2015 6:40 pm
Re: AirCurve 10 ASV pressure settings for pillows vs full face
I can tell my Mask settings are too high for Pillows because it feels too high, like I wouldn't be able to sleep and I have trouble breathing out. Also my wife said it seemed way too high compared to hers.
I mostly breathed through my nose with full face, but sometimes my mouth and nose. Breathing just through my nose isn't a problem.
I changed the Mask setting from Full Face to Pillows before posting. I've edited my post to add that.
I use SleepyHead, which OSCAR is based on.
I can probably fine tune this using how it feels plus SleepyHead data but I need a reasonable setting for pillows to start with.
Is a Full Face setting supposed to work for Pillows? I would think they'd be different pressures.
Thank you.
I mostly breathed through my nose with full face, but sometimes my mouth and nose. Breathing just through my nose isn't a problem.
I changed the Mask setting from Full Face to Pillows before posting. I've edited my post to add that.
I use SleepyHead, which OSCAR is based on.
I can probably fine tune this using how it feels plus SleepyHead data but I need a reasonable setting for pillows to start with.
Is a Full Face setting supposed to work for Pillows? I would think they'd be different pressures.
Thank you.
_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Additional Comments: Machine: Resmed Aircurve 10 ASVauto bilevel with heated humidifier |
Got on CPAP 2015, AHI 147. Much better now. Have central & obstructive apnea. Smoked 30 years, quit when got on O2 & CPAP. I'm not on O2 during day anymore, only when I sleep, into CPAP machine. Ahhhh....oxygen! Refreshing!
Re: AirCurve 10 ASV pressure settings for pillows vs full face
I'd recommend *the same settings*, since that's what you need.MichaelWD64 wrote: ↑Tue Mar 21, 2023 2:31 amI am adept at changing clinical settings on my machine. My full-face mask settings are WAY too high with pillows. What would you recommend I change to to try with pillows?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: AirCurve 10 ASV pressure settings for pillows vs full face
"Feelings" are usually wrong.MichaelWD64 wrote: ↑Tue Mar 21, 2023 7:15 amI can tell my Mask settings are too high for Pillows because it feels too high,
Sleepyhead is *DEAD*, killed by it's creator. Plus it has a number of bugs that have been fixed in OSCAR.
Yes, there's no such thing as "full face settings" vs "pillow settings". In other words, you think incorrectly.MichaelWD64 wrote: ↑Tue Mar 21, 2023 7:15 amIs a Full Face setting supposed to work for Pillows? I would think they'd be different pressures.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- ChicagoGranny
- Posts: 14468
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: AirCurve 10 ASV pressure settings for pillows vs full face
Then your pressure settings with the FFM would be the correct settings for nasal pillows.
Don't judge one person's pressure needs by what another person needs.MichaelWD64 wrote: ↑Tue Mar 21, 2023 7:15 amAlso my wife said it seemed way too high compared to hers.
If you could start using OSCAR, and post a Daily Details chart of a FFM night in this thread (Don't start another one.), some ASV users could take a look and give you some more detailed advice.
BTW, you don't have the correct machine in your equipment profile.
- MichaelWD64
- Posts: 59
- Joined: Sat Dec 12, 2015 6:40 pm
Re: AirCurve 10 ASV pressure settings for pillows vs full face
Will try that.
And I've updated my gear in the signature line. Good call.
I think that even my settings for full face were too high. I recently got a new machine and wasn't comfortable using it since I got it. I was very comfortable with the old one. I'm going to call my equipment provider and see if they changed (upped) my settings from my old machine.
Then if I still have trouble, I'll get Oscar and post some data.
Thank you.
And I've updated my gear in the signature line. Good call.
I think that even my settings for full face were too high. I recently got a new machine and wasn't comfortable using it since I got it. I was very comfortable with the old one. I'm going to call my equipment provider and see if they changed (upped) my settings from my old machine.
Then if I still have trouble, I'll get Oscar and post some data.
Thank you.
_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Additional Comments: Machine: Resmed Aircurve 10 ASVauto bilevel with heated humidifier |
Got on CPAP 2015, AHI 147. Much better now. Have central & obstructive apnea. Smoked 30 years, quit when got on O2 & CPAP. I'm not on O2 during day anymore, only when I sleep, into CPAP machine. Ahhhh....oxygen! Refreshing!
- ChicagoGranny
- Posts: 14468
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: AirCurve 10 ASV pressure settings for pillows vs full face
Do you still have your old machine? Unless they are kaput, most of us keep our old machines for backup.
- MichaelWD64
- Posts: 59
- Joined: Sat Dec 12, 2015 6:40 pm
Re: AirCurve 10 ASV pressure settings for pillows vs full face
Yes, my old one gives the error of "past motor life", but works. I checked it a couple hours ago, the settings were the same they put for me on the new machine.
For some reason that pressure is now uncomfortable for me with mask or pillows.
I think it might be because I've just become a lot healthier over time. I got on CPAP when I was nearly dying, had just quit smoking after 30 years. Am a lot better now after not smoking and using CPAP for 7 years.
I have recording pulse ox, my O2 is fine with the lower pressure I'm trying with pillows.
Will talk to provider today, see what they say.
Thank you.
For some reason that pressure is now uncomfortable for me with mask or pillows.
I think it might be because I've just become a lot healthier over time. I got on CPAP when I was nearly dying, had just quit smoking after 30 years. Am a lot better now after not smoking and using CPAP for 7 years.
I have recording pulse ox, my O2 is fine with the lower pressure I'm trying with pillows.
Will talk to provider today, see what they say.
Thank you.
_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Additional Comments: Machine: Resmed Aircurve 10 ASVauto bilevel with heated humidifier |
Got on CPAP 2015, AHI 147. Much better now. Have central & obstructive apnea. Smoked 30 years, quit when got on O2 & CPAP. I'm not on O2 during day anymore, only when I sleep, into CPAP machine. Ahhhh....oxygen! Refreshing!
Re: AirCurve 10 ASV pressure settings for pillows vs full face
Learn how to get into the Clinical menu. Then you can change your own settings without the need to play phone-tag with a DME that most likely doesn't care.MichaelWD64 wrote: ↑Tue Mar 21, 2023 8:25 pmI think that even my settings for full face were too high. I recently got a new machine and wasn't comfortable using it since I got it. I was very comfortable with the old one. I'm going to call my equipment provider and see if they changed (upped) my settings from my old machine.
If you want to do dial wingin' on your own, you need to get Oscar and download the data regularly.Then if I still have trouble, I'll get Oscar and post some data.
Here's the thing: Your perception of the problem---the pressure setting is too high for your comfort---is important. And it is possible that a small reduction in minimum pressure by 1 or 2 cm might be enough to fix the comfort problem. But the question then becomes whether that small pressure reduction winds up letting too many events happen during the night. Oscar can tell you that quite quickly---far quicker than playing phone-tag with the DME and the doctor who prescribed the machine can.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Also use a P10 mask |
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls
- MichaelWD64
- Posts: 59
- Joined: Sat Dec 12, 2015 6:40 pm
Re: AirCurve 10 ASV pressure settings for pillows vs full face
Thank you.robysue1 wrote: ↑Wed Mar 22, 2023 8:01 amLearn how to get into the Clinical menu. Then you can change your own settings without the need to play phone-tag with a DME that most likely doesn't care.MichaelWD64 wrote: ↑Tue Mar 21, 2023 8:25 pmI think that even my settings for full face were too high. I recently got a new machine and wasn't comfortable using it since I got it. I was very comfortable with the old one. I'm going to call my equipment provider and see if they changed (upped) my settings from my old machine.
If you want to do dial wingin' on your own, you need to get Oscar and download the data regularly.Then if I still have trouble, I'll get Oscar and post some data.
Here's the thing: Your perception of the problem---the pressure setting is too high for your comfort---is important. And it is possible that a small reduction in minimum pressure by 1 or 2 cm might be enough to fix the comfort problem. But the question then becomes whether that small pressure reduction winds up letting too many events happen during the night. Oscar can tell you that quite quickly---far quicker than playing phone-tag with the DME and the doctor who prescribed the machine can.
I can get into the Clinical menu. That's how I changed what I was talking about in the first post.
My provider here (we moved last year) is more solid than some I've heard about.
Thank you.
_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Additional Comments: Machine: Resmed Aircurve 10 ASVauto bilevel with heated humidifier |
Got on CPAP 2015, AHI 147. Much better now. Have central & obstructive apnea. Smoked 30 years, quit when got on O2 & CPAP. I'm not on O2 during day anymore, only when I sleep, into CPAP machine. Ahhhh....oxygen! Refreshing!
- MichaelWD64
- Posts: 59
- Joined: Sat Dec 12, 2015 6:40 pm
Re: AirCurve 10 ASV pressure settings for pillows vs full face
ChicagoGranny wrote: ↑Tue Mar 21, 2023 1:17 pmIf you could start using OSCAR, and post a Daily Details chart of a FFM night in this thread
Installed OSCAR, ran it, and started a new post with OSCAR and pulse-ox F50 data here:
viewtopic.php?f=1&t=186391&p=1436695#p1436695
I started a new thread because I have more info to add and it would get confusing here if people read the new post but not the top one.
Thank you.
_________________
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Additional Comments: Machine: Resmed Aircurve 10 ASVauto bilevel with heated humidifier |
Got on CPAP 2015, AHI 147. Much better now. Have central & obstructive apnea. Smoked 30 years, quit when got on O2 & CPAP. I'm not on O2 during day anymore, only when I sleep, into CPAP machine. Ahhhh....oxygen! Refreshing!