Posting this here for completeness.
Dano1465 wrote: ↑Tue Feb 28, 2023 4:57 pm
Hello
On the Oscar reports There is a header called statistics. There are numbers associated to Min, Med, 95% and Max. I'm not sure what numbers mean with regards to leak rate. Last night my numbers were min - 0, med - 5, 95% - 35 and max - 50. I'm not sure if these are good, bad or awful. I hope this makes sense.
I have attached a link
https://imgur.com/Qbmt94W
Thanks so much
Your data looks like this:
The units for leak rates are Liters/min, and this represents airflow being lost to the semi-closed, pressurized system comprising the blower, the hose, the mask, and your upper airway. Because you are using a PR DreamStation, your machine actually reports two sets of leak numbers: The
Total Leak Rate and the
Leak Rate. The Total Leak Rate includes the intentional leak rate for your mask that is designed to prevent you from rebreathing the CO2 you exhale with each breath. The Leak Rate is an approximation of the
excess leakage---i.e. the difference between the Total Leak Rate and the expected, intentional leak rate for your mask.
The Min Leak Rate, Median Leak Rate, 95% Leak Rate, and Max Leak Rate are Liters/min numbers have
statistical definitions. For a very detailed explanation of how Median and 95% numbers for a given data set are defined, go to my
blog entry about Average, Median, and 95% numbers
The medium length version:
The Min Leak Rate is the smallest leak rate recorded during the entire night. The fact that your Min Leak Rate = 0 L/min means there were some periods during the night when the only leaks detected were the
expected, intentional leaks designed into your mask.
The Median Leak Rate is the number where for 50% of the night, your leaks were AT or BELOW this number, and for 50% of the night, your leaks were AT or ABOVE this number. The fact that your Median Leak Rate = 5 L/min means that for 50% of the night (about 4 hours and 25 minutes), your excess leaks were AT or BELOW 5 L/min. And for 50% of the time (about 4 hours and 25 minutes), your excess leaks were AT or ABOVE 5 L/min.
Whether a Median Leak Rate of 5 L/min is acceptable depends on information we don't yet know because you haven't told us enough information. In general, leaks in the neighborhood of 5 L/min are so small that they don't affect the efficacy of therapy, regardless of how long they last. But many people find long lasting leaks of 5 L/min pesky in the sense that they wake up frequently to fiddle (unsuccessfully) with the mask. And all those mini-wakes needed to fiddle with the mask disrupt the quality of the sleep. And if that's the case, the person definitely needs to figure out a way to fix the leaks.
The 95% number is the number where for 95% of the night, your leaks were AT or BELOW this number and for 5% of the night, your leaks were AT or ABOVE this number. The fact that your 95% Leak Rate = 35 L/min means that for 95% of the night (about 8 hours and 25 minutes), your Leak Rat was AT or BELOW 35 L/min. And for 5% of the night (about 25 minutes) your Leak was AT or ABOVE 35 L/min. Since the "Large Leak territory" for PR machines is greater than that for Resmed machines, this means that it is highly
unlikely that your largest leaks lasted long enough to have a severe affect on the efficacy of your therapy. And sure enough, the Oscar report says that you spent 4.65% of the night (less than 20 minutes) with leaks that were large enough to be flagged as official Large Leaks. So once again, the importance of this number is whether the leaks you are experiencing are disrupting your sleep rather than affecting the efficacy of your therapy.
The Max Leak Rate is just the largest leak rate recorded during the entire night. At some point your leak rate reached 50 L/min. Since the Oscar report indicates that you did spend time in official Large Leak territory, that tells us that and excess leak rate of 50 L/min is large enough (at your pressures) to trigger a Large Leak Flag. Since the time spent in Large Leak territory is minimal, whether to worry about the leaks goes back to whether the leaks you are experiencing are disrupting your sleep rather than affecting the efficacy of your therapy.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.
Correct number of posts is 7250 as robysue + what I have as robysue1
Profile pic: Frozen Niagara Falls