Bookbear wrote: ↑Wed Mar 01, 2023 3:20 pm
Any recent changes in medication, either the med or the dosage? A number of medications cause dry mouth. BTW, Xylamelts come in two flavors. There are other brands, but I only know of Xylamelts having the adhesive back that sticks to your cheek, gum, or roof of mouth. This is important so you do not swallow/choke on the tablet in your sleep. The adhesive itself melts away once the tab is gone. One lasts me about 4 hours.
Good luck!
Meds all the same but I will try to find those mints!
-
zonker
- Posts: 11048
- Joined: Fri Jun 19, 2015 4:36 pm
Post
by zonker » Wed Mar 01, 2023 6:38 pm
JSmithy wrote: ↑Wed Mar 01, 2023 3:55 pm
Meds all the same but I will try to find those mints!
the xlymelts should be in your grocery or drugstore in the dental hygiene section along side the toothpaste.
or if not there, on amazon.
they work a treat!
-
JSmithy
- Posts: 13
- Joined: Tue Feb 28, 2023 7:47 am
Post
by JSmithy » Wed Mar 01, 2023 7:39 pm
zonker wrote: ↑Wed Mar 01, 2023 6:38 pm
JSmithy wrote: ↑Wed Mar 01, 2023 3:55 pm
Meds all the same but I will try to find those mints!
the xlymelts should be in your grocery or drugstore in the dental hygiene section along side the toothpaste.
or if not there, on amazon.
they work a treat!
Thx. I ordered them on Amazon earlier. Not sure I need them but am curious about the adhesion that keeps you from choking during sleep but then dissolved too. Lol
-
Bookbear
- Posts: 1149
- Joined: Tue Apr 25, 2006 7:47 pm
-
Contact:
Post
by Bookbear » Wed Mar 01, 2023 11:08 pm
There are two sides to a xylamelt tab.... a white, domed side, and a tan, more flat side. It's the tan side that is the adhesive side and goes against your cheek, gum, or the roof of your mouth.
_________________
Machine | Mask | |
| |
Additional Comments: Avg. AHI .4 |
Getting old doesn't make you 'forgetful'. Having too damn many things to remember makes you 'forgetful'.
-
Pugsy
- Posts: 64026
- Joined: Thu May 14, 2009 9:31 am
- Location: Missouri, USA
Post
by Pugsy » Thu Mar 02, 2023 5:53 am
JSmithy wrote: ↑Wed Mar 01, 2023 1:02 pm
I really suspect it’s vaping that causes it. The nights I don’t get sticky mouth probably are edible nights. I prefer edibles but sometimes I forget to do one and they can take 45 minutes or more to work. Vape is close to two minutes.
You may have found your culprit here and it may just be medication side effects.
We don't necessarily think of THC stuff as a medication but it's just a drug like any other drug/medication that we might take and it has side effects as well.
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
-
Davy7
- Posts: 1
- Joined: Thu Mar 02, 2023 6:14 am
Post
by Davy7 » Thu Mar 02, 2023 6:22 am
I have the same problem. Mouth sticking .
Usually happens in the morning. But I know how powerful the stick is, very uncomfortable.
I'll try Biotene. I know it keeps your mouth moist. Will keep you posted on the results.
Thanks for the tip.
-
Conrad
- Posts: 509
- Joined: Mon Jul 30, 2012 9:10 am
- Location: Northern Illinois
Post
by Conrad » Thu Mar 02, 2023 9:05 am
JSmithy wrote: ↑Wed Mar 01, 2023 9:15 am
Never cared for any mouthwash. I don’t drink either so maybe it’s that alcohol taste. Last night was fine. It’s not an every night occurrence. Scares me when it happens though.
Say what? Alcohol taste? You do know that not every mouthwash contains alcohol, right? Biotene doesn't.
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
-
chunkyfrog
- Posts: 34397
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Post
by chunkyfrog » Thu Mar 02, 2023 9:59 am
Many mouthwashes produce dry mouth because of the alcohol they contain.
Look for a non-alcohol mouthwash.
-
JSmithy
- Posts: 13
- Joined: Tue Feb 28, 2023 7:47 am
Post
by JSmithy » Thu Mar 02, 2023 10:38 am
chunkyfrog wrote: ↑Thu Mar 02, 2023 9:59 am
Many mouthwashes produce dry mouth because of the alcohol they contain.
Look for a non-alcohol mouthwash.
It’s not mouthwash bc I don’t use it. But thank you.
-
Oughtsix
- Posts: 92
- Joined: Tue Jun 21, 2011 3:34 pm
Post
by Oughtsix » Thu Mar 02, 2023 12:10 pm
JSmithy wrote: ↑Tue Feb 28, 2023 7:51 am
Been on cpap for more than 20 yrs. Use nasal pillows since they came out. Last 6 months or so several times at night my mouth sticks together. It wakes me and is scary! I use a humidifier. I don’t mouth breathe. No mask leaks.
I sleep on my left side only. Can’t sleep in right side due to CGM sensor. I’ve slept on my side since I was little. Any ideas?
When I had this issue the heater in the cpap humidifier had died. I get a little bit of moisture from natural evaporation but not enough to keep everything moist.
When you get up in the morning is the heater under the water chamber warm to the touch or is it cold? Have you noticed if the water chamber has more water in it after a nights sleep than it used to before your mouth sticking problem?
_________________
Machine | Mask | |
| |
-
JSmithy
- Posts: 13
- Joined: Tue Feb 28, 2023 7:47 am
Post
by JSmithy » Thu Mar 02, 2023 12:55 pm
Oughtsix wrote: ↑Thu Mar 02, 2023 12:10 pm
JSmithy wrote: ↑Tue Feb 28, 2023 7:51 am
Been on cpap for more than 20 yrs. Use nasal pillows since they came out. Last 6 months or so several times at night my mouth sticks together. It wakes me and is scary! I use a humidifier. I don’t mouth breathe. No mask leaks.
I sleep on my left side only. Can’t sleep in right side due to CGM sensor. I’ve slept on my side since I was little. Any ideas?
When I had this issue the heater in the cpap humidifier had died. I get a little bit of moisture from natural evaporation but not enough to keep everything moist.
When you get up in the morning is the heater under the water chamber warm to the touch or is it cold? Have you noticed if the water chamber has more water in it after a nights sleep than it used to before your mouth sticking problem?
I hadn’t considered that. I only remove the humidifier chamber before bed to fill and weekly to clean but that’s early evening. Thank you! I will check that.
-
JSmithy
- Posts: 13
- Joined: Tue Feb 28, 2023 7:47 am
Post
by JSmithy » Thu Mar 02, 2023 10:56 pm
JSmithy wrote: ↑Thu Mar 02, 2023 12:55 pm
Oughtsix wrote: ↑Thu Mar 02, 2023 12:10 pm
JSmithy wrote: ↑Tue Feb 28, 2023 7:51 am
Been on cpap for more than 20 yrs. Use nasal pillows since they came out. Last 6 months or so several times at night my mouth sticks together. It wakes me and is scary! I use a humidifier. I don’t mouth breathe. No mask leaks.
I sleep on my left side only. Can’t sleep in right side due to CGM sensor. I’ve slept on my side since I was little. Any ideas?
When I had this issue the heater in the cpap humidifier had died. I get a little bit of moisture from natural evaporation but not enough to keep everything moist.
When you get up in the morning is the heater under the water chamber warm to the touch or is it cold? Have you noticed if the water chamber has more water in it after a nights sleep than it used to before your mouth sticking problem?
I hadn’t considered that. I only remove the humidifier chamber before bed to fill and weekly to clean but that’s early evening. Thank you! I will check that.
Just woke to pee after being in bed over 3.5 hrs. Humidifier plate ice cold!!! Problem solved. Thank you!
-
JSmithy
- Posts: 13
- Joined: Tue Feb 28, 2023 7:47 am
Post
by JSmithy » Fri Mar 03, 2023 7:54 am
JSmithy wrote: ↑Thu Mar 02, 2023 10:56 pm
[quote=JSmithy post_id=<a href="tel:1435451">1435451</a> time=<a href="tel:1677783331">1677783331</a> user_id=102700]
[quote=Oughtsix post_id=<a href="tel:1435448">1435448</a> time=<a href="tel:1677780612">1677780612</a> user_id=57291]
[quote=JSmithy post_id=<a href="tel:1435326">1435326</a> time=<a href="tel:1677592267">1677592267</a> user_id=102700]
Been on cpap for more than 20 yrs. Use nasal pillows since they came out. Last 6 months or so several times at night my mouth sticks together. It wakes me and is scary! I use a humidifier. I don’t mouth breathe. No mask leaks.
I sleep on my left side only. Can’t sleep in right side due to CGM sensor. I’ve slept on my side since I was little. Any ideas?
When I had this issue the heater in the cpap humidifier had died. I get a little bit of moisture from natural evaporation but not enough to keep everything moist.
When you get up in the morning is the heater under the water chamber warm to the touch or is it cold? Have you noticed if the water chamber has more water in it after a nights sleep than it used to before your mouth sticking problem?
[/quote]
I hadn’t considered that. I only remove the humidifier chamber before bed to fill and weekly to clean but that’s early evening. Thank you! I will check that.
[/quote]
Just woke to pee after being in bed over 3.5 hrs. Humidifier plate ice cold!!! Problem solved. Thank you!
[/quote]
JSmithy wrote: ↑Thu Mar 02, 2023 10:56 pm
JSmithy wrote: ↑Thu Mar 02, 2023 12:55 pm
Oughtsix wrote: ↑Thu Mar 02, 2023 12:10 pm
JSmithy wrote: ↑Tue Feb 28, 2023 7:51 am
Been on cpap for more than 20 yrs. Use nasal pillows since they came out. Last 6 months or so several times at night my mouth sticks together. It wakes me and is scary! I use a humidifier. I don’t mouth breathe. No mask leaks.
I sleep on my left side only. Can’t sleep in right side due to CGM sensor. I’ve slept on my side since I was little. Any ideas?
When I had this issue the heater in the cpap humidifier had died. I get a little bit of moisture from natural evaporation but not enough to keep everything moist.
When you get up in the morning is the heater under the water chamber warm to the touch or is it cold? Have you noticed if the water chamber has more water in it after a nights sleep than it used to before your mouth sticking problem?
I hadn’t considered that. I only remove the humidifier chamber before bed to fill and weekly to clean but that’s early evening. Thank you! I will check that.
Just woke to pee after being in bed over 3.5 hrs. Humidifier plate ice cold!!! Problem solved. Thank you!
EDIT: I turned humidifier up to 6 just because. Machine say heat okay. Woke a few hours later and plate was very warm and water chamber down quite a bit. I don’t recall when or why I lowered the humidity. I used to run it at 5. Possible 3 was too dry although as I originally said this does not happen every night! Well it’s a start for now!
-
Oughtsix
- Posts: 92
- Joined: Tue Jun 21, 2011 3:34 pm
Post
by Oughtsix » Mon Mar 27, 2023 11:42 pm
JSmithy wrote: ↑Thu Mar 02, 2023 10:56 pm
JSmithy wrote: ↑Thu Mar 02, 2023 12:55 pm
Oughtsix wrote: ↑Thu Mar 02, 2023 12:10 pm
JSmithy wrote: ↑Tue Feb 28, 2023 7:51 am
Been on cpap for more than 20 yrs. Use nasal pillows since they came out. Last 6 months or so several times at night my mouth sticks together. It wakes me and is scary! I use a humidifier. I don’t mouth breathe. No mask leaks.
I sleep on my left side only. Can’t sleep in right side due to CGM sensor. I’ve slept on my side since I was little. Any ideas?
When I had this issue the heater in the cpap humidifier had died. I get a little bit of moisture from natural evaporation but not enough to keep everything moist.
When you get up in the morning is the heater under the water chamber warm to the touch or is it cold? Have you noticed if the water chamber has more water in it after a nights sleep than it used to before your mouth sticking problem?
I hadn’t considered that. I only remove the humidifier chamber before bed to fill and weekly to clean but that’s early evening. Thank you! I will check that.
Just woke to pee after being in bed over 3.5 hrs. Humidifier plate ice cold!!! Problem solved. Thank you!
Great! I am happy to hear you found the problem! I have 2 Philips system one CPAPs with bad heaters. Luckily Philips is replacing them under the recall.