Desperately Frustrated sleeper seeking help
-
- Posts: 8
- Joined: Thu Mar 16, 2023 10:56 pm
Desperately Frustrated sleeper seeking help
Hi. I’ve been on CPAP/APAP for years.
I had to get a replacement new ResMed Airsense 10 APAP machine on Feb 9 as the motor broke on the other machine. they gave me a new machine.
Since then, I’ve been waking every single day with a dry mouth. I’m also having more events than I’ve ever had in years. Before I got the new machine, no dry mouth, close to 0 events and I was using that machine since 2021. I have no clue what the humidity or tube temp was set on the other machine I returned. How do I fix this issue? No one seems to know when I ask around the community so I’m asking here. Nothing changed except a replacement machine. I’m using the Bleep mask, same mask I’ve been using since 2021. I do know what the sleep settings were on old machine and I changed the new machine to reflect this. This issues started that night so I know it’s the machine settings.
I also have a humidifier in my bedroom and I do use that. I also used that for the last year with the other machine and never encountered any issues with dry mouth and a lot of events per hour.
What do I need to set the tube temp at and the humidity level on the machine? The humidity in my bedroom gets to between 53-59%. It’s Arizona so really dry. I’m about to cry cuz this is so frustrating. I hope someone can please help guide me. Thanks so much
I had to get a replacement new ResMed Airsense 10 APAP machine on Feb 9 as the motor broke on the other machine. they gave me a new machine.
Since then, I’ve been waking every single day with a dry mouth. I’m also having more events than I’ve ever had in years. Before I got the new machine, no dry mouth, close to 0 events and I was using that machine since 2021. I have no clue what the humidity or tube temp was set on the other machine I returned. How do I fix this issue? No one seems to know when I ask around the community so I’m asking here. Nothing changed except a replacement machine. I’m using the Bleep mask, same mask I’ve been using since 2021. I do know what the sleep settings were on old machine and I changed the new machine to reflect this. This issues started that night so I know it’s the machine settings.
I also have a humidifier in my bedroom and I do use that. I also used that for the last year with the other machine and never encountered any issues with dry mouth and a lot of events per hour.
What do I need to set the tube temp at and the humidity level on the machine? The humidity in my bedroom gets to between 53-59%. It’s Arizona so really dry. I’m about to cry cuz this is so frustrating. I hope someone can please help guide me. Thanks so much
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Re: Desperately Frustrated sleeper seeking help
Did you have and SD card in your old machine? Did you keep it?
If you did, there's a program that will tell us everything about the settings on your old machine.
Please read the first announcement the forum, and follow the intructions:viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html
If you don't have the SD card from the old machine, check if you have one in the new machine. If you don't, buy one, insert it, keep it there when you're sleeping and upload the data as instructed.
The will help us see all present setting, and what's going on with your breathing as you sleep.
We need the data in order to help you figure out what's going on.
If you did, there's a program that will tell us everything about the settings on your old machine.
Please read the first announcement the forum, and follow the intructions:viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html
If you don't have the SD card from the old machine, check if you have one in the new machine. If you don't, buy one, insert it, keep it there when you're sleeping and upload the data as instructed.
The will help us see all present setting, and what's going on with your breathing as you sleep.
We need the data in order to help you figure out what's going on.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Desperately Frustrated sleeper seeking help
You could try various settings for the humidifier over a couple of nights each and see what works for you... just be careful not to congest your nose with a very high setting unless it works well for you.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: Desperately Frustrated sleeper seeking help
_________________
Machine: Airsense 10 Card to Cloud |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p
-
- Posts: 8
- Joined: Thu Mar 16, 2023 10:56 pm
Re: Desperately Frustrated sleeper seeking help
I've already read this and have been using Resmed Airsense for years, but thank you.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
-
- Posts: 8
- Joined: Thu Mar 16, 2023 10:56 pm
Re: Desperately Frustrated sleeper seeking help
The old SD card was in the old machine. I bought a new one. I posted them here. Thanks for reviewing them.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
-
- Posts: 8
- Joined: Thu Mar 16, 2023 10:56 pm
Re: Desperately Frustrated sleeper seeking help
Here is the data. I appreciate anyone looking at it. Thanks.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
-
- Posts: 8
- Joined: Thu Mar 16, 2023 10:56 pm
Re: Desperately Frustrated sleeper seeking help
Here is the data. Would love some guidance, anything, I need to sleep as I feel like I'm losing my mind.
I turned the humidity level up last night to a 4 and woke up with dry throat, headache, and water chamber close to empty (which has never ever happened before this new stupid machine).
Thanks.
I turned the humidity level up last night to a 4 and woke up with dry throat, headache, and water chamber close to empty (which has never ever happened before this new stupid machine).
Thanks.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Re: Desperately Frustrated sleeper seeking help
viewtopic/t183806/Attachment-storage-is ... MENTS.htmldoriluigi28jk92 wrote: ↑Fri Mar 17, 2023 5:07 pmHere is the data. Would love some guidance, anything, I need to sleep as I feel like I'm losing my mind.
I turned the humidity level up last night to a 4 and woke up with dry throat, headache, and water chamber close to empty (which has never ever happened before this new stupid machine).
Thanks.
and please stay in one thread.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 8
- Joined: Thu Mar 16, 2023 10:56 pm
Re: Desperately Frustrated sleeper seeking help
I don’t follow what you’re saying. I am in one thread.
Where do I post the Oscar data for someone to look at it?
Thx
Where do I post the Oscar data for someone to look at it?
Thx
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Re: Desperately Frustrated sleeper seeking help
Zonker is right, you created two threads:
This is the second one, unnecessary, and containing no data:
viewtopic/t186349/Sleep-Deprived--Despe ... -Help.html
This is the second one, unnecessary, and containing no data:
viewtopic/t186349/Sleep-Deprived--Despe ... -Help.html
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Desperately Frustrated sleeper seeking help
If you have OSCAR data from both machines, that's great.doriluigi28jk92 wrote: ↑Fri Mar 17, 2023 4:54 pmThe old SD card was in the old machine. I bought a new one. I posted them here. Thanks for reviewing them.
If you don't know how to post the images, Zonker will no doubt help you.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- chunkyfrog
- Posts: 34394
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Desperately Frustrated sleeper seeking help
You need to use an image hosting service, as direct attachments are impossible.
This is because the forum sponsors refuse to fix the problem.
This is because the forum sponsors refuse to fix the problem.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Desperately Frustrated sleeper seeking help
ozij wrote: ↑Fri Mar 17, 2023 11:30 pmIf you have OSCAR data from both machines, that's great.doriluigi28jk92 wrote: ↑Fri Mar 17, 2023 4:54 pmThe old SD card was in the old machine. I bought a new one. I posted them here. Thanks for reviewing them.
If you don't know how to post the images, Zonker will no doubt help you.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 8
- Joined: Thu Mar 16, 2023 10:56 pm
Re: Desperately Frustrated sleeper seeking help
Hi, here are the Oscar reports:
https://www.flickr.com/photos/197939421 ... d080r69815
I attached last 3 months and the one from last night. If this doesn't work, I give up. I'm too exhausted to try to figure this out. Thanks to whoever looks at this and lets me know what you see.
Recap:
1. Been using the same Resmed Airsense Auto APAP since 2021. Same Bleep mask, AHI's close to 0 and I was sleeping okay, everything fine.
2. Feb 9, 2023-the motor was going out, my DME gave me a new replacement machine; same model.
3. Since Feb 9th, haven't slept much at all. I'm about to lose my mind.
I stopped using Oscar after Sept 2022 as I was sleeping okay.
I started using it again when I got this new machine Feb 9, 2023.
Last night, humidity in my bedroom was 48%.
I lowered the min and max pressure settings and I was gasping for air, so I increased them back to what they were.
I had the tube temp on 79 and humidity level set to 5, ended up woken up with rainout.
I changed the tube temp to Auto and humidity to 4. That fixed the rainout.
Used Hyperfix mouth tape last night. Woke up dry mouth, dry nose, headache and dry and sore throat.
The night before last, I didn't run my humidifier and slept better, but still AHI's.
thanks.
https://www.flickr.com/photos/197939421 ... d080r69815
I attached last 3 months and the one from last night. If this doesn't work, I give up. I'm too exhausted to try to figure this out. Thanks to whoever looks at this and lets me know what you see.
Recap:
1. Been using the same Resmed Airsense Auto APAP since 2021. Same Bleep mask, AHI's close to 0 and I was sleeping okay, everything fine.
2. Feb 9, 2023-the motor was going out, my DME gave me a new replacement machine; same model.
3. Since Feb 9th, haven't slept much at all. I'm about to lose my mind.
I stopped using Oscar after Sept 2022 as I was sleeping okay.
I started using it again when I got this new machine Feb 9, 2023.
Last night, humidity in my bedroom was 48%.
I lowered the min and max pressure settings and I was gasping for air, so I increased them back to what they were.
I had the tube temp on 79 and humidity level set to 5, ended up woken up with rainout.
I changed the tube temp to Auto and humidity to 4. That fixed the rainout.
Used Hyperfix mouth tape last night. Woke up dry mouth, dry nose, headache and dry and sore throat.
The night before last, I didn't run my humidifier and slept better, but still AHI's.
thanks.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |